Peter,

Be careful, some agencies have taken the 1.0 mm thickness requirement for
transformers (57.9.4.e, first dash) and applied it to potted components the
need reinforced insulation.  IEC 60950, I believe only requires 0.4 mm for
reinforced.


Ned Devine
Program Manager III
Entela, Inc.
3033 Madison Ave. SE
Grand Rapids, MI  49548

616 248 9671 Phone
616 574 9752 Fax
[email protected]  e-mail




-----Original Message-----
From: Peter Merguerian [mailto:[email protected]]
Sent: Friday, August 31, 2001 3:38 AM
To: 'Ned Devine'; [email protected]
Subject: RE: Medical application-Isolation voltage


Ned,

I agree with you except for one point. Most ac/dc convertors are
encapsulated. Therefore no internal claearances or creepages. Most likely
the external clearance/creepage of 5.0/8.0mm betwen the primary and
secondary pins will be met. The IEC 601/UL2601-1 does not have a requirement
for oven aging of the potting as in IEC950/UL60950. As such, I would have to
say that a 950 ac/dc convertor could pass as long as it meets the stricter
leakage current and dielectric strength requirements of the
IEC601-1/UL2601-1 standard. 

One final comment. Check the UL Conditions of Acceptability for the ac/dc
converor and the TUV or VDE license. Most of these convertors specify that
the abnormals were tested with a specific type fuse at the input and if any
other fuse is used in the end-product, it will be necesary to repeat the
abnormal tests. It is very important to use the correct or better fuse
characteristics in order to reduce the amount of testing required in the
end-product (ie if the convertor manufacturer is willing to share their
schematics and providing upotted samples to conduct the tests!).




PETER S. MERGUERIAN
Technical Director
I.T.L. (Product Testing) Ltd.
26 Hacharoshet St., POB 211
Or Yehuda 60251, Israel
Tel: + 972-(0)3-5339022  Fax: + 972-(0)3-5339019
Mobile: + 972-(0)54-838175






-----Original Message-----
From: Ned Devine [mailto:[email protected]]
Sent: Thursday, August 30, 2001 3:31 PM
To: [email protected]
Subject: RE: Medical application-Isolation voltage



Hi,

The voltages in IEC 60601-1 are rms.  See paragraph 20.4.b.

You need a converter that is certified for medical.  Not only is the
dielectric higher, but the spacing (8mm CD, 5mm AC) are higher.  Most power
supplies designed for IEC 60950 will not pass.


Ned Devine
Program Manager III
Entela, Inc.
3033 Madison Ave. SE
Grand Rapids, MI  49548

616 248 9671 Phone
616 574 9752 Fax
[email protected]  e-mail




-----Original Message-----
From: John Woodgate [mailto:[email protected]]
Sent: Wednesday, August 29, 2001 3:18 PM
To: [email protected]
Subject: Re: Medical application-Isolation voltage



I read in !emc-pstc that Ron <[email protected]> wrote (in <00d501c1309e$f4
[email protected]>) about 'Medical application-Isolation
voltage', on Wed, 29 Aug 2001:
>
>I have a medical application that requires "double insulation" from the
>Mains to an "applied part" according to IEC60601.  IEC60601 states that for
>a reference voltage of 250v and double insulation, the test voltage is
4000v
>for 1 min.  I am assuming that the voltages are RMS values.

Yes, unless another measure is specified. Look carefully: it may be a
peak value.
>
>How does one go about meeting the 4000v test when the highest rated ac-dc
>converter I can find is 3500v RMS  continuous.
>
Complain to the converter manufacturers that you can't buy their product
because it doesn't meet the standard. It's their problem, not yours!
-- 
Regards, John Woodgate, OOO - Own Opinions Only. http://www.jmwa.demon.co.uk

Eat mink and be dreary!

-------------------------------------------
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/

To cancel your subscription, send mail to:
     [email protected]
with the single line:
     unsubscribe emc-pstc

For help, send mail to the list administrators:
     Michael Garretson:        [email protected]
     Dave Heald                [email protected]

For policy questions, send mail to:
     Richard Nute:           [email protected]
     Jim Bacher:             [email protected]

All emc-pstc postings are archived and searchable on the web at:
    No longer online until our new server is brought online and the old
messages are imported into the new server.

-------------------------------------------
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/

To cancel your subscription, send mail to:
     [email protected]
with the single line:
     unsubscribe emc-pstc

For help, send mail to the list administrators:
     Michael Garretson:        [email protected]
     Dave Heald                [email protected]

For policy questions, send mail to:
     Richard Nute:           [email protected]
     Jim Bacher:             [email protected]

All emc-pstc postings are archived and searchable on the web at:
    No longer online until our new server is brought online and the old
messages are imported into the new server.

-------------------------------------------
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/

To cancel your subscription, send mail to:
     [email protected]
with the single line:
     unsubscribe emc-pstc

For help, send mail to the list administrators:
     Michael Garretson:        [email protected]
     Dave Heald                [email protected]

For policy questions, send mail to:
     Richard Nute:           [email protected]
     Jim Bacher:             [email protected]

All emc-pstc postings are archived and searchable on the web at:
    No longer online until our new server is brought online and the old 
messages are imported into the new server.

Reply via email to