Ron,

You are right, the test is 4000V RMS for one minute.

Medical device manufacturers often use a 1:1 isolation transformer to solve
this problem. There are many transformer manufacturers who have catalog
and/or custom isolation transformers that are designed to comply with IEC
60601-1. Their construction ensures compliance with the creepage, clearance,
leakage current and electric strength requirements of IEC 60601-1. The input
side can also be arranged so that they can easily be switched from 230VAC to
120VAC operation.

Note that any 'AC-DC converter' or transformer which is not specifically
designed for compliance with the medical standard, is very unlikely to
comply with this standard. Once your product is 'front-ended' by a medical
grade isolation transformer, you can then use other non-medical components
in many cases.

I hope this helps.


Jon Griver
http://www.601help.com The Medical Device Developers' Guide to IEC 60601-1


>
>
> I have a medical application that requires "double insulation" from the
> Mains to an "applied part" according to IEC60601.  IEC60601
> states that for
> a reference voltage of 250v and double insulation, the test
> voltage is 4000v
> for 1 min.  I am assuming that the voltages are RMS values.
>
> How does one go about meeting the 4000v test when the highest rated ac-dc
> converter I can find is 3500v RMS  continuous.
>
> Any help is appreciated.
>  Thanks.
>
>
>
>
> Ronald Cyprowski
> Vascor Inc.
> 566 Alpha Drive
> Pittsburgh PA 15238
> 412-963-7438 x139
> [email protected]
>
>
> -------------------------------------------
> This message is from the IEEE EMC Society Product Safety
> Technical Committee emc-pstc discussion list.
>
> Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/
>
> To cancel your subscription, send mail to:
>      [email protected]
> with the single line:
>      unsubscribe emc-pstc
>
> For help, send mail to the list administrators:
>      Michael Garretson:        [email protected]
>      Dave Heald                [email protected]
>
> For policy questions, send mail to:
>      Richard Nute:           [email protected]
>      Jim Bacher:             [email protected]
>
> All emc-pstc postings are archived and searchable on the web at:
>     No longer online until our new server is brought online and
> the old messages are imported into the new server.
>


-------------------------------------------
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/

To cancel your subscription, send mail to:
     [email protected]
with the single line:
     unsubscribe emc-pstc

For help, send mail to the list administrators:
     Michael Garretson:        [email protected]
     Dave Heald                [email protected]

For policy questions, send mail to:
     Richard Nute:           [email protected]
     Jim Bacher:             [email protected]

All emc-pstc postings are archived and searchable on the web at:
    No longer online until our new server is brought online and the old 
messages are imported into the new server.

Reply via email to