With regards to your questions:
Mobile or not does not make a difference. We manufacturer diagnostic
equipment and use "off the shelf" PC's to do the number crunching. CE
marking on the PC has always been an issue with us. About two years ago I
did a PC search regarding CE marked equipment under the MDD - no such luck!
I contacted UL in Northbrook and asked the question if they have ever
evaluated a PC under UL2601 - no luck! What we have done is used a line
isolation transformer for the PC, opto-isolated the RS232 to eliminate any
ground loop currents. Test the combination per 60601-1-1. We also assure
that the PC's, monitors and printers we re-sell have CE (low voltage
directive) with our equipment being CE marked under the MDD.
CE marking under the MDD can become quite involved, depending on what you do
in your facility you may need to pursue quality certification by a Notified
Body (NB). If you already have an NB I would definitely contact them
regarding this issue, if you don't I would get one, at least to consult
with!
With regards to 60601-1-2, there is a second draft of this document. I have
a copy which I received from a colleague. If there is a web site for the
draft I'm not aware of the address. Perhaps one of the other forum members
knows of a site and will share with us!
We just tested one of our systems, with the recommendation of our NB we
added 61000-4-11 (voltage dips), 61000-3-2 (harmonics), 61000-3-3 (flicker).
This addresses some of the changes (additions) coming with 60601-1-2. With
regards to EN 55011, we tested group 1, class A. I'm not aware of the Sweden
and Germany requirement for class B.
With regards to the "demo" equipment. You can distribute to Europe without
CE marking, however the unit needs to be clearly labeled as "Demo Use Only"
(or similar) and must be used only by your people or your authorized
representative / distributor. They must be made aware of any potential harm
that may be imposed upon the patient, operator or bystander or the
environment (EMI/RFI, landfill waste, etc.). You are still held responsible
for your equipment!
I hope this sheds some light on your issues, & good luck!
  



-----Original Message-----
From: Binnom, Cyril A [mailto:[email protected]]
Sent: Friday, April 27, 2001 9:46 AM
To: emc-pstc
Subject: Medical Directive



Group:

I am attempting to send ITE equipment to the European Union, and the unit
will be used in a hospital environment. I have no problem testing to the
Medical or EMC Directive,  whichever is applicable, but I am in need of some
clarification

My questions are:

*       I have been told that if the unit is mobile, (which it is) then the
Medical Directive applies. There is no patient interface but it will be in a
hospital environment.
*       Will EN 60601-1-2 (Medical Immunity) be upgraded in the near future
as I do not want to test to the current standard and have to go back in a
year or two and retest the unit, and if so is there a proposed standard for
review.
*       Does the unit under EN 55011 have to be tested to Class B or Class A
limits for hospital use. I have been told that in Sweden and Germany, Class
B is mandatory.
*       A demo has been requested A.S.A.P. Does the demo have to be CE mark,
(it will not be sold) or is there room or a  clause that allows this one
unit to be shipped to the EU while compliance is being completed. I am under
the assumption, that the demo will be used in the hospital during this time.

Regards,

Cyril A. Binnom Jr.
EMI/EMC Approvals Engineer
LXE, Inc.
(770) 447-4224 Ext. 3240
(770) 447-6928 Fax
[email protected]   

     



 



-------------------------------------------
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/

To cancel your subscription, send mail to:
     [email protected]
with the single line:
     unsubscribe emc-pstc

For help, send mail to the list administrators:
     Michael Garretson:        [email protected]
     Dave Heald                [email protected]

For policy questions, send mail to:
     Richard Nute:           [email protected]
     Jim Bacher:             [email protected]

All emc-pstc postings are archived and searchable on the web at:
    http://www.rcic.com/      click on "Virtual Conference Hall,"

-------------------------------------------
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/

To cancel your subscription, send mail to:
     [email protected]
with the single line:
     unsubscribe emc-pstc

For help, send mail to the list administrators:
     Michael Garretson:        [email protected]
     Dave Heald                [email protected]

For policy questions, send mail to:
     Richard Nute:           [email protected]
     Jim Bacher:             [email protected]

All emc-pstc postings are archived and searchable on the web at:
    http://www.rcic.com/      click on "Virtual Conference Hall,"


Reply via email to