Hi,

The scope of the MDD (see below) says nothing about whether a device is
mobile or not.  Please look at IEC 60601-1-1 for systems.  This gives
examples of what certifications devices need when used in a hospital.  i.e.
in the patient vicinity.

For the EMC questions, I am sure someone has this information.

If you place the device on the market or into service, it must have the CE
marking.  There is an Annex in the MDD (X) that talks about Clinical
Evaluations.  You might be able to use this.

Ned

Ned Devine
Program Manager III
Entela, Inc.
3033 Madison Ave. SE
Grand Rapids, MI  49548

616 248 9671 Phone
616 574 9752 Fax
[email protected]  e-mail








2.      For the purposes of this Directive, the following definitions shall
apply:

(a)     medical device' means any instrument, apparatus, appliance, material
or other article, whether used alone or in combination, including the
software necessary for its proper application intended by the manufacturer
to be used for human beings for the purpose of:

-       diagnosis, prevention, monitoring, treatment or alleviation of
disease,

-       diagnosis, monitoring, treatment, alleviation of or compensation for
an injury or handicap,

-       investigation, replacement or modification of the anatomy or of a
physiological process,

-       control of conception,

and which does not achieve its principal intended action in or on the human
body by pharmacological, immunological or metabolic means, but which may be
assisted in its function by such means;

-----Original Message-----
From: Binnom, Cyril A [mailto:[email protected]]
Sent: Friday, April 27, 2001 10:46 AM
To: emc-pstc
Subject: Medical Directive



Group:

I am attempting to send ITE equipment to the European Union, and the unit
will be used in a hospital environment. I have no problem testing to the
Medical or EMC Directive,  whichever is applicable, but I am in need of some
clarification

My questions are:

*       I have been told that if the unit is mobile, (which it is) then the
Medical Directive applies. There is no patient interface but it will be in a
hospital environment.
*       Will EN 60601-1-2 (Medical Immunity) be upgraded in the near future
as I do not want to test to the current standard and have to go back in a
year or two and retest the unit, and if so is there a proposed standard for
review.
*       Does the unit under EN 55011 have to be tested to Class B or Class A
limits for hospital use. I have been told that in Sweden and Germany, Class
B is mandatory.
*       A demo has been requested A.S.A.P. Does the demo have to be CE mark,
(it will not be sold) or is there room or a  clause that allows this one
unit to be shipped to the EU while compliance is being completed. I am under
the assumption, that the demo will be used in the hospital during this time.

Regards,

Cyril A. Binnom Jr.
EMI/EMC Approvals Engineer
LXE, Inc.
(770) 447-4224 Ext. 3240
(770) 447-6928 Fax
[email protected]   

     



 



-------------------------------------------
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/

To cancel your subscription, send mail to:
     [email protected]
with the single line:
     unsubscribe emc-pstc

For help, send mail to the list administrators:
     Michael Garretson:        [email protected]
     Dave Heald                [email protected]

For policy questions, send mail to:
     Richard Nute:           [email protected]
     Jim Bacher:             [email protected]

All emc-pstc postings are archived and searchable on the web at:
    http://www.rcic.com/      click on "Virtual Conference Hall,"

-------------------------------------------
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/

To cancel your subscription, send mail to:
     [email protected]
with the single line:
     unsubscribe emc-pstc

For help, send mail to the list administrators:
     Michael Garretson:        [email protected]
     Dave Heald                [email protected]

For policy questions, send mail to:
     Richard Nute:           [email protected]
     Jim Bacher:             [email protected]

All emc-pstc postings are archived and searchable on the web at:
    http://www.rcic.com/      click on "Virtual Conference Hall,"


Reply via email to