Ken,

I couldn't disagree with you more.  As a Professional Engineer in Oregon
who previously worked for an NRTL, I can assure you that the current state
law does not permit individual PE's to field label equipment - nor has it
ever, to my knowledge.  Equipment approval activity is limited specifically
to labs approved by the state of Oregon.  If there is a PE not associated
with a state-approved lab that is conducting this work, I would doubt
whether it met the requirement of the current law.  Intel, Weyerhaeuser and
others have argued unsuccessfully for this approach for years.

With regard to the original question, NRTL approval (through OSHA) is a
first line approval for most of the larger labs.  While it may be possible
to obtain approval with some jurisdictions without NRTL status, I would be
cautious about working with such a lab.  NRTL approval can then generally
be leveraged to obtain approvals in specific jurisdictions.  Some
jurisdictions accept the Federal NRTL designation as sufficient for
approving products in their area (some, such as Texas, have little
oversight of such activity in any case).  Others, like Oregon and North
Carolina, as well as probably the cities of Los Angeles and Chicago require
labs to meet additional requirements, as well as support periodic audits of
their activity in the jurisdiction.

Even within Oregon, however, the playing field is not perfectly level in
the application of the requirement for existing law.  The Portland
Metropolitan area sees far more enforcement of these requirements than the
rural, southeast corner of the state, for instance.  This may mean that in
addition to issues where the equipment can be installed in a facility in
Austin, TX without further evaluation, while Portland, OR may require it,
you also have the potential of installing a system in a community within
Oregon where these requirements are not universally applied, while you will
still run into difficulty within Portland (one of the most consistent
jurisdictions in the application of these requirements).

In the case of the equipment referenced by Mr. Selva, I would encourage him
to obtain the list of labs currently approved by the State of Oregon.  This
can be obtained on their website (
http://www.cbs.state.or.us/external/bcd/tag/advisory/fieldevalfirms.pdf).
This document contains both the list of labs - for listing - and field
evaluation firms that are approved by the state.  You will note that there
are no individuals, such as PE's listed in this document.  The engineering
firm Edan Engineering is the closest to this mark, but I believe you will
find that their pricing will be similar to the other firms and that their
activities follow the same requirements placed by state law.  Products
listed by an approved lab are generally accepted as long as they carry the
listing mark from the lab.  Products that have already left the
manufacturer's site without a listing mark will be required to be field
labeled.  Recognize that labs are approved for different categories of
equipment, so the correct category will have to be approved for the lab
that is selected to perform the work.

In short, however, I would encourage you to visit the website in question
and select the lab to perform this work with extreme care.  As someone who
has worked in this field for ten years (and is now on the other side of
things as the manufacturer), it is very easy to get lost in misinformation
about the specific requirements, especially when separated physically by
6-8 time zones and several thousand airmiles.  I would categorize the
comment about PE's performing equivalent approvals to be in the category of
misinformation.

Finally, I will mention that for a product that is not exceptionally unique
- that is, something that is produced in significant quantity - the view
has frequently been - and the law is written such - that this type of
product should be listed by an NRTL instead of being field evaluated.  If
the product in question - a spectrometer, I believe - fall into this
category, which I would think is likely, this could be an additional
barrier to using field evaluation for the approval of thie equipment.  This
has been an issue that is coming up more an more in recent years.

If you have specific questions about this process or wish to discuss
individual labs in Oregon, I would encourage you to contact me directly
either via e-mail or at the phone number shown below.

Regards,

Michael Garretson
Sr. Compliance Engineer
RadiSys Corporation
+1 503 615-1227


                                                                                
                                 
                    "Matsuda, Ken"                                              
                                 
                    <MATSUDAK@curti        To:     "'Pierre SELVA'" 
<[email protected]>, Forum         
                    sinst.com>             Safety-emc 
<[email protected]>                              
                    Sent by:               cc:                                  
                                 
                    owner-emc-pstc@        Subject:     RE: Safety in Oregon    
                                 
                    ieee.org                                                    
                                 
                                                                                
                                 
                                                                                
                                 
                    01/29/01 06:44                                              
                                 
                    AM                                                          
                                 
                    Please respond                                              
                                 
                    to "Matsuda,                                                
                                 
                    Ken"                                                        
                                 
                                                                                
                                 
                                                                                
                                 





For Oregon, you can actually have a PE field label your product( in your
case for UL's version of 1010---I think it is UL 3101), provided he/she
fills out the proper documentation , alot cheaper than an NRTL.


Regards,

Ken Matsuda


 -----Original Message-----
From: Pierre SELVA [mailto:[email protected]]
Sent: Saturday, January 27, 2001 6:48 AM
To: Forum Safety-emc
Subject: Safety in Oregon



Dear colleagues,

I would like to know if UL mark (or other certification mark) is required,
or mandatory in Oregon state, for safety purpose. This applies to a
laboratory equipment (spectrometer).

If you know some web links, I'll appreciate them.

Thanks a lot for your cooperation,

====================================
Pierre SELVA
2 route de la Grobelle
73000 JACOB BELLECOMBETTE
Tel : 33 (0) 6 60 52 04 96
Fax : 33 (0)6 61 37 87 48
e-mail : [email protected]
====================================




-------------------------------------------
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

To cancel your subscription, send mail to:
     [email protected]
with the single line:
     unsubscribe emc-pstc

For help, send mail to the list administrators:
     Jim Bacher:              [email protected]
     Michael Garretson:        [email protected]

For policy questions, send mail to:
     Richard Nute:           [email protected]







-------------------------------------------
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

To cancel your subscription, send mail to:
     [email protected]
with the single line:
     unsubscribe emc-pstc

For help, send mail to the list administrators:
     Jim Bacher:              [email protected]
     Michael Garretson:        [email protected]

For policy questions, send mail to:
     Richard Nute:           [email protected]

Reply via email to