What an EMC engineer who understands the physics of field-to-wire coupling
would say is that the operation of non-antenna connected electronics
associated with one subsystem will not be degraded by close proximity with
the non-antenna connected electronics of another subsystem.  Forget 10
meters.  Are the PCs in your office separated by 10 m?  Would you expect two
PCs stacked side-by-side or one on top of the other to interact in any
manner?  These are rhetorical questions.

About the blood pressure monitor example.  Not enough info here to back out
what is wrong, but basic logic theory says when the conclusion is
impossible, you must re-examine your assumptions.  If 92 dBuV/m were enough
to make the device malfunction, it would malfunction a lot and there would
have been enough trouble reports to get it fixed or withdrawn. And 92 dBuV/m
at 10 meters is SCREAMING!!!  I am on location and don't have FCC regs
easily available, but the limits stair-step around 40 dBuV/m at 3 m, per my
recollection.

Would I feel comfortable placing a CISPR compliant PC next to a medical
device qualified to 1 V/m?  There is an inherent (not planned) margin of
safety here that is many orders of magnitude.   The answer is absolutely
yes.  If there were a problem, I would expect it more to occur below 30 MHz,
at the power supply switching frequency, IF the medical device processed
extremely low levels of electrical signals and was poorly shielded.  But I
believe there are separate immunity requirements which cover this
eventuality as well.




on 1/5/02 12:23 PM, [email protected] at [email protected] wrote:

Dear Cortland 
People can't simply say: "ordinary semiconductors won't demodulate RF levels
produced by an unintentional radiator" ­ even the smallest amount of RF can
be demodulated ­ there are no hysteresis or threshold effects in a PN
semiconductor junction or FET that is biased into its conduction region (at
least not until you get below signal levels equivalent to less than a single
electron). 

What I am sure most engineers would really mean to say is:
"ordinary semiconductors exposed to RF levels from an information technology
product which is fully compliant with all relevant EMC emissions standards
and is at 10 metres distance will generally not demodulate a sufficient
level of interference to make an appreciable difference to most electronic
systems." 

Now we have a statement which has some scientific rigor and some engineering
validity to it. 
(Although I do worry that in Europe our harmonised EMC standards only test
emissions up to 1GHz, so what does that say about the possible emitted
fields strengths from a PC with a 1.2GHz clock frequency?)

Let's see if we can put some meat into this discussion with a real-life
example... 

I once tested a blood sample incubator for RF field immunity. The incubator
was used during screening programs (for cancer and other diseases) and kept
about 100 test tubes at 37.1C (normal blood temperature), while the reagents
in the test tubes changed colour. After 24 hours of incubation medical staff
would inspect the test tubes and write letters to people telling them they
were sick, or that they were clear of the disease. I don't know what
temperature tolerance the reagents had to give an accurate medical
diagnosis, so assume ±0.1C.

On the front panel of the incubator was a display of its temperature, which
was of course 37.1C. We found that field strengths as low as 1V/m would
cause the incubation temperature to range over full scale, from heaters
fully off (in which case the temperature would decline to ambient) to
maximum (in which case the water used to incubate the test tubes would
boil). 
We could use the RF test frequency to control the temperature between plus
and minus full scale over the frequency range 80 to 1000MHz at 1V/m (and did
not test beyond 1GHz).

Most worryingly, the front panel display would only show temporary
variations from its 37.1C when the RF field was turned off or on, and would
continue to show 37.1C even when the water in the incubator was stone cold
or actually boiling.

Most demodulation effects in bipolar and FET devices approximate to a square
law - for example a 1dB fall in the field strength (keeping everything else
constant) would typically result in a 2dB fall in the demodulated
'interference' error signal, as John Woodgate has recently pointed out.

If we assume that the 1V/m field strength was causing a 60C temperature
error, how low would we need to make the RF field to get down to the 0.1C
accuracy of the front panel display?

Assuming square-law characteristics for the device doing the demodulation I
calculate a field strength of around 40mV/m or 92dBmicrovolts/metre.

You will notice that I have been generous to the incubator and assumed that
the 1V/m field just about caused its temperature error to increase by 60C to
boil the water, whereas it could have been overdriving the internal circuits
by a considerable margin and still suffered a 60C error at 0.1V/m. We didn't
test this possibility as our focus was (as in most of these cases) on
quickly modifying the product so it passed the immunity test - which we
did.. 

92dBmicrovolts/metre is not a very high RF field level for a PC without any
EMC precautions at a distance of 10 metres.

How many people reading this would be now be quite happy to place even a
fully-compliant PC (compliant at 10 metres distance, that is) right next to
the unmodified incubator?

If it helps, imagine that it is your young daughter whose blood sample is in
the incubator to discover which drugs she needs to survive.

Shall we have a vote on how close we would be prepared to place the PC?
Might be interesting.

Let's not even think about the problems of proximity to cellphones and other
intentional radiators.

I didn't mention that the incubator was a small model used for mobile
screening, for installation in a truck adapted for medical screening
purposes which travels to various communities and parks there for a few days
while it tests the local people for disease - hardly a very well controlled
electromagnetic environment.

What does the above imply for similar incubators in countries that do not
have mandatory EMC immunity standards? Or for older incubators in the EU
that have never had to meet the EMC directive?

(Please don't reply with the old chestnut that "we haven't heard of any
problems so far, so everything must be OK" - people who should have known
better were using that phrase before September 11th. It is just not an
acceptable argument where safety issues are involved, as any expert in
safety law will tell you. Try: "I've been driving past that school at 40mph
for ten years and haven't hit a kid yet, so it must be safe mustn't it?" as
a test of the concept.)

Regards, Keith Armstrong

In a message dated 04/01/02 17:54:23 GMT Standard Time,
[email protected] writes:

Subj:Re: EMC-related safety issues
List-Post: [email protected]
Date:04/01/02 17:54:23 GMT Standard Time
From:    [email protected] (Cortland Richmond)
To:    [email protected]
CC:    [email protected]

I don't believe this is what people are saying here. What they are saying
is, ordinary semiconductors won't demodulate RF levels produced by an
unintentional radiator. Cortland
(What I write here is mine alone.
My employer does not
Concur, agree or else endorse
These words, their tone, or thought.)
 
 [email protected] wrote:
Does anyone else think that ordinary semiconductors doesn't respond to RF? I
have tested a product which was little more than an LM324 quad op-amp for RF
immunity using IEC 61000-4-3. This op-amp has a slew rate of 1V/micro-second
on a good day with the wind in its favour. It was housed in an unshielded
plastic enclosure. Demodulated noise that exceeded the (not very tough)
product specification were seen all the way up to 500MHz at a number of spot
frequencies that appeared to be due to the natural resonances of the input
and output cables. 