Dear Cortland
People can't simply say: "ordinary semiconductors won't demodulate RF levels 
produced by an unintentional radiator" – even the smallest amount of RF can 
be demodulated – there are no hysteresis or threshold effects in a PN 
semiconductor junction or FET that is biased into its conduction region (at 
least not until you get below signal levels equivalent to less than a single 
electron).

What I am sure most engineers would really mean to say is: 
"ordinary semiconductors exposed to RF levels from an information technology 
product which is fully compliant with all relevant EMC emissions standards 
and is at 10 metres distance will generally not demodulate a sufficient level 
of interference to make an appreciable difference to most electronic systems."

Now we have a statement which has some scientific rigor and some engineering 
validity to it. 
(Although I do worry that in Europe our harmonised EMC standards only test 
emissions up to 1GHz, so what does that say about the possible emitted fields 
strengths from a PC with a 1.2GHz clock frequency?)

Let's see if we can put some meat into this discussion with a real-life 
example...

I once tested a blood sample incubator for RF field immunity. The incubator 
was used during screening programs (for cancer and other diseases) and kept 
about 100 test tubes at 37.1C (normal blood temperature), while the reagents 
in the test tubes changed colour. After 24 hours of incubation medical staff 
would inspect the test tubes and write letters to people telling them they 
were sick, or that they were clear of the disease. I don't know what 
temperature tolerance the reagents had to give an accurate medical diagnosis, 
so assume ±0.1C.

On the front panel of the incubator was a display of its temperature, which 
was of course 37.1C. We found that field strengths as low as 1V/m would cause 
the incubation temperature to range over full scale, from heaters fully off 
(in which case the temperature would decline to ambient) to maximum (in which 
case the water used to incubate the test tubes would boil). 
We could use the RF test frequency to control the temperature between plus 
and minus full scale over the frequency range 80 to 1000MHz at 1V/m (and did 
not test beyond 1GHz). 

Most worryingly, the front panel display would only show temporary variations 
from its 37.1C when the RF field was turned off or on, and would continue to 
show 37.1C even when the water in the incubator was stone cold or actually 
boiling.

Most demodulation effects in bipolar and FET devices approximate to a square 
law - for example a 1dB fall in the field strength (keeping everything else 
constant) would typically result in a 2dB fall in the demodulated 
'interference' error signal, as John Woodgate has recently pointed out.

If we assume that the 1V/m field strength was causing a 60C temperature 
error, how low would we need to make the RF field to get down to the 0.1C 
accuracy of the front panel display? 

Assuming square-law characteristics for the device doing the demodulation I 
calculate a field strength of around 40mV/m or 92dBmicrovolts/metre.

You will notice that I have been generous to the incubator and assumed that 
the 1V/m field just about caused its temperature error to increase by 60C to 
boil the water, whereas it could have been overdriving the internal circuits 
by a considerable margin and still suffered a 60C error at 0.1V/m. We didn't 
test this possibility as our focus was (as in most of these cases) on quickly 
modifying the product so it passed the immunity test - which we did.

92dBmicrovolts/metre is not a very high RF field level for a PC without any 
EMC precautions at a distance of 10 metres. 

How many people reading this would be now be quite happy to place even a 
fully-compliant PC (compliant at 10 metres distance, that is) right next to 
the unmodified incubator? 

If it helps, imagine that it is your young daughter whose blood sample is in 
the incubator to discover which drugs she needs to survive.

Shall we have a vote on how close we would be prepared to place the PC? 
Might be interesting.

Let's not even think about the problems of proximity to cellphones and other 
intentional radiators. 

I didn't mention that the incubator was a small model used for mobile 
screening, for installation in a truck adapted for medical screening purposes 
which travels to various communities and parks there for a few days while it 
tests the local people for disease - hardly a very well controlled 
electromagnetic environment.

What does the above imply for similar incubators in countries that do not 
have mandatory EMC immunity standards? Or for older incubators in the EU that 
have never had to meet the EMC directive?

(Please don't reply with the old chestnut that "we haven't heard of any 
problems so far, so everything must be OK" - people who should have known 
better were using that phrase before September 11th. It is just not an 
acceptable argument where safety issues are involved, as any expert in safety 
law will tell you. Try: "I've been driving past that school at 40mph for ten 
years and haven't hit a kid yet, so it must be safe mustn't it?" as a test of 
the concept.)

Regards, Keith Armstrong

In a message dated 04/01/02 17:54:23 GMT Standard Time, 
[email protected] writes:

> Subj:Re: EMC-related safety issues
> Date:04/01/02 17:54:23 GMT Standard Time
> From:    [email protected] (Cortland Richmond)
> To:    [email protected]
> CC:    [email protected]
> 
> I don't believe this is what people are saying here. What they are saying 
> is, ordinary semiconductors won't demodulate RF levels produced by an 
> unintentional radiator. Cortland 
> (What I write here is mine alone. 
> My employer does not 
> Concur, agree or else endorse 
> These words, their tone, or thought.) 
>   
>   [email protected] wrote: 
> >> Does anyone else think that ordinary semiconductors doesn't respond to RF?
>>  I have tested a product which was little more than an LM324 quad op-amp 
>> for RF immunity using IEC 61000-4-3. This op-amp has a slew rate of 
>> 1V/micro-second on a good day with the wind in its favour. It was housed 
>> in an unshielded plastic enclosure. Demodulated noise that exceeded the 
>> (not very tough) product specification were seen all the way up to 500MHz 
>> at a number of spot frequencies that appeared to be due to the natural 
>> resonances of the input and output cables. 
>> 

Reply via email to