I was working off IEC 1000-4-6 (1996), Figures 2 and 5b, which cover direct
injection and show the EUT isolated above ground.  Figure 6 shows clamp
injection and EUT isolation above ground.  What you are referencing is a
special case in the annex when the cm 150 Ohm impedance cannot be met.  The
theory is, I believe, as I stated before.  Incidentally clamp injection is
fundamentally different than CDN injection.  CDN injects line-to-ground, and
clamp injects in series with the cable.  The standard specifically says that
when using a clamp installed between EUT and AE, the AE must be an equal or
lower impedance than the EUT.  This is opposite of what you want with a CDN.
My paper covers that issue in detail.  It appeared in the 1997 IEEE EMC
Symposium record, page 479.

Ken Javor

> From: "Dan Kinney (A)" <[email protected]>
> Date: Fri, 14 Feb 2003 16:22:37 -0500
> To: "'Ken Javor'" <[email protected]>, Oliver Betz
> <[email protected]>, [email protected]
> Subject: RE: Grounding of EUT in EN61000-4-6 conducted immunity test
> 
> I have a hard time following Mr. Javor's reasoning.  The standard provides
> the option to use current clamp injection.  Paragraph 7.3 states "The AE and
> EUT with clamp injection shall represent the functional installation
> conditions as close as possible e.g. either the EUT shall be connected to
> the ground reference plane or placed on an insulating support (see figures
> A.6 and A.7)."  Figures A.6 and A.7 both clearly indicate the EUT may be
> connected to ground.  And I would contend, must be connected to ground if
> the EUTs normal installation is such that the conductive case is connected
> to grounded metal structures.   Furthermore, if shielded signal are used in
> the system, the shields must be connected to ground if they are to serve
> their intended function.  Often, this is through the conductive case of the
> EUT.  I would be interested in reading the article Mr. Javor wrote on this
> topic.  If I misunderstand the standard, perhaps the article will show me
> the errors in my reasoning.  Is the article available sir?
> Dan Kinney
> Horner APG
> 
> 
> 
>> -----Original Message-----
>> From: Ken Javor [SMTP:[email protected]]
>> Sent: Thursday, February 13, 2003 1:46 PM
>> To: Oliver Betz; [email protected]
>> Subject: Re: Grounding of EUT in EN61000-4-6 conducted immunity test
>> 
>> 
>> I haven't seen a response to this yet, so I will take a stab at it.  Full
>> disclosure:  Except for the following paragraph, I am basically passing
>> along what I learned from others on this excellent forum not all that long
>> ago.
>> 
>> First off, my own comment.  My expert :-) opinion is that EN61000-4-6 is
>> really meant for equipment that was NOT designed to be electrically bonded
>> to a metallic ground plane.  I wrote and presented an article in part on
>> that very subject several years ago.
>> 
>> That said, here is what I learned on this forum.  The test set-up is very
>> clear that the EUT case and indeed its safety ground are NOT grounded, at
>> least at the frequencies associated with the test.  The power cord
>> including
>> safety ground go through a CDN which presents a 150 Ohm common mode
>> impedance in series with the power cord.  Any other EUT-connected cables
>> get
>> the same treatment.  The EUT chassis and connected cables are supported
>> above the ground plane at a  height that is supposed to provide a 150 Ohm
>> characteristic impedance.  The purpose of all of this is to have a
>> controlled impedance test set-up so that reflections are minimized.
>> 
>> So the answer relative to the standard is that the EUT chassis should
>> float
>> and attached cables should be common mode ungrounded.  Any grounds in
>> these
>> cables will see an rf impedance of 150 Ohms to ground from 150 kHz to 80
>> MHz.  My own opinion again: From a technical point-of-view, if the
>> equipment
>> in service will be bonded to a ground plane and the attached cables will
>> be
>> be held in close proximity to a ground plane then the levels injected by
>> EN61000-4-6 are higher relative to the assumed incident fields than they
>> would be in situ.  Equipment and cables installed close to and referenced
>> to
>> a ground plane have a lower antenna effective height than equipment
>> installed far from and not referenced to a ground plane.
>> 
>> Hope this helps.
>> 
>> Ken Javor
>> 
>> 
>> 
>>> From: "Oliver Betz" <[email protected]>
>>> Reply-To: "Oliver Betz" <[email protected]>
>>> Date: Wed, 12 Feb 2003 08:59:22 +0100
>>> To: [email protected]
>>> Subject: Grounding of EUT in EN61000-4-6 conducted immunity test
>>> 
>>> 
>>> Hello All,
>>> 
>>> how shall conductive cases of a EUT (intended to be mounted to a
>>> grounded metal construction in real life) be connected to the
>>> reference plane during a EN61000-4-6 test: directly, via a resistor
>>> or not at all?
>>> 
>>> I remember only that (auxiliary?) earth connectors have to be
>>> connected to ground via 150 Ohms, but I don't remember any word about
>>> the case of the EUT.
>>> 
>>> Thanks in advance,
>>> 
>>> Oliver
>>> 
>>> -------------------------------------------
>>> This message is from the IEEE EMC Society Product Safety
>>> Technical Committee emc-pstc discussion list.
>>> 
>>> Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/
>>> 
>>> To cancel your subscription, send mail to:
>>> [email protected]
>>> with the single line:
>>> unsubscribe emc-pstc
>>> 
>>> For help, send mail to the list administrators:
>>> Ron Pickard:              [email protected]
>>> Dave Heald:               [email protected]
>>> 
>>> For policy questions, send mail to:
>>> Richard Nute:           [email protected]
>>> Jim Bacher:             [email protected]
>>> 
>>> Archive is being moved, we will announce when it is back on-line.
>>> All emc-pstc postings are archived and searchable on the web at:
>>> http://www.ieeecommunities.org/emc-pstc
>>> 
>> 
>> 
>> -------------------------------------------
>> This message is from the IEEE EMC Society Product Safety
>> Technical Committee emc-pstc discussion list.
>> 
>> Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/
>> 
>> To cancel your subscription, send mail to:
>> [email protected]
>> with the single line:
>> unsubscribe emc-pstc
>> 
>> For help, send mail to the list administrators:
>> Ron Pickard:              [email protected]
>> Dave Heald:               [email protected]
>> 
>> For policy questions, send mail to:
>> Richard Nute:           [email protected]
>> Jim Bacher:             [email protected]
>> 
>> Archive is being moved, we will announce when it is back on-line.
>> All emc-pstc postings are archived and searchable on the web at:
>> http://www.ieeecommunities.org/emc-pstc
> 



This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/

To cancel your subscription, send mail to:
     [email protected]
with the single line:
     unsubscribe emc-pstc

For help, send mail to the list administrators:
     Ron Pickard:              [email protected]
     Dave Heald:               [email protected]

For policy questions, send mail to:
     Richard Nute:           [email protected]
     Jim Bacher:             [email protected]

Archive is being moved, we will announce when it is back on-line.
All emc-pstc postings are archived and searchable on the web at:
    http://www.ieeecommunities.org/emc-pstc

Reply via email to