Dear Bill and others,

Clause 4.1 of IEC 60601-1 states that the tests in the standard are type
tests, i.e. they are 
intended to test the design of a product.

Clause 4.1 of Appendix A (rationale) discusses production tests and suggests
leakage current, 
dielectric strength and earth continuity, where applicable.

Production testing is closely linked with the subject of product liability. If
you have a license 
>from a test house to put their label on your product, the test house is also
in the product 
liability firing range, and therefore usually demands production testing as
part of the terms of 
the license for the use of their label. The actual tests required depend on
the nature of the 
product, and may, to some extent, be the subject of negotiation on the basis
of engineering 
judgement.

If you do not use a test house label, and self declare, e.g. for CE marking,
then its up to you to 
what extent you do production testing. Nevertheless, it is your responsibility
to show that your 
production, and not just your design, is compliant; so both for this reason,
as well as product 
liability reasons, you would be strongly advised to perform thorough
production safety testing, as 
well as to make sure the results are recorded and traceable.

Regards,

Jon Griver
http://www.601help.com
The Medical Device Designer's Guide to IEC 60601-1



Hi,

I am relatively new in this field, but here are some questions and my
thoughts.

How do agencies require 100% testing for electrical medical products? Are
the requirements published somewhere, or are the tests specified by each
agency, per product? I know one standard, UL 2601-1, that specifies
production line tests, but it also states that the tests are "not an
integral part" of the standard.

At a particular 3rd party lab, the report that is issued after completion of
'type' testing includes the requirements for production line tests. This
usually is based off Clauses 700-702 of UL 2601-1. The tests specified in
these clauses include dielectric and ground continuity tests, but does not
include leakage current testing.

At another lab, the report doesn't specify any production tests and they
expect you to determine the relevant production tests on your own. In this
case, I do not know what happens if you decide not to do any production
tests as my company typically does conduct production safety testing.

Thanks,
Delphina
 


From: [email protected] [mailto:[email protected]] 
Sent: Monday, February 10, 2003 1:53 PM
To: 'FastWave'
Cc: [email protected]
Subject: RE: Electrical Medical Product Production Testing


I would find it hard to think of ANY agencies that WOULD NOT require 100%
testing.

I was at a product liability seminar at British Standards Institute many
years ago and someone asked about sampling levels for Earth Bond and Earth
Leakage on consumer goods. The BSI panel took some time to understand that
the questioner was serious.

A lot of us took mental (and written) note of the company name - to ensure
that we would not purchase anything from them.

Best regards

Gregg






From: [email protected]
[mailto:[email protected]]On Behalf Of FastWave
Sent: Monday, February 10, 2003 2:18 PM
To: '[email protected]'
Subject: Electrical Medical Product Production Testing


Does anybody know any Certification Agencies and/or Standards anywhere in
the world requiring Production Line Leakage Current Testing on Electrical
Medical Products? If so, does anybody have it in writing?

Thanks for the help,

Bill Bisenius
ED&D, Inc.
[email protected] <mailto:[email protected]>
www.productsafet.com <http://www.productsafet.com>




This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/

To cancel your subscription, send mail to:
     [email protected]
with the single line:
     unsubscribe emc-pstc

For help, send mail to the list administrators:
     Ron Pickard:              [email protected]
     Dave Heald:               [email protected]

For policy questions, send mail to:
     Richard Nute:           [email protected]
     Jim Bacher:             [email protected]

Archive is being moved, we will announce when it is back on-line.
All emc-pstc postings are archived and searchable on the web at:
    http://www.ieeecommunities.org/emc-pstc

Reply via email to