Hi Joe,

As I understand it, there is a difference between an authorised
representative and an importer. An authorised representative has to be in
the EU, and would normally work on behalf of the manufacturer, and in this
case may also sign the D of C. Effectively the AR is the manufacturer, but
you don't have to have an authorised representative in the EU.

Paragraph 3.3 of the New Approach guide
(http://europa.eu.int/comm/enterprise/n
wapproach/legislation/guide/document/chap03.pdf)

 talks about the responsibilities of the importer or person responsible for
placing the product on the market. The way it is worded, the importer in
the EU  must be able to provide the documents to the surveillance authority
on demand. One point though, if you go this route the importer has to be
listed as a point of contact in the User documentation. This approach is
fine if you only have one dealer or importer in the EU. If you have more
than one, then there you have to change the contact information for each
importer.
I personally think the authorised representative route, is better because
you now have a single point of contact and one technical file , regardless
of how many distributors you have.
I don't think the EU surveillance organisations like DTI care whether you
have the documents in hard copy or soft copy, but I think the point is that
they have to be available on demand, so effectively the technical file is
in the EU. I would just put them on a CD and send it to the
importer/distributor, or put them on a secure website, and they can print
them off when needed.

Regards

Doug


                                                                              
                                      
                    Joe Finlayson                                             
                                      
                    <jfinlayson@tel      To:    
"'[email protected]'" <[email protected]>,        
                    ica.com>               [email protected]                 
                                      
                                         cc:     [email protected]  
                                      
                    04/29/03 12:57       Subject:     RE: D of C again        
                                      
                    PM                                                        
                                      
                                                                              
                                      
                                                                              
                                      




Doug,

     I agree with the D of C, but where does it state that the technical
file must also be held by the representative in the EU?  Maybe I missed
that
one.  I am speaking with regards to the EMCD and LVD.  It's not a big deal
if it is required, although it would be easier to keep the files in a
central location and minimize copying.

Thx,


Joe

*************************************
Joe Finlayson
Manager, Compliance Engineering
Telica, Inc.
734 Forest Street, Bldg. G, Suite 100
Marlboro, MA 01752
Tel: (508) 804-8212
Fax: (508) 480-0922
Email: [email protected]


From: [email protected] [mailto:[email protected]]
Sent: Tuesday, April 29, 2003 11:19 AM
To: [email protected]
Cc: [email protected]
Subject: Re: D of C again




And the answer is........................

The D of C must be signed by the responsible person designated by the
manufacturer. The signatory does not have to be located in the EU.
The manufacturer must designate a representative in the EU who will hold
the certificate and the technical file. This could be an agent,
distributor.
A copy of the D of C that states the directives complied with must be
placed in the user manual. Note, this does not have to be signed, but it
must state where the signed original can be obtained. i.e. the EU
representative's contact details.


Doug Beckwith

(OOPO)




                    [email protected]

                    Sent by:                     To:
[email protected]
                    owner-emc-pstc@majordom      cc:

                    o.ieee.org                   Subject:     D of C again





                    04/29/03 09:13 AM

                    Please respond to

                    Lfresearch









Hi folks,

I've been asked by a US manufacturer who's signature if any should appear
on the D of C, and should they be located on European soil.

Is there a clear consensus?

Thanks,

Derek N. Walton
Owner L F Research EMC Design and Test Facility
Poplar Grove,
Illinois,  USA
www.lfresearch.com





This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/

To cancel your subscription, send mail to:
     [email protected]
with the single line:
     unsubscribe emc-pstc

For help, send mail to the list administrators:
     Ron Pickard:              [email protected]
     Dave Heald:               [email protected]

For policy questions, send mail to:
     Richard Nute:           [email protected]
     Jim Bacher:             [email protected]

Archive is being moved, we will announce when it is back on-line.
All emc-pstc postings are archived and searchable on the web at:
    http://www.ieeecommunities.org/emc-pstc





This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/

To cancel your subscription, send mail to:
     [email protected]
with the single line:
     unsubscribe emc-pstc

For help, send mail to the list administrators:
     Ron Pickard:              [email protected]
     Dave Heald:               [email protected]

For policy questions, send mail to:
     Richard Nute:           [email protected]
     Jim Bacher:             [email protected]

Archive is being moved, we will announce when it is back on-line.
All emc-pstc postings are archived and searchable on the web at:
    http://www.ieeecommunities.org/emc-pstc

Reply via email to