Greetings to all,

In EN61010-1 section 10 and EN60950 safety standards, there are surface
temperature limits specification for Normal Condition based on an Ambient
Temperature of 40ºC. Our company makes analytical instrumentation and a new
product we are working on is functionally very susceptible to room
temperatures.
So, the operational specification in our manual says it must be in a room no
warmer than 30ºC.

Question is,  is it proper to consider surface temperatures for safety based on
a 30ºC ambient which would  give us 10º more range to work with?  Someone
told
me you could do this but it doesn't seem right to me.  You are relying on the
building's air conditioning as a level of protection. Then I was told that if
the AC broke down it would be considered a Fault and then the 105ºC limit
would
be applied.  

Is this correct thinking?  Knowing the correct way to apply these requirements
effect the design of our instruments so it is important for me to know.  

Thanks to all in advance.
Brian Kunde
LECO Corp.




This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/

To cancel your subscription, send mail to:
     [email protected]
with the single line:
     unsubscribe emc-pstc

For help, send mail to the list administrators:
     Ron Pickard:              [email protected]
     Dave Heald:               [email protected]

For policy questions, send mail to:
     Richard Nute:           [email protected]
     Jim Bacher:             [email protected]

Archive is being moved, we will announce when it is back on-line.
All emc-pstc postings are archived and searchable on the web at:
    http://www.ieeecommunities.org/emc-pstc

Reply via email to