I may have missed something, but my understanding of clamp injection is that
it is a transformer in series with the cable-under-test (CUT).  That being
so, the CUT has loads at either end.  Even if the AE side is an open
circuit, the EUT side could be a dead short and draw a lot of current.
Further I will confess to not understanding the rationale for raising the
impedance on the AE side (if the CUT is unshielded).  If the AE end is a
high impedance relative to the EUT end, then the injected rf potential will
be dropped across the AE end, instead of the EUT end, which is where I would
think the injected potential should end up.  What obvious point am I
missing?

> From: <[email protected]>
> Reply-To: <[email protected]>
> Date: Fri, 23 Jan 2004 19:45:35 +0200
> To: <[email protected]>
> Subject: RE: Clamp Injection EN61000-4-3 Question
> 
> 
> Well,
> if you can maintain the 150 ohm to AE, your current cannot exceed Uo/150,
> right?
> 
> 
>> -----Original Message-----
>> From: [email protected]
>> [mailto:[email protected]]On Behalf Of ext
>> [email protected]
>> Sent: 23 January, 2004 19:08
>> To: [email protected]
>> Subject: RE: Clamp Injection EN61000-4-3 Question
>> 
>> 
>> 
>> My product is a relatively large, 19"-rack mounted device.
>> And I made a (kind of obvious) mistake, it is -6 (conducted
>> immunity) and not -3.
>> 
>> The chassis is supposed to be bonded to Safety Ground.
>> 6.2.2.1 provides requirement to connect the EUT chassis to
>> ground reference plane through CDN-M1 (to increase the CM impedance).
>> 
>> 150 Ohm (plus/minus tolerancefrom chassis to GND is a moot
>> point at the higher end of the tested frequency range, due to
>> the impedance of the chassis and the attached cables vs.
>> environment, that is in parallel to the CDN.
>> 
>> Let me rephrase my original question:
>> 
>> Am I allowed to measure current and stop the test if the
>> current into EUT exceeds Uo/150, even if I can meet the
>> formas requirements on the EUT and AE setup/impedances?
>> 
>> Thanks, Neven
>>> 
>>> You're right, not always (i.e. with safety class II).
>>> But if there's earth terminal a M-1 is connected to it, see 6.2.1.1
>>> of 2nd Ed.
>>> Ari
>>> 
>>>> -----Original Message-----
>>>> From: [email protected]
>>>> [mailto:[email protected]]On Behalf Of
>> ext Oliver Betz
>>>> Sent: 23 January, 2004 11:49
>>>> To: [email protected]
>>>> Subject: RE: Clamp Injection EN61000-4-3 Question
>>>> 
>>>> 
>>>> 
>>>> <[email protected]> wrote:
>>>> 
>>>> [EN61000-4-6]
>>>> 
>>>>> Some comments to Neven:
>>>>> -When you use BCI clamp you can measure and limit the
>>>> current as you describe.
>>>>> -The impedance is not >150 ohms but with +/- tolerance,
>> see Table 3.
>>>>> -The EUT needs to have 150 ohms from chassis to earth so
>>>> the impedance is not as low as you say
>>>> 
>>>> Really "_chassis_ to earth" - where do you read this?
>>>> 
>>>> Oliver
>>>> -- 
>>>> Oliver Betz, Muenchen
>>>> 
>>>> 
>>>> -------------------------------------------
>>>> This message is from the IEEE EMC Society Product Safety
>> 
>>>> Technical Committee emc-pstc discussion list.
>>>> 
>>>> Visit our web site at:  http://www.ieee-pses.org/
>>>> 
>>>> To cancel your subscription, send mail to:
>>>> [email protected]
>>>> with the single line:
>>>> unsubscribe emc-pstc
>>>> 
>>>> For help, send mail to the list administrators:
>>>> Ron Pickard:              [email protected]
>>>> Dave Heald:               [email protected]
>>>> 
>>>> For policy questions, send mail to:
>>>> Richard Nute:           [email protected]
>>>> Jim Bacher:             [email protected]
>>>> 
>>>> All emc-pstc postings are archived and searchable on the web at:
>>>> http://www.ieeecommunities.org/emc-pstc
>>>> 
>>> 
>>> 
>>> -------------------------------------------
>>> This message is from the IEEE EMC Society Product Safety
>>> Technical Committee emc-pstc discussion list.
>>> 
>>> Visit our web site at:  http://www.ieee-pses.org/
>>> 
>>> To cancel your subscription, send mail to:
>>> [email protected]
>>> with the single line:
>>> unsubscribe emc-pstc
>>> 
>> 
>>> For help, send mail to the list administrators:
>>> Ron Pickard:              [email protected]
>>> Dave Heald:               [email protected]
>>> 
>>> For policy questions, send mail to:
>>> Richard Nute:           [email protected]
>>> Jim Bacher:             [email protected]
>>> 
>>> All emc-pstc postings are archived and searchable on the web at:
>>> http://www.ieeecommunities.org/emc-pstc
>> 
>> 
>> -------------------------------------------
>> This message is from the IEEE EMC Society Product Safety
>> Technical Committee emc-pstc discussion list.
>> 
>> Visit our web site at:  http://www.ieee-pses.org/
>> 
>> To cancel your subscription, send mail to:
>> [email protected]
>> with the single line:
>> unsubscribe emc-pstc
>> 
>> For help, send mail to the list administrators:
>> Ron Pickard:              [email protected]
>> Dave Heald:               [email protected]
>> 
>> For policy questions, send mail to:
>> Richard Nute:           [email protected]
>> Jim Bacher:             [email protected]
>> 
>> All emc-pstc postings are archived and searchable on the web at:
>> http://www.ieeecommunities.org/emc-pstc
>> 
> 
> 
> -------------------------------------------
> This message is from the IEEE EMC Society Product Safety
> Technical Committee emc-pstc discussion list.
> 
> Visit our web site at:  http://www.ieee-pses.org/
> 
> To cancel your subscription, send mail to:
> [email protected]
> with the single line:
> unsubscribe emc-pstc
> 
> For help, send mail to the list administrators:
> Ron Pickard:              [email protected]
> Dave Heald:               [email protected]
> 
> For policy questions, send mail to:
> Richard Nute:           [email protected]
> Jim Bacher:             [email protected]
> 
> All emc-pstc postings are archived and searchable on the web at:
> http://www.ieeecommunities.org/emc-pstc
> 



This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ieee-pses.org/

To cancel your subscription, send mail to:
     [email protected]
with the single line:
     unsubscribe emc-pstc

For help, send mail to the list administrators:
     Ron Pickard:              [email protected]
     Dave Heald:               [email protected]

For policy questions, send mail to:
     Richard Nute:           [email protected]
     Jim Bacher:             [email protected]

All emc-pstc postings are archived and searchable on the web at:
    http://www.ieeecommunities.org/emc-pstc

Reply via email to