Gary, 

I think you also should consider comments in Annex AAA of IEC 60601-1-2:2001,
Subclause 36.2020.1.d): 

"Consideration should be given to ISO 14971 and IEC 60601-1-4 for this risk
analysis. 
Non-medical electrical equipment often has IMMUNITY requirements that are
different from 
those specified by this standard. The exclusion from IMMUNITY testing
according to this 
standard of non-medical electrical equipment that is not expected to affect
the ESSENTIAL 
PERFORMANCE or safety of the SYSTEM, if the non-medical electrical equipment
exhibits 
DEGRADATION, permits the use in a SYSTEM of non-medical electrical equipment
that may not 
comply with the requirements of this standard. An example of such equipment is
a printer 
used in a SYSTEM in which the information to be printed remains in the memory
of an 
EQUIPMENT until it is intentionally deleted, and the information can be
re-transmitted to the 
printer and re-printed in the case of interference with printer operation.
Instructions for use of 
the MEDICAL ELECTRICAL SYSTEM should caution the user to verify proper
operation of the 
printer before deleting stored data. In contrast, a non-medical electrical
equipment (e.g. ITE) 
used as a PATIENT monitoring central station would likely be subject to the
IMMUNITY 
requirements of this standard, depending upon the results of a risk analysis,
because loss or 
corruption of PATIENT information at the monitoring central station can
reasonably be expected 
to affect the safety of the PATIENT. 
Once equipment that could affect the ESSENTIAL PERFORMANCE or safety of the
SYSTEM has 
been identified, that equipment should meet the requirements of 36.202." 


Michael Gusel
Principal Engineer
Regulatory Affairs Department
Tel. 201-995-8362
Fax. 201-995-8605 



"Jon Griver" <[email protected]> 
Sent by: [email protected] 


03/12/2004 05:51 AM 


Please respond to
"Jon Griver" <[email protected]>


To
<[email protected]> 

cc

Subject
Fw: Medical Electrical Equipment

        





Gary, 
  
A couple of points here, not necessarily directly related to your question on
EMC. 
  
1. If the 'medical' software can have a direct or indirect effect on the
treatment the patient receives, then the software may be subject to the
requirements of the FDA in the USA and the Medical Devices Directive in the
EU. 
  
2. If the PC is physically connected to a medical device (according to the
definition below), then the PC is subject to the requirements of IEC
60601-1-1, the safety standard for electrical medical systems. The
consequences of this are dependent on whether the PC is used close to the
patient's bed. Even if it is not connected to a medical device, but is located
close to the patient, IEC 60601-1-1 should be taken into consideration, as the
patient could close a current loop by touching the PC. 
  
3. The paragraph you quote from the EMC standard can be paraphrased as saying
that if you do not test the PC to the immunity requirements of IEC 60601-1-2,
then you must perform a risk analysis to examine whether immunity problems
could detract from the patient's treatment. 
  
  
Regards, 
  
Jon Griver 
 <http://www.601help.com/> http://www.601help.com 
The Medical Device Designer's Guide to IEC 60601-1 
  
  
  
  
----- Original Message ----- 
From:  <mailto:[email protected]> Brent DeWitt 
To:  <mailto:[email protected]> Gary Stuyvenberg ; 
<mailto:[email protected]> [email protected] 
Sent: Friday, March 12, 2004 4:32 AM 
Subject: RE: Medical Electrical Equipment 

Hi Gary, 
  
If your example is just a PC with medical software, then the answer to you
question is found in the definition of Medical Electrical Equipment in
60601-1: 
  
----------------------------------- 
1.1                   medical electrical equipment  (hereinafter me equipment) 
Electrical equipment, provided with not more than one connection to a
particular supply mains; and intended by its manufacturer to be used in the
diagnosis, treatment, or monitoring of a patient; and that makes physical or
electrical contact with the patient or transfers energy to or from the patient
or detects such energy transfer to or from the patient. 
----------------------------------- 
Stand-alone equipment which has no direct influence on the patient is not MEE
it's ITE, no matter what the software does. 
Brent DeWitt 
member IEC SC-62A MT-23 
  
  


From: [email protected]
[mailto:[email protected]]On Behalf Of Gary Stuyvenberg
Sent: Thursday, March 11, 2004 4:22 PM
To: [email protected]
Subject: Medical Electrical Equipment

I am looking for help in understanding what the following paragraph from IEC
60601-1-2 
(Medical Electrical Equipment) means and the extent to which it
applies...........for example, 
a PC with medical software installed on it.  Thanks in advance.   
  
Gary 
  
  
 Page 35  
 60601-1-2 ) IEC:2001(E) 
  
*36.202.1 GENERAL 
*d) Non-medical electrical equipment 
Non-medical electrical equipment that is supplied as part of a SYSTEM and the
use of which 
in the SYSTEM can be reasonably expected not to affect the ESSENTIAL
PERFORMANCE or 
safety of the SYSTEM if the non-medical electrical equipment exhibits
DEGRADATION, is 
exempt from the IMMUNITY testing requirements of this standard, provided the
non-medical 
electrical equipment complies with applicable international IMMUNITY
standards.  (For example 
see CISPR 24 for ITE and IEC 61326-1 for measurement, control and laboratory
equipment.)   
The determination of reasonable expectation not to affect the ESSENTIAL
PERFORMANCE or 
safety of the SYSTEM shall be based upon a risk analysis. This risk analysis
is not required 
if the non-medical electrical equipment supplied as part of a SYSTEM is tested
for IMMUNITY 
in accordance with 36.202. 
  
EOM 
  