I am looking for help in understanding what the following paragraph from IEC
60601-1-2 
(Medical Electrical Equipment) means and the extent to which it
applies...........for example, 
a PC with medical software installed on it.  Thanks in advance.  
 
Gary
 
 
– Page 35 –
 60601-1-2 © IEC:2001(E) 
 
*36.202.1 GENERAL
*d) Non-medical electrical equipment 
Non-medical electrical equipment that is supplied as part of a SYSTEM and the
use of which 
in the SYSTEM can be reasonably expected not to affect the ESSENTIAL
PERFORMANCE or 
safety of the SYSTEM if the non-medical electrical equipment exhibits
DEGRADATION, is 
exempt from the IMMUNITY testing requirements of this standard, provided the
non-medical 
electrical equipment complies with applicable international IMMUNITY
standards.  (For example 
see CISPR 24 for ITE and IEC 61326-1 for measurement, control and laboratory
equipment.)  
The determination of reasonable expectation not to affect the ESSENTIAL
PERFORMANCE or 
safety of the SYSTEM shall be based upon a risk analysis. This risk analysis
is not required 
if the non-medical electrical equipment supplied as part of a SYSTEM is tested
for IMMUNITY 
in accordance with 36.202. 
 
EOM
 