I read in !emc-pstc that Pat Lawler <[email protected]> wrote (in
<[email protected]>)
about 'Medical devices and voltage dip testing' on Thu, 29 Apr 2004:
>EN60601-1-2:2001 (EMC requirements for general medical devices) requires
various 
>voltage dip tests at the extremes of the rated input voltage range.
>If a product is rated for 100-240Vac operation, this means test levels like 
>40Vac for 100ms (100Vac/50Hz nominal input, 60% dip for 5 cycles). The 
>performance criteria makes this especially difficult - "The system shall
provide 
>essential performance".
>
>This looks like a major design change for many of our power supplies.
>
>What has been the experience of list members designing power supplies or
systems 
>to meet this requirement?  Have you found it to be a minor design challenge,
or 
>are you still figuring out how to meet the test requirement?

It looks very much as though the 'world-wide' power supplies weren't
properly taken into account by TC 62 or SC62A. I suggest you contact
people concerned with those committees and ask them. There is a list of
officers available form the public part of the IEC web site:

http://www/iec.ch
-- 
Regards, John Woodgate, OOO - Own Opinions Only. 
The good news is that nothing is compulsory.
The bad news is that everything is prohibited.
http://www.jmwa.demon.co.uk Also see http://www.isce.org.uk 


This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ieee-pses.org/

To cancel your subscription, send mail to:
     [email protected]
with the single line:
     unsubscribe emc-pstc

For help, send mail to the list administrators:
     Ron Pickard:              [email protected]
     Dave Heald:               [email protected]

For policy questions, send mail to:
     Richard Nute:           [email protected]
     Jim Bacher:             [email protected]

All emc-pstc postings are archived and searchable on the web at:
    http://www.ieeecommunities.org/emc-pstc

Reply via email to