Hi David:

>   We both have the same analyzer (Vitrek 944i), EUT set-up, and are
conducting
>   dielectrics (reinforced and basic insulation) per IEC 60065. Only
difference
>   is that I have my analyzers ARC detect setting set to slow and test current
>   set to 30 mA.  The NRTL have their analyzer set to the defaults (fast ARC
>   detect / 10 mA).  When the test is conducted (after abnormals) our DUT
>   passes and theirs fail. Just to double check when I input the same setting
>   into the analyzer and re-conduct the test our EUT fails.

Today's hi-pot testers are just too sophisticated.
They tend to give indications of performance of the
various dielectrics and not necessarily indication
of the true electric strength of a safety-required
insulation.

The hi-pot settings are of no consequence whatsover
in your dealings with the NRTL.

The only thing that counts is the *CAUSE* of the
alleged breakdown.

In order to answer the question, you need to find
where the breakdown is occurring.  You need to find
the specific component or the spacing that is
breaking down (or causing the arc detect to trip).

When you know the cause, then you can determine
whether the breakdown indicates inadequate electric
strength of an insulator (which is the requirement).

Finding the cause of the "breakdown" is tricky.  You
need to set the hi-pot tester to the voltage at which
the tester trips due to arc detect.  Set the tester
for continuous operation.

Then, using a cardboard tube from a roll of paper
towels, point the tube to various parts of your
circuit, and listen for the "hiss" or "snap" of the
partial breakdown.

If this doesn't work, then remove, one at a time,
the components that bridge the isolation that you
are testing.  After each component removal, repeat
the hi-pot test.  When the arc detect no longer
trips, then the last component you removed was the
cause of the hi-pot trip.  You should be able to
measure that component as a stand-alone and repeat
the arc trip.

My suspicion is that your construction includes an
air gap and a solid insulation in series, where the
air gap is very small, say less than 1 mm.  This
construction is that of two capacitors in series,
where the voltage divides inversely with the value
of each capacitor.  The air gap is the smaller
capacitor, so most of the voltage is dropped across
that air gap.  When doing the hi-pot test, the solid
insulation limits the current, and the air gap
breaks down.  The current isn't enough to trip the
hi-pot tester, but may very well trip the arc
detect function.

Another way of finding the cause of the failure is
to increase the hi-pot voltage until you can clearly
see and hear the breakdown.  Unfortunately, this may
break something else, and fail to find the original
cause (due to the construction I mentioned above).

Both you and the NRTL are focusing on the test, not
on the component or spacing that has failed.  If
the NRTL prevails, then you MUST find the failure
and fix it.  But, depending on the cause of the
hi-pot failure, you may or may not need to fix the
cause of the failure.


Good luck, and best regards,
Rich


This message is from the IEEE Product Safety Engineering Society
emc-pstc discussion list.    Website:  http://www.ieee-pses.org/

To post a message to the list, send your e-mail to [email protected]

Instructions:  http://listserv.ieee.org/listserv/request/user-guide.html

List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:

     Ron Pickard:              [email protected]