Hi Pierre: 

I think the real question is whether the equipment should be compliant with the 
Medical Device Directive or the Low Voltage Directive/EMC Directive. 

This is the criteria specified in the MDD in Article 1 clause 2a: 
" 'medical device' means any instrument, apparatus, appliance, material or 
other article, whether used alone or in combination, including the software 
necessary for its proper application intended by the manufacturer to be used 
for human beings for  the purpose of: 
 - diagnosis, prevention, monitoring, treatment or alleviation of disease, 
 - diagnosis, monitoring, treatment, alleviation of or compensation for an 
injury or handicap, 
 - investigation, replacement or modification of the anatomy or of a 
physiological process, 
 - control of conception, 
and which does not achieve its principal intended action in or on the human 
body by pharmacological, immunological or metabolic means, but which may be 
assisted in its function by such means; " 

It sounds like if your equipment malfunctions, it could affect the treatment of 
disease (consider the effects of early termination of haemodialysis ). 

Pat Lawler
EMC Engineer 
Condor DC Power Supplies 



        "Pierre Selva" <[email protected]> 
Sent by: [email protected] 


03/22/2005 02:56 AM         
        To:        "Forum Safety-emc" <[email protected]> 
        cc:         
        Subject:        Medical or laboratory product ?






Hello all, 

  

One of my customers is making a product which is intended to make measurement 
on liquid issued from a haemodialysis system.  The product is close to the 
haemodialysis system, and it give only information on the liquid quality to 
indicate if it’s time to stop the dialyse procedure. It doesn’t have capability 
to stop the system,It gives only information to medical personnel. There is no 
electrical links between products, only a tubular link with liquid. Product is 
not an In Vitro Diagnostic system. 

My question is : 

What is the most appropriate standard to be use ?60601-1 or 61010-1 ? 

  

Many thanks for your inputs, 

Best regards, 

Pierre 

  

Pierre SELVA 

Responsable certifications produits 

EMITECH Rhône Alpes 

1 rue des Essarts 

38610 GIERES 

Ph : +33 (0)4 38 370 007 

Fx : +33 (0)4 38 371 004 