Wendy,

Intermediate checks means verification between calibration intervals.  You
need to be sure your equipment is working properly at regular intervals.  The
intent is to reduce the impact and magnitude of an equipment problem during a
verification or calibration.  If you do not check your equipment through
verification and a problem is found during your annual calibration, then the
magnitude of the problem is a whole years worth of projects that have to be
investigated.  Talk about tedious work!

 

If you had weekly or monthly checks of the system then you have a reference
point in time that you know the equipment was good.  So the magnitude of the
problem is from the time the problem was discovered back to the last time the
equipment was determined to be good through verification.  Which hopefully is
measured in days, or at worst a few weeks depending on your verification cycle.

 

Hope this covered your question…

 

 

Kind Regards, 

 

Sam Wismer

Technical Director

ACS, Inc.

 

*Tel: (770) 831-8048 
*Fax: (770) 831-8598 
*Web:  <BLOCKED::http://www.acstestlab.com/> www.acstestlab.com 
*Email:  [email protected] <BLOCKED::mailto:[email protected]> 

 

****************CONFIDENTIAL****************

This e-mail and any attachments may contain information which is confidential,
proprietary, privileged or otherwise protected by law. The information is
solely intended for the named addressee (or a person responsible for
delivering it to the addressee). If you are not the intended recipient of this
message, you are not authorized to read, print, retain, copy or disseminate
this message or any part of it. If you have received this e-mail in error,
please notify the sender immediately by return e-mail and delete it from your
computer.

  _____  

From: [email protected] [mailto:[email protected]] On Behalf Of wendy
Sent: Wednesday, August 03, 2005 11:44 PM
To: Emc-Pstc@Ieee. Org
Subject: ISO 17025 Section 5.5.10 Intermediate Checks

 

Dear All,

 

I am preparing my laboratory for accreditation. Referring to ISO 17025 Section
5.5.10 Intermediate Checks, I am uncertain on the extent of checks required. 

 

For example, IEC 61000-4-11 Section 6.1.2 Verification of the characteristics
of the voltage dips, short interruptions generators states that “the
generator characteristics has to be verified according to ……”. Does the
intermediate checks mention in 17025 coincides or refer to such tedious work?
Or does this verification actually means ‘calibration’?

 

What is the extent of intermediate checks required by ISO17025?

 

Appreciate advice from those who works in an accredited EMC lab. Thanks.

 

Regards,

Wendy Nya

EMC Manager

LabOne Singapore Pte Ltd

---------------------------------------------------------------- This message
is from the IEEE Product Safety Engineering Society emc-pstc discussion list.
Website: http://www.ieee-pses.org/ 

To post a message to the list, send your e-mail to [email protected] 


Instructions: http://listserv.ieee.org/listserv/request/user-guide.html 


List rules: http://www.ieee-pses.org/listrules.html 


For help, send mail to the list administrators: 


Scott Douglas [email protected] Mike Cantwell [email protected] 


For policy questions, send mail to: 


Richard Nute: [email protected] Jim Bacher: [email protected] 


All emc-pstc postings are archived and searchable on the web at: 


http://www.ieeecommunities.org/emc-pstc
---------------------------------------------------------------- This message
is from the IEEE Product Safety Engineering Society emc-pstc discussion list.
Website: http://www.ieee-pses.org/ 


To post a message to the list, send your e-mail to [email protected] 


Instructions: http://listserv.ieee.org/listserv/request/user-guide.html 


List rules: http://www.ieee-pses.org/listrules.html 


For help, send mail to the list administrators: 


Scott Douglas [email protected] Mike Cantwell [email protected] 


For policy questions, send mail to: 


Richard Nute: [email protected] Jim Bacher: [email protected] 


All emc-pstc postings are archived and searchable on the web at: 


http://www.ieeecommunities.org/emc-pstc


Reply via email to