In message <[email protected]>, 
dated Fri, 17 Mar 2006, "Barker, Neil" <[email protected]> writes

>It is not clear whether your external power supply is Class 1 (basic 
>insulation + protective earth) or Class 2 (double insulated). If the 
>former, then the connection between the power supply and your 
>instrument should be providing the safety grounding. If the latter, 
>then providing the instrument does not generate any hazardous voltages 
>there is no need for a protective earth as all parts including the case 
>will be SELV. The complication arises when you connect to the PC, as it 
>is almost certainly a Class 1 apparatus, but since the PC interface 
>would be considered safe, the 'inadvertent' connection of the 
>instrument to Ground via the data cable is of no concequence as it is 
>either already connected (in the Class 1 case) or is double insulated 
>(in the Class 2 case).

There is one other consideration. What does this instrument measure? Is 
it connected to something that could supply it with a large fault 
current?
-- 
OOO - Own Opinions Only. Try www.jmwa.demon.co.uk and www.isce.org.uk
2006 is YMMVI- Your mileage may vary immensely.

John Woodgate, J M Woodgate and Associates, Rayleigh, Essex UK

-

This message is from the IEEE Product Safety Engineering Society
emc-pstc discussion list.    Website:  http://www.ieee-pses.org/

To post a message to the list, send your e-mail to [email protected]

Instructions:  http://listserv.ieee.org/request/user-guide.html

List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:

     Scott Douglas           [email protected]
     Mike Cantwell           [email protected]

For policy questions, send mail to:

     Jim Bacher:             [email protected]
     David Heald:            [email protected]

All emc-pstc postings are archived and searchable on the web at:

    http://www.ieeecommunities.org/emc-pstc

Reply via email to