Dave,
 
It is not clear whether your external power supply is Class 1 (basic
insulation + protective earth) or Class 2 (double insulated). If the former,
then the connection between the power supply and your instrument should be
providing the safety grounding. If the latter, then providing the instrument
does not generate any hazardous voltages there is no need for a protective
earth as all parts including the case will be SELV. The complication arises
when you connect to the PC, as it is almost certainly a Class 1 apparatus, but
since the PC interface would be considered safe, the 'inadvertent' connection
of the instrument to Ground via the data cable is of no concequence as it is
either already connected (in the Class 1 case) or is double insulated (in the
Class 2 case).
Best regards 

Neil R. Barker CEng MIEE FSEE MIEEE 
Manager 
Compliance Engineering 
e2v technologies (uk) ltd 
106 Waterhouse Lane 
Chelmsford 
Essex CM1 2QU 
UK 

Tel: (+44) 1245 453616 
Fax: (+44) 1245 453410 
Mob: (+44) 7801 723735 


From: [email protected] [mailto:[email protected]]
Sent: 17 March 2006 16:05
To: [email protected]
Subject: lab equipment connected to PC



Gents,

Situation: I am building an instrument that interfaces to a DAQ card in a PC.
The instrument is powered by an external AC-DC power supply supplying +/- 12
volts at 100 mA. The instrument metal case is connected to the PC Common via
the DAQ card cable. 

Question: Is this safety grounding arrangement legal under EN61010? Do we need
to provide more solid earthing of the instrument case? 

   Dave Cuthbert

   Micron Technology, Inc.

- ---------------------------------------------------------------- This
message is from the IEEE Product Safety Engineering Society emc-pstc
discussion list. Website: http://www.ieee-pses.org/ 

To post a message to the list, send your e-mail to [email protected] 


Instructions: http://listserv.ieee.org/request/user-guide.html 


List rules: http://www.ieee-pses.org/listrules.html 


For help, send mail to the list administrators: 


Scott Douglas [email protected] Mike Cantwell [email protected] 


For policy questions, send mail to: 


Jim Bacher: [email protected] David Heald: [email protected] 


All emc-pstc postings are archived and searchable on the web at: 


http://www.ieeecommunities.org/emc-pstc

- ---------------------------------------------------------------- This
message is from the IEEE Product Safety Engineering Society emc-pstc
discussion list. Website: http://www.ieee-pses.org/ 

To post a message to the list, send your e-mail to [email protected] 


Instructions: http://listserv.ieee.org/request/user-guide.html 


List rules: http://www.ieee-pses.org/listrules.html 


For help, send mail to the list administrators: 


Scott Douglas [email protected] Mike Cantwell [email protected] 


For policy questions, send mail to: 


Jim Bacher: [email protected] David Heald: [email protected] 


All emc-pstc postings are archived and searchable on the web at: 


http://www.ieeecommunities.org/emc-pstc 


Reply via email to