Hi,
 
Your question 1: There is no specific phase noise requirement in the standards
you mention. In 61000-4-3- the test system as whole needs to comply with the
requirement for harmonics per Clause 6. In 61000-4-6 it's in Clause 6.1.
 
Your question 2: To do the tests without power meter and directional coupler
is questionable. In that case you need to trust that your PA gain does not
change from the calibration. And without those you really cannot check that
your PA is not saturated, which you need to do.
61000-4-3 strongly recommends the use of directional coupler and power meter
(some people read the text as it being mandatory), and while you have it it
can be used for 4-6, too.
 
Best Regards
 
Ari Honkala


  _____  

From: [email protected] [mailto:[email protected]] On Behalf Of ext Grace Lin
Sent: 27. huhtikuuta 2006 16:55
To: [email protected]
Subject: Phase Noise per 61000-4-3 and 61000-4-6 and More


Dear Experts,
 
We are in the process of building a new engineering building with a 3m chamber
(for emission and immunicty measurement).  We are looking for A2LA
accreditation after the chamber is built.  A lot of test equipment need to be
purchased.  And a lot of questions arise. 
 
What is phase noise requirement for a signal generator to perform 61000-4-3
and 61000-4-6 tests?  Since I used a Gigatronics 6082A signal generator
before, I use its specifications to look for one.  An agent returned to me
with a $32k quote and told me if I can lower phase noise requirement, he can
find a much cheaper model. 
 
An used equipment is not an option as the owner of the company will not
approve any used equipment.
 
Personally, I prefer to have a signal generator in a chamber control room to
perform 61000-4-3 test and a signal generator outside of chamber area to
perform 61000-4-6 test.  I am not sure if I can achieve this as I do have a
budget limit. 
 
Another question:  do I need to purchase a directional coupler and a power
meter for these two tests?  For 61000-4-3 test, since I have a field probe and
a field monitor, I think I am OK.  For 61000-4-6, this is questionable. 
 
Your help is highly appreciated!
 
Best regards,
Grace Lin
 
 
I need your help to find inex
- ---------------------------------------------------------------- This
message is from the IEEE Product Safety Engineering Society emc-pstc
discussion list. Website: http://www.ieee-pses.org/ 

To post a message to the list, send your e-mail to [email protected] 


Instructions: http://listserv.ieee.org/request/user-guide.html 


List rules: http://www.ieee-pses.org/listrules.html 


For help, send mail to the list administrators: 


Scott Douglas [email protected] Mike Cantwell [email protected] 


For policy questions, send mail to: 


Jim Bacher: [email protected] David Heald: [email protected] 


All emc-pstc postings are archived and searchable on the web at: 


http://www.ieeecommunities.org/emc-pstc 

- ---------------------------------------------------------------- This
message is from the IEEE Product Safety Engineering Society emc-pstc
discussion list. Website: http://www.ieee-pses.org/ 

To post a message to the list, send your e-mail to [email protected] 


Instructions: http://listserv.ieee.org/request/user-guide.html 


List rules: http://www.ieee-pses.org/listrules.html 


For help, send mail to the list administrators: 


Scott Douglas [email protected] Mike Cantwell [email protected] 


For policy questions, send mail to: 


Jim Bacher: [email protected] David Heald: [email protected] 


All emc-pstc postings are archived and searchable on the web at: 


http://www.ieeecommunities.org/emc-pstc 


Reply via email to