It seems clear to me, from this post and what I know, that the 10 meter site
is the gold-standard.  If I were a commercial test facility, or an in-house
test facility for a company making large products, a ten meter site would be
the way to go.

But if I were a PC manufacturer, or if I qualified items that size or smaller,
I can see no justification whatsoever for the huge cost of the ten meter
chamber vs. a three meter chamber.  As long as I am already using a biconical
antenna instead of a tuned dipole at the low end, I see no cost-effective
rationale for the larger chamber.

Even if my small test item might have a long cable attached to it, I ought to
be able to control the cable common-mode current emissions using a current
probe or absorbing clamp and be able to generate a limit for those currents
commensurate with the radiated emission limit at three or ten meters.

This approach is not new, it dates back to at least the 1980s, with a series
of papers given at the Zurich symposia by Mr. Balint Szentkuti (e.g.,
Szentkuti, B., Give Up Radiation Testing In Favour Of Conduction Testing,
Proceedings, EMC Zurich 1989).  

I'm not advocating elimination of radiated testing, but simply recognizing
that emissions emanating from long cables are more accurately measured and
controlled as currents than as fields, not to mention much less expensively. 
For precisely the same reason that 61000-4-6 is performed in lieu of 61000-4-3
below 80 MHz.



From: "Kunde, Brian" <[email protected]>
List-Post: [email protected]
Date: Mon, 18 Jun 2007 15:32:32 -0400
To: "Grace Lin" <[email protected]>, <[email protected]>
Subject: RE: RE Measurement Distance per EU




Grace,

Do they make a fully compliant 3 meter semi-anechoic chamber?  I wasn't aware
that a 3 meter chamber passed the NSA requirements down to 30Mhz. 

The purpose of having a Test Standard is to standardize on a method in which
testing is performed so the results can be repeated at other test sites and
hopefully have the same or similar results. When you stray from what the
standard calls out, then you take the risk of your results being brought into
question. 

I think your biggest problem with testing at 3 meter is going to be with near
field below 100Mhz. Because of the common differences seen when testing a
product at 3 meter verses 10 meter, I think there will always be a high level
of doubt in your test results. 

In my opinion, "compliance testing" can only be confidently done at a 10 meter
OATS or equivalent.  

BTW, the note in section 10.2.1 is a method to test signals that fall near or
on a high ambient. Most all OATS have at least a few areas with high ambient
and this section gives them a work around at only those frequencies. It is not
intended to allow full testing at distances than 10 meter. In a chamber you
wouldn't have any ambient signals so you couldn't use this section to justify
3 meter testing. 

I am very curious to know the confidence level of 3 meter chambers are. Maybe
some who have such chambers will post. 

The Other Brian.



  _____  


From: [email protected] [mailto:[email protected]] On Behalf Of Grace Lin
Sent: Monday, June 18, 2007 1:43 PM
To: [email protected]
Subject: RE Measurement Distance per EU

Dear Members,

Is 3m measurement distance for radiated emission not accepted by EU?

I propose a 3m full compliant chamber.  A chamber manufacturer informs me that
EU only accepts 10m measurement distance.  I checked EN 55022: 1998
incorprating Corrigendum Nol 1 and A1 and A2.  Clause 6 states:

"The EUT shall meet the limits of Table 5 or Table 6 when measured at the
measuring distance R in accordance with the methods described in Clause 10.
...  "

The title of Table 5 is: "Table 5 - Limits for radiated disturbance of class A
ITE at a measuring distance of 10m".  The title of Table 6 is: "Table 6 -
Limits for radiated disturbanc eof class B ITE at a measuring distance of
10m".  Subclause 10.2.1 states:

"
10.2.1 Antenna-to-EUT distance
Measurements of the radiated field shall be made with the antenna located at
the horizontal distance from the boundary of the EUT as specified in Clasue 6.
  

NOTE If the field-strength measurement at 10m cannot be made because of high
ambient noise levels, or fo rother reasons, measurement of class B EUT's may
be made at a closer distance, for example 3m.  An inverse proportionality
factor of 20 dB per decade should be used to normalize the measured data to
the specified distance for determining compliance.  Care should be taken in
the measurement of large EUTs at 3m at frequencies near 30 MHz, due to near
field effects. 
"

I look forward to your comments.

Best regards,
Grace Lin
Crestron Electronics, Inc.
www.crestron.com <http://www.crestron.com> 

- ---------------------------------------------------------------- This
message is from the IEEE Product Safety Engineering Society emc-pstc
discussion list. Website: http://www.ieee-pses.org/ 

To post a message to the list, send your e-mail to [email protected] 

Instructions: http://listserv.ieee.org/request/user-guide.html 

List rules: http://www.ieee-pses.org/listrules.html 

For help, send mail to the list administrators: 

Scott Douglas [email protected] Mike Cantwell [email protected] 

For policy questions, send mail to: 

Jim Bacher: [email protected] David Heald: [email protected] 

All emc-pstc postings are archived and searchable on the web at: 

http://www.ieeecommunities.org/emc-pstc Notice:  This communication may
contain confidential information intended for the named recipient(s)only. If
you received this by mistake, please destroy it and notify us of the error. 
Thank you. - ----------------------------------------------------------------
This message is from the IEEE Product Safety Engineering Society emc-pstc
discussion list.    Website:  http://www.ieee-pses.org/ 

To post a message to the list, send your e-mail to [email protected] 

Instructions:  http://listserv.ieee.org/request/user-guide.html 

List rules: http://www.ieee-pses.org/listrules.html 

For help, send mail to the list administrators: 

Scott Douglas           [email protected] Mike Cantwell          
[email protected] 

For policy questions, send mail to: 

Jim Bacher:             [email protected] David Heald:           
[email protected] 

All emc-pstc postings are archived and searchable on the web at: 

http://www.ieeecommunities.org/emc-pstc 




______________________________________________________________________
This email has been scanned by the MessageLabs Email Security System.
For more information please visit http://www.messagelabs.com/email 
______________________________________________________________________

- ---------------------------------------------------------------- This
message is from the IEEE Product Safety Engineering Society emc-pstc
discussion list. Website: http://www.ieee-pses.org/ 

To post a message to the list, send your e-mail to [email protected] 


Instructions: http://listserv.ieee.org/request/user-guide.html 


List rules: http://www.ieee-pses.org/listrules.html 


For help, send mail to the list administrators: 


Scott Douglas [email protected] Mike Cantwell [email protected] 


For policy questions, send mail to: 


Jim Bacher: [email protected] David Heald: [email protected] 


All emc-pstc postings are archived and searchable on the web at: 


http://www.ieeecommunities.org/emc-pstc 


Reply via email to