Your explanation is rather lengthy and a bit hard to follow so I will 
only generalize a bit.

First you might want to look up the wave guide effect in coax. Coax has 
upper limits because it changes modes from coax and starts behaving like 
waveguide. I don't recall how sudden the effect is.

Second you talk about coax but you don't mention connectors. They have a 
cut off too. It's usually lower than coax! I would be working with SMAs 
if I was going to 10 GHz.

RG-214 is 7.3 db/100' at a GHz and 31.6 db/100' at 10 GHz.  That's a 25 
db slope! I'd say a slippery slope at that.

Third when you eliminated the antennas, did you simply join the coax 
using a 'barrel' connector. Barrels as all connectors have significant  
VSWR at 10 GHz.  I think I would replace the two coax with one equal to 
the two combined lengths.

Fred Townsend
DC to Light

Ken Javor wrote:

>Forum Members,
>
>This is a simple question with a long intro.  The question mainly concerns
>the use of coax at frequencies higher than what it was designed for.  What
>effects are to be expected?
>
>I just completed checking an antenna factor calibration per SAE ARP-958. The
>antenna separation was one meter.  The antenna was a microwave log-conical
>or log-spiral used from 1 - 10 GHz. Measurements were made at every GHz.
>
>Per ARP-958, two identical antennas are set up one meter apart with one
>driven and the other connected to a receiver.  A reading is taken, then the
>cables are disconnected from the antennas and connected together. The
>antenna factor can be determined from the difference between the straight
>through reading and the reading through the antennas.
>
>This measurement is pure substitution, it does not rely on traceability to
>any standard.  It does assume the two antennas are identical.
>
>I used two similar but slightly different set-ups. Below 5 GHz, I used
>RG-214 cables, and lots of it. The antennas were about five feet above the
>ground (grass) and the separation of the antennas from anything above ground
>was over two meters. The signal source and receivers were directly behind
>each antenna  and close to the ground at at least two meters separation.
>The antennas have about 2 dBi gain looking forward.
>
>With this set-up, I got excellent agreement with manufacturer's generic
>antenna factors published in their catalog.  All measurements were within 1
>dB, and only one measurement was close to 1 dB off.
>
>But there was too much loss in the RG-214 to go above 6 GHz, and at 6 GHz
>the error was 1.1 dB (not bad, but the trend was increasing error, which I
>attributed to lack of measurement dynamic range).
>
>So I changed out the RG-214 for lower loss cable, and shorter cables. This
>necessitated lowering the antennas to about four feet above the ground, and
>bringing the signal source and receiver within about one meter of their
>respective antennas, still directly behind the antennas.  It also
>necessitated bring the antennas closer to the concrete driveway, which
>paralleled the axis between the antennas, but still over one meter away.
>
>Some of the low loss cable was good to 5-6 GHz, but some was only good to
>3-4 GHz.  This cable was very large diameter and stiff with 7/16" DIN
>connectors.  I used the smaller diameter lighter cable to connect to the
>antennas, and the heavier stuff hung off that, with the connection between
>the two types of coax near the ground to minimize moment-arm torque on the
>antennas.
>
>The vswr rating is excellent on these antennas below 5 GHz, so I didn't pad
>the antennas at all when using the RG-214, on account of the RG-214 was all
>the pad I could stand.  But the vswr increases some at higher frequencies,
>so with the low loss coax I placed a 20 dB pad at the transmit antenna only.
>The pad was good to 18 GHz.
>
>With this set-up, I had plenty of signal to work with and easily made
>measurements to 10 GHz.  The measurement at 6 GHz yielded an antenna factor
>0.5 dB closer to manufacturer's specification, so I thought I was on my way.
>
>But the errors grew with increasing frequency. The worst was 2.5 dB off at
>10 GHz.  Now that isn't at all bad - I would have been happy with that if
>the other errors were random.  But they aren't - the errors increase with
>increasing frequency and the antenna factor is always larger than it is
>supposed to be.
>
>The point of all this information is to show the controls that were in place
>and see if anyone reading this can find any source of uncontrolled error
>that would have been systematic, not random.  The only thing I could think
>of that was not perfect was the antennas were closer to the ground for the
>high frequency portion of the testing then I would have liked and the cables
>were not spec'd to 10 GHz.  But I would think both of those would have
>contributed errors that did not increase monotonically with increasing
>frequency.
>
>Any suggestions?
>
>Thanks in advance,
>
>Ken Javor
>
>-
>----------------------------------------------------------------
>This message is from the IEEE Product Safety Engineering Society
>emc-pstc discussion list.    Website:  http://www.ieee-pses.org/
>
>To post a message to the list, send your e-mail to [email protected]
>
>Instructions:  http://listserv.ieee.org/request/user-guide.html
>
>List rules: http://www.ieee-pses.org/listrules.html
>
>For help, send mail to the list administrators:
>
>     Scott Douglas           [email protected]
>     Mike Cantwell           [email protected]
>
>For policy questions, send mail to:
>
>     Jim Bacher:             [email protected]
>     David Heald:            [email protected]
>
>All emc-pstc postings are archived and searchable on the web at:
>
>    http://www.ieeecommunities.org/emc-pstc
>
>
>
>  
>

-

This message is from the IEEE Product Safety Engineering Society
emc-pstc discussion list.    Website:  http://www.ieee-pses.org/

To post a message to the list, send your e-mail to [email protected]

Instructions:  http://listserv.ieee.org/request/user-guide.html

List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:

     Scott Douglas           [email protected]
     Mike Cantwell           [email protected]

For policy questions, send mail to:

     Jim Bacher:             [email protected]
     David Heald:            [email protected]

All emc-pstc postings are archived and searchable on the web at:

    http://www.ieeecommunities.org/emc-pstc

______________________________________________________________________
This email has been scanned by the MessageLabs Email Security System.
For more information please visit http://www.messagelabs.com/email 
______________________________________________________________________

Reply via email to