Hi Dennis,

we have common ground, and I'm in agreement, except for table size. If we have 
to show it's not influencing the measurement, it's not a factor.  If you want 
to argue about height.... I'll pass on that. It's 80 cm.

Cheers,

Derek.

On 10/11/2011 2:14 PM, Dennis Ward wrote: 

        Heaven forbid we ever go back to the early 80’s when there was no 
acceptable methods, no accepted procedures, no accepted setup and when the FCC 
expected cable manipulation till the cows came home to find the maximum 
emissions levels.  Talk about a waste of time – days upon days moving every 
cable into every likely position finding the ‘worse’ case.  Then to find out 
that the FCC or another lab found that one cable position that showed 10dB more 
than the other lab measured.   Everyone was right in their own eyes and most of 
those eyes were blind except to their own erroneous methods and practices.

        No, conformity of setup, table size, configuration, support equipment 
etc, may not be the best, but it is infinitely better than 30, even 25 years 
ago.

         

        Dennis Ward 
        
        

        Director of Engineering

        American Certification Body 
        Certification Resource for the Wireless Industry http://www.acbcert.com
        703-847-4700 fax 703-847-6888 
        direct - 703-880-4841

         

        From: [email protected] [mailto:[email protected]] On Behalf Of Ken 
Javor
        Sent: Tuesday, October 11, 2011 11:29 AM
        To: 'EMC-PSTC'
        Subject: Re: Table Size in Emissions test

         

        Within his frame of reference, Ward is correct, which is why this 
entire accreditation process is flawed and broken.  We got along without it 
before, and it is adding nothing but extra costs and bureaucracy, and negative 
value to the process of controlling EMI.
         
        Ken Javor
        
        Phone: (256) 650-5261
        
        

        
________________________________


        From: Dennis Ward <[email protected]> <mailto:[email protected]> 
        Reply-To: <[email protected]> <mailto:[email protected]> 
        Date: Tue, 11 Oct 2011 11:15:00 -0700
        To: 'Derek Walton' <[email protected]> <mailto:[email protected]> 
        Cc: 'Grace Lin' <[email protected]> <mailto:[email protected]> , 
'WNya' <[email protected]> <mailto:[email protected]> , 'EMC-PSTC' 
<[email protected]> <mailto:[email protected]> 
        Subject: RE: Table Size in Emissions test
        
        HI Derek
        The point of anything ‘nominal’ is that it starts with what is expected 
and allows reasonable variations of sorts. And you hit the right words 
“relating to a designated or theoretical size”.  If you say you must follow the 
standard, then follow it. The designated size of a table is 1m x 1.5m x 0.8m.  
This then is the size generally expected.  It is not a micrometer reading with 
0.01m measurement uncertainty tolerances; it is a general designated and 
expected size – a nominal size.  
         
        Now, and without going into uncertainties and its minutia, a 0.8m x 
1.2m table, while not typically or generally expected, could be stretched to be 
within a range of what could be considered within the nominal range.  Likewise 
a table 1.2m x 1.8m would or should still be considered within an expected or 
nominal value. To the contrary, a table 0.2m x 0.2m, for many reasons, would 
not be generally considered expected and thus not nominal, but abnormal in size 
for the intent and purposes of the standard.  
         
        Remembering that there are three dimensions given (i.e. 1mx1.5mx0.8m), 
if we take your ‘exception’, then I could place a device on the ground plane as 
long as it was on a nonconducting surface.  Yet I know of no lab nor auditor 
that would accept this as a ‘nominal’ height condition for any of the standards 
mentioned.  I would not be able to reject or challenge their claim that my 
table height was not 80cm by saying, nominal is “of, being, or relating to a 
designated or theoretical size that may vary from the actual, and I don’t have 
to have a table 80cm high.”  
        
        One might say, yes, but height matters. And they would be correct, 
however, the dimensions are together in the same frame of reference in the same 
sentence. So any exceptions to the dimensions would have to follow the same 
process.  The issue then is not uncertainty and its issues, but what is 
generally expected and accepted as a reasonable table for the intent of the 
standard.  
         
        Remembering that standards are produced as much to make something 
consistent wherever used, then the nominal would be that which is generally 
consistent within the population that uses the standard. 
         
        As to ‘bigger fish to fry’, this would be a good size fish for a lab 
that was seeking accreditation. Yes there are allowable variations, equipment 
considerations, etc that can be discussed with the accrediting organization.  
To say that because ‘nominal’ is used and because ‘shall’ is not, I can do what 
I want, is simply not the case if they want to become accredited. It is fairly 
easy and cheap to construct a nominal table of dimensions in the standards.  
There is probably more money, time and effort in trying to challenge or reject 
the assessment than to simply hire a carpenter and build one.
         
        But the choice is the labs and how they wish to relate to their 
accreditation body.
         
        
        Dennis Ward 
        
        
        Director of Engineering
        American Certification Body 
        Certification Resource for the Wireless Industry http://www.acbcert.com
        703-847-4700 fax 703-847-6888 
        direct - 703-880-4841
        
        
        From: Derek Walton [mailto:[email protected]] 
        Sent: Tuesday, October 11, 2011 10:09 AM
        To: Dennis Ward
        Cc: 'Grace Lin'; 'WNya'; 'EMC-PSTC'
        Subject: Re: Table Size in Emissions test
        
        Sorry Dennis, 
        
        you are not correct. Nominal means:
        
        b : of, being, or relating to a designated or theoretical size that may 
vary from the actual.
        
        This is from Websters.
        
        As you indicate, there is NOTHING that says it has to be 1 by 1.5m
        
        I'm sorry to be so anal about this, but it is happening too much where 
assessors are assessing to opinions and personal agendas. We must follow the 
standard. Usually there are MUCH bigger fish to fry than quibbling over 
something like this
        
        I repeat, again, that there is NOTHING that says the table SHALL be 1 
by 1.5m. Only then could a deficiency be written.
        
        Sincerely,
        
        Derek.
        
        
        On 10/11/2011 11:43 AM, Dennis Ward wrote: 
        Both ANSI C63.10 and ANSI C63.4, the typical standards for which 
ISO17025 accreditation is used, contain the following statement, “Tabletop 
devices shall be placed on a nonconducting platform, of nominal size 1 m by 1.5 
m, raised 80 cm above the reference ground plane.
        
        While a bit more open to variation due to size of equipment, CISPR 22 
has the statement “Equipment intended for tabletop use shall be placed on a 
non-conductive table. The size of the table will nominally be 1,5 m ◊ 1,0 m but 
may ultimately be dependent on the horizontal dimensions of EUT.
         
        So while you may challenge the accreditation organization, they are 
correct, your table does not meet the ‘nominal’ size requirements for at least 
the two standards ANSI C63.4 and C6310.
         
        Other standards may also have the nominal size issue as well.
         
        Thanks 
        Dennis Ward 
        
        
        
        Director of Engineering
        American Certification Body 
        Certification Resource for the Wireless Industry http://www.acbcert.com
        703-847-4700 fax 703-847-6888 
        direct - 703-880-4841
         
        
        From: [email protected] [mailto:[email protected]] On Behalf Of Grace 
Lin
        Sent: Tuesday, October 11, 2011 4:16 AM
        To: WNya
        Cc: EMC-PSTC
        Subject: Re: Table Size in Emissions test
        
        
        Wendy,
        
         
        
        When I look for an accredited ANSI C63.4 laboratory, I expect the 
laboratory has the facility as stated in the standard, including a standard 
size of the table as defined in the standard.  For this reason, unless the 
accreditation certificate bears a restriction note, I support the auditor's 
comment.  
        
         
        
        From the other point of view, many manufacturers' laboratories are for 
internal use only, including my employer's.  For this reason, the auditor may 
be willing to accept the smaller size of the table.  The question is how to 
determine if the laboratory is for internal use only (for testing certain type 
of products).  An example is Alcatel-Lucent's EMC laboratory in New Jersey, 
USA.  It opens to the general public.
        
         
        
        With regards,
        
        Grace
        
         
        
         
        
        On Mon, Oct 10, 2011 at 7:54 PM, WNya <[email protected]> 
<mailto:[email protected]>  wrote:
        Dear Experts,
        Recently my company went through the first ISO17025 audit. We have a 
table smaller than the standard requirement of 1.5m x 1m since our products are 
small, typically 10cm x 10cm x 10cm. The height of our table was 0.8m. The 
auditor wanted us to change the table size to follow the standard.
        What does it matter since we never use the extra space on the table? I 
do agree we must keep to the height requirement since the floor is a ground 
plane and thus it sets a fixed capacitance to the EUT and also controls the 
lengths of any attached cables.
        
        Can we reject or challenge the auditor's request? Has anyone experience 
the same situation?
        
        Sent from Wendy.Nya iPhone
        
        -
        ----------------------------------------------------------------
        This message is from the IEEE Product Safety Engineering Society 
emc-pstc discussion list. To post a message to the list, send your e-mail to 
<[email protected]> <mailto:[email protected]> 
        
        All emc-pstc postings are archived and searchable on the web at:
        http://product-compliance.oc.ieee.org/
        Graphics (in well-used formats), large files, etc. can be posted to 
that URL.
        
        Website:  http://www.ieee-pses.org/
        Instructions:  http://listserv.ieee.org/request/user-guide.html
        List rules: http://www.ieee-pses.org/listrules.html
        
        For help, send mail to the list administrators:
        Scott Douglas <[email protected]> 
<mailto:[email protected]> 
        Mike Cantwell <[email protected]> <mailto:[email protected]> 
        
        For policy questions, send mail to:
        Jim Bacher:  <[email protected]> <mailto:[email protected]> 
        David Heald: <[email protected]> <mailto:[email protected]> 
        
        -
        ----------------------------------------------------------------
        This message is from the IEEE Product Safety Engineering Society 
emc-pstc discussion list. To post a message to the list, send your e-mail to 
<[email protected]> <mailto:[email protected]> 
        
        All emc-pstc postings are archived and searchable on the web at 
http://product-compliance.oc.ieee.org/
        Graphics (in well-used formats), large files, etc. can be posted to 
that URL. 
        Website: http://www.ieee-pses.org/
        Instructions: http://listserv.ieee.org/request/user-guide.html
        List rules: http://www.ieee-pses.org/listrules.html 
        
        For help, send mail to the list administrators:
        Scott Douglas <[email protected]> 
<mailto:[email protected]> 
        Mike Cantwell <[email protected]> <mailto:[email protected]>  
        
        For policy questions, send mail to:
        Jim Bacher <[email protected]> <mailto:[email protected]> 
        David Heald <[email protected]> <mailto:[email protected]>  
        
        -
        ----------------------------------------------------------------
        This message is from the IEEE Product Safety Engineering Society 
emc-pstc discussion list. To post a message to the list, send your e-mail to 
<[email protected]> <mailto:[email protected]> 
        
        All emc-pstc postings are archived and searchable on the web at 
http://product-compliance.oc.ieee.org/
        Graphics (in well-used formats), large files, etc. can be posted to 
that URL. 
        Website: http://www.ieee-pses.org/
        Instructions: http://listserv.ieee.org/request/user-guide.html
        List rules: http://www.ieee-pses.org/listrules.html 
        
        For help, send mail to the list administrators:
        Scott Douglas <[email protected]> 
<mailto:[email protected]> 
        Mike Cantwell <[email protected]> <mailto:[email protected]>  
        
        For policy questions, send mail to:
        Jim Bacher <[email protected]> <mailto:[email protected]> 
        David Heald <[email protected]> <mailto:[email protected]>  
        -
        ----------------------------------------------------------------
        This message is from the IEEE Product Safety Engineering Society 
emc-pstc discussion list. To post a message to the list, send your e-mail to 
<[email protected]> <mailto:[email protected]> 
        
        All emc-pstc postings are archived and searchable on the web at 
http://product-compliance.oc.ieee.org/
        Graphics (in well-used formats), large files, etc. can be posted to 
that URL. 
        
        Website:      http://www.ieee-pses.org/
        Instructions:  http://listserv.ieee.org/request/user-guide.html
        List rules:     http://www.ieee-pses.org/listrules.html 
        
        For help, send mail to the list administrators:
        Scott Douglas <[email protected]> 
<mailto:[email protected]> 
        Mike Cantwell <[email protected]> <mailto:[email protected]>  
        
        For policy questions, send mail to:
        Jim Bacher  <[email protected]> <mailto:[email protected]> 
        David Heald <[email protected]> <mailto:[email protected]>  

        -
        ----------------------------------------------------------------
        This message is from the IEEE Product Safety Engineering Society 
emc-pstc discussion list. To post a message to the list, send your e-mail to 
<[email protected]>
        
        All emc-pstc postings are archived and searchable on the web at 
http://product-compliance.oc.ieee.org/
        Graphics (in well-used formats), large files, etc. can be posted to 
that URL. 

        Website: http://www.ieee-pses.org/
        Instructions: http://listserv.ieee.org/request/user-guide.html
        List rules: http://www.ieee-pses.org/listrules.html 

        For help, send mail to the list administrators:
        Scott Douglas <[email protected]>
        Mike Cantwell <[email protected]> 

        For policy questions, send mail to:
        Jim Bacher <[email protected]>
        David Heald <[email protected]> 

        -
        ----------------------------------------------------------------
        This message is from the IEEE Product Safety Engineering Society 
emc-pstc discussion list. To post a message to the list, send your e-mail to 
<[email protected]>
        
        All emc-pstc postings are archived and searchable on the web at 
http://product-compliance.oc.ieee.org/
        Graphics (in well-used formats), large files, etc. can be posted to 
that URL. 

        Website: http://www.ieee-pses.org/
        Instructions: http://listserv.ieee.org/request/user-guide.html
        List rules: http://www.ieee-pses.org/listrules.html 

        For help, send mail to the list administrators:
        Scott Douglas <[email protected]>
        Mike Cantwell <[email protected]> 

        For policy questions, send mail to:
        Jim Bacher <[email protected]>
        David Heald <[email protected]> 

-
----------------------------------------------------------------
This message is from the IEEE Product Safety Engineering Society emc-pstc 
discussion list. To post a message to the list, send your e-mail to 
<[email protected]>

All emc-pstc postings are archived and searchable on the web at 
http://product-compliance.oc.ieee.org/
Graphics (in well-used formats), large files, etc. can be posted to that URL. 

Website: http://www.ieee-pses.org/
Instructions: http://listserv.ieee.org/request/user-guide.html
List rules: http://www.ieee-pses.org/listrules.html 

For help, send mail to the list administrators:
Scott Douglas <[email protected]>
Mike Cantwell <[email protected]> 

For policy questions, send mail to:
Jim Bacher <[email protected]>
David Heald <[email protected]> 


Reply via email to