Hi John: 

No, the article doesn't mention field strength.  However, the test procedure
is one I'm not familiar with: 
   ANSI C63.18-1997, 'American national standard recommended practice for an
on-site, ad hoc test method for 
   estimating radiated electromagnetic immunity of medical devices to specific
radio-frequency transmitters' 
In short, they vary the separation between the RFID device and the medical
equipment from 600 cm (20 feet) down to 0 cm (touching the equipment case),
and record the distance where the equipment failed.  Random RFID devices,
random power levels, with a test distance down to zero? 

I appreciate RFID interference is something to be considered.  However, it's
not fair to compare the immunity standard used in the report, to the pass/fail
testing with the fixed amplitude/fixed distance conditions in present IEC
immunity tests (mentioned elsewhere in the email thread.) 

Pat Lawler
EMC Engineer
SL Power Electronics Corp.

John Woodgate <[email protected]> wrote on 06/27/2008 09:28:20 AM:
> In message <[email protected]>,
> dated Fri, 27 Jun 2008, "Dean Gerard (Medical Physics)"
> <[email protected]> writes:
> >
> >Whilst I've seen the original article (Journal of the American Medical
> >Association), unfortunately I can't circulate it without breaking our
> >subscription license agreement.

> Thanks for the information you have given, anyway. The elephant in the
> room is the field strength information. Is there anything about that in
> the paper?
> --
> OOO - Own Opinions Only. Try www.jmwa.demon.co.uk and www.isce.org.uk
> Either we are causing global warming, in which case we may be able to stop
it,
> or natural variation is causing it, and we probably can't stop it. You
choose!
> John Woodgate, J M Woodgate and Associates, Rayleigh, Essex UK -
---------------------------------------------------------------- This message
is from the IEEE Product Safety Engineering Society emc-pstc discussion list.
Website: http://www.ieee-pses.org/ 

To post a message to the list, send your e-mail to [email protected] 

Instructions: http://listserv.ieee.org/request/user-guide.html 

List rules: http://www.ieee-pses.org/listrules.html 

For help, send mail to the list administrators: 

Scott Douglas [email protected] Mike Cantwell [email protected] 

For policy questions, send mail to: 

Jim Bacher: [email protected] David Heald: [email protected] 

All emc-pstc postings are archived and searchable on the web at: 

http://www.ieeecommunities.org/emc-pstc 

Reply via email to