I'm working on a device (laboratory equipment) which will have a standard 10 
amp IEC 60320 Inlet connector. We also want to supply AC Mains power to an 
external device through an IEC 60320 OUTlet connector. We want to control 
(switch) the OUTlet connector  with a relay so we can turn the external device 
on and off.

Questions:


1.     Does the relay have to be double pole or will a single pole relay be ok 
(the device can operate at either 115VAC or 230VAC in different power systems 
around the globe).

2.    Does the relay have to be a safety relay or can we just control one pole 
using a solid state relay which could fail open or closed? Either condition is 
not likely to cause a hazard within the external device.

3.    Does power to the OUTlet connector have to be removed with the ON/OFF 
switch (disconnect device) within the main instrument? Can this outlet always 
be HOT or at least one leg be HOT assuming a single pole relay is open? Are 
there international symbols that explain "switched" or "unswitched" outlet 
connectors?

4.    If the OUTlet connector and wiring is rated to handle the same current 
from the Mains Inlet, do we have to have a circuit breaker or fuse on the 
Outlet? And if so, does it have to be a double pole breaker or just a single 
pole ok?

As you can guess, my questions are related to there being Mains Voltages on the 
Outlet connector relative to PE or Neutral when the control relay is open, 
circuit breaker or fuse is open, or even when the instrument is turned off. Are 
any of these conditions prohibited? The external device we are powering 
(external stand-a-lone pump) has warning labels about disconnecting power cord 
prior to servicing.

Thanks for any input.

The Other Brian