If you think of "non-permanently connected equipment" as cord-and-plug 
connected equipment, it is not uncommon for cord-and-plug connected equipment 
to rely upon the facility mains overcurrent protection (i.e., a circuit breaker 
in a distribution panel). 



Mike Sherman 

Product Safety and Compliance Engineer 

Graco Inc. 


----- Original Message -----


From: "Niels Hougaard" <[email protected]> 
To: [email protected] 
Sent: Friday, November 1, 2013 4:45:23 AM 
Subject: [PSES] Concerning EN 61010-1, §9.6 Overcurrent protection 




Dear list members, 

  

EN 61010-1:2010 §9.6 is concerning Overcurrent protection. All equipment 
supplied from mains must be protected by some sort of overcurrent protection 
device. 

  

§9.6.3 for non-permanently connected equipment says “If an overcurrent 
protection device is provided, it shall be within the equipment”. 

  

Does this mean that equipment under §9.6.3 must have a built-in overcurrent 
protection device? Or is it enough to specify in the manual that an overcurrent 
protection device of certain size in the building installation is required? 

  

Any answer much appreciated 

  

Regards, and have a nice weekend 

Niels 

Niels Hougaard 

Bolls ApS 

Ved Gadekæret 11F 

DK-3660 Stenløse 

Denmark 

  

T: +45 48 18 35 66 

F: +45 48 18 35 30 

[email protected] 

www.bolls.dk 

  

  - 
---------------------------------------------------------------- 


This message is from the IEEE Product Safety Engineering Society emc-pstc 
discussion list. To post a message to the list, send your e-mail to &LT; 
[email protected] &GT; 

All emc-pstc postings are archived and searchable on the web at: 
http://www.ieee-pses.org/emc-pstc.html 

Attachments are not permitted but the IEEE PSES Online Communities site at 
http://product-compliance.oc.ieee.org/ can be used for graphics (in well-used 
formats), large files, etc. 

Website: http://www.ieee-pses.org/ 
Instructions: http://www.ieee-pses.org/list.html (including how to unsubscribe) 
List rules: http://www.ieee-pses.org/listrules.html 

For help, send mail to the list administrators: 
Scott Douglas &LT; [email protected] &GT; 
Mike Cantwell &LT; [email protected] &GT; 

For policy questions, send mail to: 
Jim Bacher &LT; [email protected] &GT; 
David Heald &LT; [email protected] &GT;

-
----------------------------------------------------------------
This message is from the IEEE Product Safety Engineering Society emc-pstc 
discussion list. To post a message to the list, send your e-mail to 
<[email protected]>

All emc-pstc postings are archived and searchable on the web at:
http://www.ieee-pses.org/emc-pstc.html

Attachments are not permitted but the IEEE PSES Online Communities site at 
http://product-compliance.oc.ieee.org/ can be used for graphics (in well-used 
formats), large files, etc.

Website:  http://www.ieee-pses.org/
Instructions:  http://www.ieee-pses.org/list.html (including how to unsubscribe)
List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:
Scott Douglas <[email protected]>
Mike Cantwell <[email protected]>

For policy questions, send mail to:
Jim Bacher:  <[email protected]>
David Heald: <[email protected]>

Reply via email to