On 7/11/2016 5:49 PM, Schaefer, David wrote:
These were all calibrated probes, and three were identical models. The
one that was way off wasn’t the odd model. Our cal house measures a DC
voltage for accuracy, and bandwidth.
What’s causing the inaccuracy? How can I prevent this problem in the
future? Moving to all active probes isn’t an option right now.
In addition to what's been written in other replies, there is a
difference in calibration if you even hold the oscilloscope probe
differently when measuring faster, high-frequency signals. It's always
better to use the short steel wire (or coaxial tip adapter) ground
connection provided with a high-frequency probe when making high-speed,
high frequency measurements – and changing the ground lead requires
*readjusting the probe's user calibration*. That's what the CAL
terminal on a 'scope front panel is for.
Metrology departments might think probe calibration is sacred and
forever, but it's not.
Cortland Richmond
-
----------------------------------------------------------------
This message is from the IEEE Product Safety Engineering Society emc-pstc discussion
list. To post a message to the list, send your e-mail to <[email protected]>
All emc-pstc postings are archived and searchable on the web at:
http://www.ieee-pses.org/emc-pstc.html
Attachments are not permitted but the IEEE PSES Online Communities site at
http://product-compliance.oc.ieee.org/ can be used for graphics (in well-used
formats), large files, etc.
Website: http://www.ieee-pses.org/
Instructions: http://www.ieee-pses.org/list.html (including how to unsubscribe)
List rules: http://www.ieee-pses.org/listrules.html
For help, send mail to the list administrators:
Scott Douglas <[email protected]>
Mike Cantwell <[email protected]>
For policy questions, send mail to:
Jim Bacher: <[email protected]>
David Heald: <[email protected]>