Hi Vincent:

 

You asked:

 

1) Why is Class III electrical protection not accepted in IEC 60601-1, standard 
for medical device safety?

 

According to IEC 60601-1, Annex 4:  

 

The term “Class III equipment” is used in some other standards to identify 
equipment that is

powered from a safety extra-low voltage (SELV) mains supply system. The term 
Class III

equipment is not formally used in this standard. The BASIC SAFETY of Class III 
equipment is

critically dependent on the installation and on other Class III equipment 
connected thereto.

These factors are outside the control of the OPERATOR and this is considered to 
be

unacceptable for ME EQUIPMENT. Additionally, limitation of voltage is not 
sufficient to ensure

safety of the PATIENT. For these reasons, this standard does not recognize 
Class III

construction.

 

My further understanding is that the only criterion for IEC 60601-1 electrical 
safety is limitation of current irrespective of voltage.  In my opinion, this 
is appropriate for ME equipment.

 

Best regards,

Rich

 

 

 

 


-
----------------------------------------------------------------
This message is from the IEEE Product Safety Engineering Society emc-pstc 
discussion list. To post a message to the list, send your e-mail to 
<emc-p...@ieee.org>

All emc-pstc postings are archived and searchable on the web at:
http://www.ieee-pses.org/emc-pstc.html

Attachments are not permitted but the IEEE PSES Online Communities site at 
http://product-compliance.oc.ieee.org/ can be used for graphics (in well-used 
formats), large files, etc.

Website:  http://www.ieee-pses.org/
Instructions:  http://www.ieee-pses.org/list.html (including how to unsubscribe)
List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:
Scott Douglas <sdoug...@ieee.org>
Mike Cantwell <mcantw...@ieee.org>

For policy questions, send mail to:
Jim Bacher:  <j.bac...@ieee.org>
David Heald: <dhe...@gmail.com>

Reply via email to