Hi Bernd

It’s a mess but at this time no one seems to want to address the limit. If the 
public exposure limit is 400mT then all equipment including pacemakers should 
operate at this level or higher.

Some progress was made towards the slightly older ICNIRP limit of 40mT

BS ISO 14117:2012 

4.7 Protection from static magnetic fields of flux density up to 50 mT
4.7.1 General considerations
The DUT shall not remain functionally affected after exposure to static 
magnetic fields of flux density of up to 50 mT.
4.7.2 Pacemakers and CRT-P devices
Test equipment: use a field coil that is capable of generating a uniform 
magnetic field of flux density of up to
50 mT ± 5 mT in the region to be occupied by the DUT.
Test procedure: the required field flux density shall be generated before 
placing the DUT in the field. Then the
DUT shall be slowly placed in the centre of the test coil. After at least 15 s 
of exposure to the magnetic field,
the DUT shall be slowly removed from the field.

http://www.rbht.nhs.uk/about/news-events/archive/pacemaker/

The easiest was around it is to actually contact the manufacturer and try and 
get some form of reassurance from them. 

Regards

Andy


-----Original Message-----
From: Dürrer Bernd [mailto:[email protected]] 
Sent: 06 July 2018 09:53
To: McCallum, Andy <[email protected]>; [email protected]
Subject: AW: [PSES] EN62311 - Human Exposure to Electromagnetic Fields

Hello Andy,

Thank you for pointing out the ICNIRP limits: Actually, I recently had a 
request concerning a service technician with a pacemaker and exposure to static 
magnetic fields when handling permanent magnet rotors.
However, when I check ICNIRP resources, I do not find any publication to 
justify higher limits. On the contrary, the “ICNIRP GUIDELINES ON LIMITS OF 
EXPOSURE TO STATIC MAGNETIC FIELDS” 
(https://emea01.safelinks.protection.outlook.com/?url=https%3A%2F%2Fwww.icnirp.org%2Fcms%2Fupload%2Fpublications%2FICNIRPstatgdl.pdf&data=01%7C01%7CAndy.McCallum%40mottmac.com%7C15d40e00326b4219a4df08d5e31ddb08%7Ca2bed0c459574f73b0c2a811407590fb%7C0&sdata=GtqO0ps18xjo1yE6siOKfeeYPdNzy4Ws7Tv6OcFLzmM%3D&reserved=0)
 contain the statement:

“Electromagnetic interference from low-intensity static magnetic fields has 
been observed to affect the operation of pacemakers, particularly those with 
magnetic switches, and other types of medical electronic devices, including 
cardiac defibrillators, hormone infusion pumps (e.g., for insulin), 
neuromuscular stimulation devices (e.g., for the sphincter muscle of the 
bladder), neurostimulators, and electronically operated prosthetic devices 
(e.g., for the limbs and inner ear). In general, the operation of these devices 
is not adversely affected by static magnetic fields below 0.5 mT."

The "ICNIRP GUIDELINES FOR LIMITING EXPOSURE TO TIME-VARYING ELECTRIC AND 
MAGNETIC FIELDS (1 Hz TO 100 kHz)" 
(https://emea01.safelinks.protection.outlook.com/?url=https%3A%2F%2Fwww.icnirp.org%2Fcms%2Fupload%2Fpublications%2FICNIRPLFgdl.pdf&data=01%7C01%7CAndy.McCallum%40mottmac.com%7C15d40e00326b4219a4df08d5e31ddb08%7Ca2bed0c459574f73b0c2a811407590fb%7C0&sdata=TKolfMPIxVlXj8Gu%2FcuYTd2jpKm2TQpJJY8R%2B4ihA3k%3D&reserved=0)
 contain the following statement:

"Compliance with the present guidelines may not necessarily preclude 
interference with, or effects on, medical devices such as metallic prostheses, 
cardiac pacemakers and implanted defibrillators and cochlear implants. 
Interference with pacemakers may occur at levels below the recommended 
reference levels. Advice on avoiding these problems is beyond the scope of the 
present document but is available elsewhere (IEC 2005b)." (IEC2005b refers to 
IEC 60601-1-2)

Kind regards,

Bernd



Von: McCallum, Andy <[email protected]>
Gesendet: Freitag, 6. Juli 2018 09:41
An: Dürrer Bernd <[email protected]>; [email protected]
Betreff: RE: [PSES] EN62311 - Human Exposure to Electromagnetic Fields

Bernd

Pacemakers have DC static magnetic field limits as low as 0.5mT as mentioned in 
the EU EMF directive. Which is a level encountered in many environments not 
least DC Trams and railways. The fact that we don’t see people with pacemakers 
staggering around the underground suggests this limit has been set far too low 
(most likely when pacemakers where first introduced). The ICNIRP public 
exposure limit is    400mT which is much more realistic and only encountered 
rarely.

Contacted a number of UK government agencies regarding this and the most honest 
verbal response was we will get rid of that limit after Brexit.

Also note a number of pacemakers are now available which are safe to wear in 
MRI scanners operating at 7Teslas.

Andy

From: Dürrer Bernd [mailto:[email protected]]
Sent: 06 July 2018 08:24
To: mailto:[email protected]
Subject: [PSES] AW: [PSES] EN62311 - Human Exposure to Electromagnetic Fields

Hi Doug,

I assume that due to your good health you don’t have a pacemaker or any other 
electronic medical implants that might be affected by EM field exposure. As 
protection and shielding options for such implants are limited (except by the 
surrounding tissue), the remaining other option to reduce the risk of 
malfunction is to limit EM fields.

Stay in the best of health and have a nice weekend,

Bernd

Von: Doug Smith <mailto:[email protected]>
Gesendet: Donnerstag, 5. Juli 2018 20:56
An: mailto:[email protected]
Betreff: Re: [PSES] EN62311 - Human Exposure to Electromagnetic Fields

Hi Michael and the group,

With the exception tissue heating generated by some sources at very high 
frequencies, I wonder if the EM field exposure worries today are a bit 
overblown. I started my experiments in high voltage  (100,000+ Volts, and NOT 
static electricity) around age 13 and by age 14 was regularly immersing myself 
in 10,000+V/m, 300 kHz fields. It was fun, heating up screw driver blades from 
induction heating, lighting incandescent light bulbs (as well as fluorescent 
tubes) without wires, and more. About 600 Watts of RF energy was concentrated 
in the relatively small space I was in and I was uninjured! My kids all have 
two arms, two legs, and one head and I am still here almost 60 years later and 
healthier than 90% of adults of any age from 18 on. Some of my friends say this 
explains a lot about me (frying brain cells) though.

I realize this is only one case, but have had a lot of high energy RF exposure 
throughout my life. Such as the field in my dorm room from my amateur radio 
transmitter being so strong that the fluorescent tubes in the room lit up with 
no power and blinked with Morse code. I had to remove the tubes from their 
fixtures and cover with towels so my roommate could sleep.

Still here,
Doug



On Thu, 5 Jul 2018 16:27:12 +0100, Michael Derby wrote:
It’s not quite an “unintentional radiator” but if you had a wireless charger 
which did not contain any communication (e.g., a dumb charging signal without 
handshaking),  then it would come within the scope of the EMCD and the LVD (not 
the RED), and RF Exposure would therefore be an issue under the LVD.

Michael.


From: McCallum, Andy [mailto:[email protected]]
Sent: 05 July 2018 15:33
To: mailto:[email protected]
Subject: Re: [PSES] EN62311 - Human Exposure to Electromagnetic Fields

Be surprised if any LVD could cause EMF large enough to pose a danger to Human 
Health. Intentional transmitters maybe, a rare beast to create those levels 
unintentionally at less than 1000V.

Andy

From: John Woodgate [mailto:[email protected]]
Sent: 05 July 2018 10:28
To: mailto:[email protected]
Subject: Re: [PSES] EN62311 - Human Exposure to Electromagnetic Fields

It applies to any LVD product, but there is a flow chart that shows that a 
simple assessment is possible if the product has no reasonable likelihood of 
producing sufficiently strong EMF.
John Woodgate OOO-Own Opinions Only
J M Woodgate and Associates 
https://emea01.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.woodjohn.uk&data=01%7C01%7CAndy.McCallum%40mottmac.com%7C3897dc1f76c0406b3a9808d5e259910b%7Ca2bed0c459574f73b0c2a811407590fb%7C0&sdata=vOXyjzMm8FtxrbHRrTRbkgWlhb8gQeuv%2FiISyiAwg24%3D&reserved=0
Rayleigh, Essex UK
On 2018-07-05 10:09, Amund Westin wrote:
EN62311 is listed in OJ and a harmonized LVD standard.
Does EN62311apply to any LVD product or only products containing radio 
transmitters?


Best regards
Amund
-
----------------------------------------------------------------
This message is from the IEEE Product Safety Engineering Society emc-pstc 
discussion list. To post a message to the list, send your e-mail to 
<mailto:[email protected]> All emc-pstc postings are archived and searchable on 
the web at: 
https://emea01.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.ieee-pses.org%2Femc-pstc.html&data=01%7C01%7CAndy.McCallum%40mottmac.com%7C3897dc1f76c0406b3a9808d5e259910b%7Ca2bed0c459574f73b0c2a811407590fb%7C0&sdata=49%2BsokQ%2Bfgbdw5PdFOVMQq9w5i6xUg%2FynDYiz5MmRso%3D&reserved=0
Attachments are not permitted but the IEEE PSES Online Communities site at 
https://emea01.safelinks.protection.outlook.com/?url=http%3A%2F%2Fproduct-compliance.oc.ieee.org%2F&data=01%7C01%7CAndy.McCallum%40mottmac.com%7C3897dc1f76c0406b3a9808d5e259910b%7Ca2bed0c459574f73b0c2a811407590fb%7C0&sdata=NXnfbl2tReFV%2FOyyCDItpDVpyeZdAnDYXJo2y7%2B2b74%3D&reserved=0
 can be used for graphics (in well-used formats), large files, etc.
Website: 
https://emea01.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.ieee-pses.org%2F&data=01%7C01%7CAndy.McCallum%40mottmac.com%7C3897dc1f76c0406b3a9808d5e259910b%7Ca2bed0c459574f73b0c2a811407590fb%7C0&sdata=iF4opy2qXpxGUDSOPjI0JO542mTk9qx9tgUZ8G8KhIE%3D&reserved=0
Instructions: 
https://emea01.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.ieee-pses.org%2Flist.html&data=01%7C01%7CAndy.McCallum%40mottmac.com%7C3897dc1f76c0406b3a9808d5e259910b%7Ca2bed0c459574f73b0c2a811407590fb%7C0&sdata=clkGQ06wwx8J0uLKatJk2TlwZP%2FLmDxRUJvlWeyFEcA%3D&reserved=0
List rules: 
https://emea01.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.ieee-pses.org%2Flistrules.html&data=01%7C01%7CAndy.McCallum%40mottmac.com%7C3897dc1f76c0406b3a9808d5e259910b%7Ca2bed0c459574f73b0c2a811407590fb%7C0&sdata=FfgMUjF4Y7aAu%2F08gj5CE%2BRbAH8ja9RnUj7jRe%2FK7Zg%3D&reserved=0
For help, send mail to the list administrators:
Scott Douglas <mailto:[email protected]> Mike Cantwell 
<mailto:[email protected]> For policy questions, send mail to:
Jim Bacher <mailto:[email protected]>
David Heald <mailto:[email protected]>

-
----------------------------------------------------------------
This message is from the IEEE Product Safety Engineering Society emc-pstc 
discussion list. To post a message to the list, send your e-mail to 
<mailto:[email protected]> All emc-pstc postings are archived and searchable on 
the web at: 
https://emea01.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.ieee-pses.org%2Femc-pstc.html&data=01%7C01%7CAndy.McCallum%40mottmac.com%7C3897dc1f76c0406b3a9808d5e259910b%7Ca2bed0c459574f73b0c2a811407590fb%7C0&sdata=49%2BsokQ%2Bfgbdw5PdFOVMQq9w5i6xUg%2FynDYiz5MmRso%3D&reserved=0
Attachments are not permitted but the IEEE PSES Online Communities site at 
https://emea01.safelinks.protection.outlook.com/?url=http%3A%2F%2Fproduct-compliance.oc.ieee.org%2F&data=01%7C01%7CAndy.McCallum%40mottmac.com%7C3897dc1f76c0406b3a9808d5e259910b%7Ca2bed0c459574f73b0c2a811407590fb%7C0&sdata=NXnfbl2tReFV%2FOyyCDItpDVpyeZdAnDYXJo2y7%2B2b74%3D&reserved=0
 can be used for graphics (in well-used formats), large files, etc.
Website: 
https://emea01.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.ieee-pses.org%2F&data=01%7C01%7CAndy.McCallum%40mottmac.com%7C3897dc1f76c0406b3a9808d5e259910b%7Ca2bed0c459574f73b0c2a811407590fb%7C0&sdata=iF4opy2qXpxGUDSOPjI0JO542mTk9qx9tgUZ8G8KhIE%3D&reserved=0
Instructions: 
https://emea01.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.ieee-pses.org%2Flist.html&data=01%7C01%7CAndy.McCallum%40mottmac.com%7C3897dc1f76c0406b3a9808d5e259910b%7Ca2bed0c459574f73b0c2a811407590fb%7C0&sdata=clkGQ06wwx8J0uLKatJk2TlwZP%2FLmDxRUJvlWeyFEcA%3D&reserved=0
List rules: 
https://emea01.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.ieee-pses.org%2Flistrules.html&data=01%7C01%7CAndy.McCallum%40mottmac.com%7C3897dc1f76c0406b3a9808d5e259910b%7Ca2bed0c459574f73b0c2a811407590fb%7C0&sdata=FfgMUjF4Y7aAu%2F08gj5CE%2BRbAH8ja9RnUj7jRe%2FK7Zg%3D&reserved=0
For help, send mail to the list administrators:
Scott Douglas <mailto:[email protected]> Mike Cantwell 
<mailto:[email protected]> For policy questions, send mail to:
Jim Bacher <mailto:[email protected]>
David Heald <mailto:[email protected]>
-
----------------------------------------------------------------
This message is from the IEEE Product Safety Engineering Society emc-pstc 
discussion list. To post a message to the list, send your e-mail to 
<mailto:[email protected]> All emc-pstc postings are archived and searchable on 
the web at: 
https://emea01.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.ieee-pses.org%2Femc-pstc.html&data=01%7C01%7CAndy.McCallum%40mottmac.com%7C22523de7d06d4570c2c408d5e31197b8%7Ca2bed0c459574f73b0c2a811407590fb%7C0&sdata=2XdRLyQbOgOH62TTRXCBhVVm18K3dKVIIxXtahKZxhk%3D&reserved=0
Attachments are not permitted but the IEEE PSES Online Communities site at 
https://emea01.safelinks.protection.outlook.com/?url=http%3A%2F%2Fproduct-compliance.oc.ieee.org%2F&data=01%7C01%7CAndy.McCallum%40mottmac.com%7C22523de7d06d4570c2c408d5e31197b8%7Ca2bed0c459574f73b0c2a811407590fb%7C0&sdata=N625izyKbER4FJGjsv2xPjQ0InnEeRU6omzltg3ANhQ%3D&reserved=0
 can be used for graphics (in well-used formats), large files, etc.
Website: 
https://emea01.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.ieee-pses.org%2F&data=01%7C01%7CAndy.McCallum%40mottmac.com%7C22523de7d06d4570c2c408d5e31197b8%7Ca2bed0c459574f73b0c2a811407590fb%7C0&sdata=uRNIxlw1y%2BiLDmupgC52VcVzgrsyK3oR8LSqCCf95pk%3D&reserved=0
Instructions: 
https://emea01.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.ieee-pses.org%2Flist.html&data=01%7C01%7CAndy.McCallum%40mottmac.com%7C22523de7d06d4570c2c408d5e31197b8%7Ca2bed0c459574f73b0c2a811407590fb%7C0&sdata=kzljLVqzCWogvZM68PXZsf%2F2mhHPotXvkdogaXboNSI%3D&reserved=0
List rules: 
https://emea01.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.ieee-pses.org%2Flistrules.html&data=01%7C01%7CAndy.McCallum%40mottmac.com%7C22523de7d06d4570c2c408d5e31197b8%7Ca2bed0c459574f73b0c2a811407590fb%7C0&sdata=Xk6Vunv806vdykueUDEPdhh1ZR0k%2BPizXrWrtNE0Jw0%3D&reserved=0
For help, send mail to the list administrators:
Scott Douglas <mailto:[email protected]> Mike Cantwell 
<mailto:[email protected]> For policy questions, send mail to:
Jim Bacher <mailto:[email protected]>
David Heald <mailto:[email protected]>
-
----------------------------------------------------------------
This message is from the IEEE Product Safety Engineering Society emc-pstc 
discussion list. To post a message to the list, send your e-mail to 
<mailto:[email protected]> All emc-pstc postings are archived and searchable on 
the web at: 
https://emea01.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.ieee-pses.org%2Femc-pstc.html&data=01%7C01%7CAndy.McCallum%40mottmac.com%7C22523de7d06d4570c2c408d5e31197b8%7Ca2bed0c459574f73b0c2a811407590fb%7C0&sdata=2XdRLyQbOgOH62TTRXCBhVVm18K3dKVIIxXtahKZxhk%3D&reserved=0
Attachments are not permitted but the IEEE PSES Online Communities site at 
https://emea01.safelinks.protection.outlook.com/?url=http%3A%2F%2Fproduct-compliance.oc.ieee.org%2F&data=01%7C01%7CAndy.McCallum%40mottmac.com%7C22523de7d06d4570c2c408d5e31197b8%7Ca2bed0c459574f73b0c2a811407590fb%7C0&sdata=N625izyKbER4FJGjsv2xPjQ0InnEeRU6omzltg3ANhQ%3D&reserved=0
 can be used for graphics (in well-used formats), large files, etc.
Website: 
https://emea01.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.ieee-pses.org%2F&data=01%7C01%7CAndy.McCallum%40mottmac.com%7C22523de7d06d4570c2c408d5e31197b8%7Ca2bed0c459574f73b0c2a811407590fb%7C0&sdata=uRNIxlw1y%2BiLDmupgC52VcVzgrsyK3oR8LSqCCf95pk%3D&reserved=0
Instructions: 
https://emea01.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.ieee-pses.org%2Flist.html&data=01%7C01%7CAndy.McCallum%40mottmac.com%7C22523de7d06d4570c2c408d5e31197b8%7Ca2bed0c459574f73b0c2a811407590fb%7C0&sdata=kzljLVqzCWogvZM68PXZsf%2F2mhHPotXvkdogaXboNSI%3D&reserved=0
List rules: 
https://emea01.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.ieee-pses.org%2Flistrules.html&data=01%7C01%7CAndy.McCallum%40mottmac.com%7C22523de7d06d4570c2c408d5e31197b8%7Ca2bed0c459574f73b0c2a811407590fb%7C0&sdata=Xk6Vunv806vdykueUDEPdhh1ZR0k%2BPizXrWrtNE0Jw0%3D&reserved=0
For help, send mail to the list administrators:
Scott Douglas <mailto:[email protected]> Mike Cantwell 
<mailto:[email protected]> For policy questions, send mail to:
Jim Bacher <mailto:[email protected]>
David Heald <mailto:[email protected]>
-
----------------------------------------------------------------
This message is from the IEEE Product Safety Engineering Society emc-pstc 
discussion list. To post a message to the list, send your e-mail to 
<mailto:[email protected]> All emc-pstc postings are archived and searchable on 
the web at: 
https://emea01.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.ieee-pses.org%2Femc-pstc.html&data=01%7C01%7CAndy.McCallum%40mottmac.com%7C22523de7d06d4570c2c408d5e31197b8%7Ca2bed0c459574f73b0c2a811407590fb%7C0&sdata=2XdRLyQbOgOH62TTRXCBhVVm18K3dKVIIxXtahKZxhk%3D&reserved=0
Attachments are not permitted but the IEEE PSES Online Communities site at 
https://emea01.safelinks.protection.outlook.com/?url=http%3A%2F%2Fproduct-compliance.oc.ieee.org%2F&data=01%7C01%7CAndy.McCallum%40mottmac.com%7C22523de7d06d4570c2c408d5e31197b8%7Ca2bed0c459574f73b0c2a811407590fb%7C0&sdata=N625izyKbER4FJGjsv2xPjQ0InnEeRU6omzltg3ANhQ%3D&reserved=0
 can be used for graphics (in well-used formats), large files, etc.
Website: 
https://emea01.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.ieee-pses.org%2F&data=01%7C01%7CAndy.McCallum%40mottmac.com%7C22523de7d06d4570c2c408d5e31197b8%7Ca2bed0c459574f73b0c2a811407590fb%7C0&sdata=uRNIxlw1y%2BiLDmupgC52VcVzgrsyK3oR8LSqCCf95pk%3D&reserved=0
Instructions: 
https://emea01.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.ieee-pses.org%2Flist.html&data=01%7C01%7CAndy.McCallum%40mottmac.com%7C22523de7d06d4570c2c408d5e31197b8%7Ca2bed0c459574f73b0c2a811407590fb%7C0&sdata=kzljLVqzCWogvZM68PXZsf%2F2mhHPotXvkdogaXboNSI%3D&reserved=0
List rules: 
https://emea01.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.ieee-pses.org%2Flistrules.html&data=01%7C01%7CAndy.McCallum%40mottmac.com%7C22523de7d06d4570c2c408d5e31197b8%7Ca2bed0c459574f73b0c2a811407590fb%7C0&sdata=Xk6Vunv806vdykueUDEPdhh1ZR0k%2BPizXrWrtNE0Jw0%3D&reserved=0
For help, send mail to the list administrators:
Scott Douglas <mailto:[email protected]> Mike Cantwell 
<mailto:[email protected]> For policy questions, send mail to:
Jim Bacher <mailto:[email protected]>
David Heald <mailto:[email protected]>

________________________________________


WILO SE
Nortkirchenstrasse 100, 44263 Dortmund
Amtsgericht Dortmund, HRB 21356
https://emea01.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.wilo.com&data=01%7C01%7CAndy.McCallum%40mottmac.com%7C15d40e00326b4219a4df08d5e31ddb08%7Ca2bed0c459574f73b0c2a811407590fb%7C0&sdata=Bez9FHPD2vM5z2GHhU6F6uUZ4dPFfRSHtW0kh0LjMvY%3D&reserved=0

Vorstand/Executive Board:
Oliver Hermes (Vorsitzender/Chairman), Carsten Krumm, Georg Weber, Mathias 
Weyers Aufsichtsratsvorsitzender/Chairman of the Supervisory Board:
Prof. Dr. Norbert Wieselhuber

Dieses Dokument ist vertraulich zu behandeln. Die Weitergabe sowie 
Vervielfaeltigung, Verwertung und Mitteilung seines Inhalts ist nur mit unserer 
ausdruecklichen Genehmigung gestattet. Alle Rechte vorbehalten, insbesondere 
fuer den Fall der Schutzrechtsanmeldung.

This document has to be treated confidentially. Its contents are not to be 
passed on, duplicated, exploited or disclosed without our expressed permission. 
All rights reserved, especially the right to apply for protective rights.
-
----------------------------------------------------------------
This message is from the IEEE Product Safety Engineering Society emc-pstc 
discussion list. To post a message to the list, send your e-mail to 
<mailto:[email protected]> All emc-pstc postings are archived and searchable on 
the web at: 
https://emea01.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.ieee-pses.org%2Femc-pstc.html&data=01%7C01%7CAndy.McCallum%40mottmac.com%7C22523de7d06d4570c2c408d5e31197b8%7Ca2bed0c459574f73b0c2a811407590fb%7C0&sdata=2XdRLyQbOgOH62TTRXCBhVVm18K3dKVIIxXtahKZxhk%3D&reserved=0
Attachments are not permitted but the IEEE PSES Online Communities site at 
https://emea01.safelinks.protection.outlook.com/?url=http%3A%2F%2Fproduct-compliance.oc.ieee.org%2F&data=01%7C01%7CAndy.McCallum%40mottmac.com%7C22523de7d06d4570c2c408d5e31197b8%7Ca2bed0c459574f73b0c2a811407590fb%7C0&sdata=N625izyKbER4FJGjsv2xPjQ0InnEeRU6omzltg3ANhQ%3D&reserved=0
 can be used for graphics (in well-used formats), large files, etc.
Website: 
https://emea01.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.ieee-pses.org%2F&data=01%7C01%7CAndy.McCallum%40mottmac.com%7C22523de7d06d4570c2c408d5e31197b8%7Ca2bed0c459574f73b0c2a811407590fb%7C0&sdata=uRNIxlw1y%2BiLDmupgC52VcVzgrsyK3oR8LSqCCf95pk%3D&reserved=0
Instructions: 
https://emea01.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.ieee-pses.org%2Flist.html&data=01%7C01%7CAndy.McCallum%40mottmac.com%7C22523de7d06d4570c2c408d5e31197b8%7Ca2bed0c459574f73b0c2a811407590fb%7C0&sdata=kzljLVqzCWogvZM68PXZsf%2F2mhHPotXvkdogaXboNSI%3D&reserved=0
List rules: 
https://emea01.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.ieee-pses.org%2Flistrules.html&data=01%7C01%7CAndy.McCallum%40mottmac.com%7C22523de7d06d4570c2c408d5e31197b8%7Ca2bed0c459574f73b0c2a811407590fb%7C0&sdata=Xk6Vunv806vdykueUDEPdhh1ZR0k%2BPizXrWrtNE0Jw0%3D&reserved=0
For help, send mail to the list administrators:
Scott Douglas <mailto:[email protected]> Mike Cantwell 
<mailto:[email protected]> For policy questions, send mail to:
Jim Bacher <mailto:[email protected]>
David Heald <mailto:[email protected]>

________________________________


WILO SE
Nortkirchenstrasse 100, 44263 Dortmund
Amtsgericht Dortmund, HRB 21356
https://emea01.safelinks.protection.outlook.com/?url=www.wilo.com&data=01%7C01%7CAndy.McCallum%40mottmac.com%7C15d40e00326b4219a4df08d5e31ddb08%7Ca2bed0c459574f73b0c2a811407590fb%7C0&sdata=KW8uuDiky2Lto6NYtb9aiFvLjWkrGeIs8qp0cQYL55U%3D&reserved=0

Vorstand/Executive Board:
Oliver Hermes (Vorsitzender/Chairman), Carsten Krumm, Georg Weber, Mathias 
Weyers Aufsichtsratsvorsitzender/Chairman of the Supervisory Board:
Prof. Dr. Norbert Wieselhuber

Dieses Dokument ist vertraulich zu behandeln. Die Weitergabe sowie 
Vervielfaeltigung, Verwertung und Mitteilung seines Inhalts ist nur mit unserer 
ausdruecklichen Genehmigung gestattet. Alle Rechte vorbehalten, insbesondere 
fuer den Fall der Schutzrechtsanmeldung.

This document has to be treated confidentially. Its contents are not to be 
passed on, duplicated, exploited or disclosed without our expressed permission. 
All rights reserved, especially the right to apply for protective rights.

-
----------------------------------------------------------------
This message is from the IEEE Product Safety Engineering Society emc-pstc 
discussion list. To post a message to the list, send your e-mail to 
<[email protected]>

All emc-pstc postings are archived and searchable on the web at:
http://www.ieee-pses.org/emc-pstc.html

Attachments are not permitted but the IEEE PSES Online Communities site at 
http://product-compliance.oc.ieee.org/ can be used for graphics (in well-used 
formats), large files, etc.

Website:  http://www.ieee-pses.org/
Instructions:  http://www.ieee-pses.org/list.html (including how to unsubscribe)
List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:
Scott Douglas <[email protected]>
Mike Cantwell <[email protected]>

For policy questions, send mail to:
Jim Bacher:  <[email protected]>
David Heald: <[email protected]>

Reply via email to