Thanks all for your responses. šŸ™‚ Youā€™ve given me a few clues that I can use to 
start a deeper dive. šŸ‘šŸ¤“

Doug Nix
d...@ieee.org
+1 (519) 729-5704

> On 5-Feb-19, at 15:59, MIKE SHERMAN <msherma...@comcast.net> wrote:
> 
> I see another probably more likely interpretation of "knife edge and finger 
> stock" in this context. Look up a knife edge fuse on the internet. Presumably 
> the socket the fuse seats in would be the finger stock.
> 
> 
> 
> Mike Sherman
> 
> Graco Inc.
> 
>> On February 5, 2019 at 2:34 PM "Schmidt, Mark" <markschm...@xrite.com> 
>> wrote: 
>> 
>> Please see test below:
>> 
>>   
>> ā€œSafety Interlocks [21 CFR 1020.40(c)(4)(i)]:Each door of a cabinet x-ray 
>> system shall have a minimum of two safety interlocks. One, but not both of 
>> the required interlocks shall be such that door opening results in physical 
>> disconnection of the energy supply circuit to the high-voltage generator, 
>> and such disconnection shall not be dependent upon any moving part other 
>> than the door. [Note: Safety interlock is defined in 21 CFR 1020.40(b)(11).]
>> 
>>   
>> The primary door interlock should be of conventional design. The second door 
>> interlock must physically disconnect the energy supply circuit to the high 
>> voltage generator. Physical disconnection means opening the energy supply 
>> circuit by removing a piece of the circuit. This is usually accomplished 
>> with a ā€˜knife-edge and finger stockā€™ or ā€˜plug and socketā€™ type connection. 
>> It is good practice to assure the primary interlock will remove power from 
>> the energy supply circuit before the physical disconnection occurs. Relays, 
>> micro switches, and ā€˜safety switchesā€™ all contain moving parts and, 
>> therefore, cannot be used to satisfy the requirement that the physical 
>> disconnect interlock is "not dependent on any moving part other than the 
>> door."ā€
>> 
>>   
>> Compliance Guide for Cabinet X-Ray Systems  
>> https://www.fda.gov/MedicalDevices/ucm094358.htm 
>> <https://www.fda.gov/MedicalDevices/ucm094358.htm>
>>   
>>   
>> Mark
>> 
>>   
>>   
>> From: Nyffenegger, Dave [mailto:dave.nyffeneg...@bhemail.com] 
>> Sent: Tuesday, February 5, 2019 12:31 PM
>> To: EMC-PSTC@LISTSERV.IEEE.ORG
>> Subject: Re: [PSES] A question about FDA language and interlocking devices
>> 
>>  
>> That seems to make more sense.  I donā€™t know anything about X-ray 
>> requirements but the FDA requirements laser interlocks are not specific, 
>> only referring to ā€œinterlocksā€.
>> 
>> -Dave
>> 
>>   
>> From: MIKE SHERMAN [mailto:msherma...@comcast.net 
>> <mailto:msherma...@comcast.net>] 
>> Sent: Tuesday, February 05, 2019 12:27 PM
>> To: EMC-PSTC@LISTSERV.IEEE.ORG <mailto:EMC-PSTC@LISTSERV.IEEE.ORG>
>> Subject: Re: [PSES] A question about FDA language and interlocking devices
>> 
>>  
>> Doug --
>> 
>>   
>> "knife-edge and finger stock" sounds more like the gasketing around the door 
>> to keep the radiation inside. I think you'll find similar gasketing on doors 
>> to EMC chambers. I've not heard of this term applied to interlock switches.
>> 
>>   
>> Mike Sherman
>> 
>> Graco Inc.
>> 
>>   
>> On February 5, 2019 at 10:35 AM Doug Nix <d...@ieee.org 
>> <mailto:d...@ieee.org>> wrote: 
>> 
>> Colleagues,
>> 
>>  
>> I had a question come my way yesterday that I need a little help with. 
>> Hereā€™s the question:
>> 
>>  
>> I have a customer that produces X-ray equipment. The FDA requires that the 
>> door that gives access to the X-ray source must have an interlock with a 
>> ā€˜knife-edge and finger stockā€™ type connection.  Also the FDA mentions that 
>> interlocks should be of conventional design. What is understood by 
>> ā€œconventional designā€? 
>> 
>>  
>> My reading on this requirement is that any conventional electromechanical 
>> interlocking device like this:
>> 
>> <image001.jpg>
>> 
>> will meet the basic requirements as described by the FDA as ā€œknife-edge and 
>> fingerstockā€ connection, but I am concerned that this may not be at all what 
>> is meant. 
>> 
>>  
>> As always, any guidance you can offer will be welcomed and appreciated!
>> 
>>  
>> Best,
>> 
>>  
>> Doug Nix 
>> d...@ieee.org <mailto:d...@ieee.org> 
>> +1 (519) 729-5704

-
----------------------------------------------------------------
This message is from the IEEE Product Safety Engineering Society emc-pstc 
discussion list. To post a message to the list, send your e-mail to 
<emc-p...@ieee.org>

All emc-pstc postings are archived and searchable on the web at:
http://www.ieee-pses.org/emc-pstc.html

Attachments are not permitted but the IEEE PSES Online Communities site at 
http://product-compliance.oc.ieee.org/ can be used for graphics (in well-used 
formats), large files, etc.

Website:  http://www.ieee-pses.org/
Instructions:  http://www.ieee-pses.org/list.html (including how to unsubscribe)
List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:
Scott Douglas <sdoug...@ieee.org>
Mike Cantwell <mcantw...@ieee.org>

For policy questions, send mail to:
Jim Bacher:  <j.bac...@ieee.org>
David Heald: <dhe...@gmail.com>

Reply via email to