Hi Stacey

we recommend processing all of your subjects the same way. Do you understand why some would have a big bias field and some would not? It was always the same coil, wasn't it?

cheers
Bruce
On Mon, 9 Jul 2018, STACEY M SCHAEFER wrote:


        External Email - Use Caution        

 

Hello Freesurfer experts,

Our current 3T T1s exhibit structural hyperintensities due to the inhomogeneity 
of the receiver coil, but not all scans have
hyperintensity issues.  The local MRI physicist suggests running 
N4BiasFieldCorrection through multiple iterations prior to
Freesurfer processing on problematic scans.

We would like your opinion as to whether running a bias correction on the 
individual problematic scans could introduce any
additional biases in Freesurfer processing. This is a longitudinal sample of ~ 
300 participants. Is it OK to run
N4BiasFieldCorrection on individual scans as needed?  Or would it be better to 
run the bias correction on all the scans (the
entire sample) routinely, even if they do not show hyperintensity issues so all 
scans are processed similarly across the
sample?   Or do you have any other suggestions?  


Your guidance is much appreciated!
Thank you.
Stacey




Stacey M. Schaefer, Ph.D.
Center for Healthy Minds

University of Wisconsin-Madison

625 W Washington Ave

Madison WI 53703

608-263-9321



_______________________________________________
Freesurfer mailing list
[email protected]
https://mail.nmr.mgh.harvard.edu/mailman/listinfo/freesurfer


The information in this e-mail is intended only for the person to whom it is
addressed. If you believe this e-mail was sent to you in error and the e-mail
contains patient information, please contact the Partners Compliance HelpLine at
http://www.partners.org/complianceline . If the e-mail was sent to you in error
but does not contain patient information, please contact the sender and properly
dispose of the e-mail.

Reply via email to