Hi Martin and Satra, The HCP itself has not extensively evaluated a 1.5 mm multiband DWI acquisition on the Trio platform, so we don't have anything concrete to share at this time. Perhaps we should rephrase that sentence in the Appendix I release guide, but it was included primarily to encourage individuals on other platforms to pilot beyond the "standard" 2 mm resolution.
I'd encourage you to start by piloting ~1.8 mm isotropic on Trio/Verio scanners with a multiband factor of 3-4. cheers, -MH -- Michael Harms, Ph.D. ----------------------------------------------------------- Conte Center for the Neuroscience of Mental Disorders Washington University School of Medicine Department of Psychiatry, Box 8134 660 South Euclid Ave. Tel: 314-747-6173 St. Louis, MO 63110 Email: [email protected] On 7/7/13 10:46 AM, "Reiß-Zimmermann, Martin" <[email protected]> wrote: >Hello Gordon, > >thank you very much for your answers. Due to our resolution limited to 2 >mm isotropic we have a substantial lower TR (3270 vs. HCP 5520 ms) and a >TE of 110.4 ms (HCP 89.50 ms). I wasn't sure if it is the one gradient >direction that needs to be changed (since it always stopped at the same >point) or something else (in this case using a longer TR/TE). >Thanks to your answer, I will now gradually increase TR in a >trial-and-error-fashion. > >Any chance, the mentioned 1.5mm-DTI-TRIO-sequence can be posted here? (I >guess, there are a couple more TRIO users in the mailing list) > >Best regards, >Martin > > >________________________________________ >Von: Xu, Junqian [[email protected]] >Gesendet: Samstag, 6. Juli 2013 16:57 >An: Reiß-Zimmermann, Martin >Cc: [email protected] >Betreff: Re: [HCP-Users] HCP-dMRI gradient table on a TRIO - any >experience? > >Martin, > >Sorry that I misunderstood your problem. However It is still gradient >duty cycle related. The Siemens VB software version doesn't estimate >gradient heating as accurately/conservatively as the Siemens VD software >version. Hence the minimum TE/TR allowed by the sequence can still trip >the gradient duty cycle limit in real time, causing scan crash. > >The solution is to relax the TE/TR little by little with try-and-error >until you find a stable combination that passes your test. > >Gordon > >On Jul 5, 2013, at 9:12 PM, "Xu, Junqian" <[email protected]> wrote: > >> Martin, >> >> The culprit is the gradient power amplifier (GPA) duty cycle too high >>on the Y axis (i.e. the consecutive high value of 0.992 and -0.975). We >>have encountered similar problem in HCP and have implemented a solution >>to solve the gradient duty cycle issue. >> >> You can find the solution to this problem in Sotiropoulos et al. >>NeuroImage special issue >> >> http://www.ncbi.nlm.nih.gov/pubmed/23702418 >> >> Section: Q-space sampling >> >> "However, even when realistic sequences of diffusion gradient >>orientations were used (even with interleaved shells), the gradient duty >>cycle, due to accumulated heating, could still limit the achievable >>minimal TR. The order of the diffusion gradient orientations in the >>final HCP dMRI protocol was therefore optimised by swapping adjacent >>orientations to reduce the auto-correlation (lag = 1) of absolute >>gradient amplitude (~ < 0.5) for each physical gradient axis." >> >> Please let me know if this is clear enough for you to solve your >>specific case. >> >> Regards >> Gordon >> >> On Jul 5, 2013, at 11:06 AM, "Reiß-Zimmermann, Martin" >><[email protected]> wrote: >> >>> Dear all, >>> >>> I tried implementing and converting the HCP dMRI-sequences for our >>>TRIO. I followed the suggestions in the Appendix I and with minor >>>modifications (2 mm isotropic) it works well with MB3 and the >>>MDDW-gradient table (60 directions). >>> >>> However, when I use the HCP gradient table the measurement stops at >>>reproducable stops within the sequence (see following) and always with >>>the same error (hope I can translate that somehow from German into >>>English): >>> Error in gradient amplifier, GPA not ready. duty cycle is to high. >>>Please check sequence parameters. >>> >>> The sequences always stop at these points: >>> >>> 95dir-dMRI: direction number 47 (-1.000 , -0.001 , -0.011) >>> 96dir-dMRI: direction number 75 (-0.094 , 0.992 , -0.082) >>> 97dir-dMRI: direction number 55 (-0.134 , -0.975 , 0.180) >>> >>> The images before look great! >>> >>> I took a look at the page of Emmanuel Caruyer - but if I create a >>>gradient table of 90 dir (the HCP 96dir-table seems to have 6 b0 and >>>the rest b1000-3000, the E.C.-table seems to encode b>0 values, right?) >>>it is totally different from the HCP one. >>> >>> I also read the note in the HCP gradient table that these gradients >>>are customized for the Connectome-Scanner. >>> >>> So my question is: Shall I just use one created on Emmanuel's page or >>>can I (or better: someone else) change the one from HCP? >>> >>> The Appendix I also mentions that the dMRI-sequence has been >>>successfully tested on a TRIO and Verio with 1.5 mm isotropic >>>resolution (and b2000). Is it possible that you could post the protocol >>>(including the gradient table). >>> >>> Thank you very much. Best regards, >>> Martin >>> ____________________________________________________ >>> Dr. med. Martin Reiss-Zimmermann >>> Abteilung für Neuroradiologie >>> Universitätsklinikum Leipzig AöR >>> Liebigstr. 20 >>> 04103 Leipzig >>> >>> Mail: [email protected] >>> Fon: +49 - 341 - 97 17 585 >>> Fax: +49 - 341 - 97 17 490 >>> Net: http://www.neuroradiologie-leipzig.de >>> >>>........................................................................ >>>........................................................................ >>>........................................................ >>> Universitätsklinikum Leipzig - Anstalt öffentlichen Rechts | >>>Aufsichtsratsvorsitzender: Prof. Dr. Wolfram Knapp >>> Medizinischer Vorstand und Sprecher: Prof. Dr. med. Wolfgang E. Fleig >>>| Kaufmännischer Vorstand: Dipl.-Kfm. Ekkehard Zimmer >>> Sitz: Liebigstraße 18, 04103 Leipzig | Telefon +49-341-971709 | >>>St.Nr.: 23114904074 | IK: 261401052 >>> Bankverbindung: Deutsche Bank | BLZ: 86070000 | Kto.: 1247998 | >>>Swift-Code: DEUTDE8L | IBAN: DE27860700000124799800 >>> Internet: www.uniklinikum-leipzig.de >>> >>> Universität Leipzig | Ritterstr. 26, 04109 Leipzig | Körperschaft des >>>Öffentlichen Rechts | Vertreten durch die Rektorin Prof. Dr. Beate >>>Schücking >>> Zuständige Aufsichtsbehörde: SMWK, Wigardstr. 17, 01097 Dresden, >>>www.smwk.de >>> Umsatzsteuer-Identifikationsnummer gemäß §27a Umsatzsteuergesetz DE >>>141510383 >>> >>> Diese E-Mail enthält vertrauliche und/oder rechtlich geschützte >>>Informationen. Sollten Sie nicht der richtige Empfänger sein oder Sie >>>haben diese E-Mail irrtümlich erhalten, informieren Sie uns bitte >>>sofort und vernichten diese E-Mail. Das unerlaubte Kopieren sowie die >>>unbefugte Weitergabe dieser E-Mail und der darin enthaltenen >>>Informationen ist nicht gestattet. >>> >>> _______________________________________________ >>> HCP-Users mailing list >>> [email protected] >>> http://lists.humanconnectome.org/mailman/listinfo/hcp-users >> >> >> _______________________________________________ >> HCP-Users mailing list >> [email protected] >> http://lists.humanconnectome.org/mailman/listinfo/hcp-users > > >_______________________________________________ >HCP-Users mailing list >[email protected] >http://lists.humanconnectome.org/mailman/listinfo/hcp-users ________________________________ The materials in this message are private and may contain Protected Healthcare Information or other information of a sensitive nature. If you are not the intended recipient, be advised that any unauthorized use, disclosure, copying or the taking of any action in reliance on the contents of this information is strictly prohibited. If you have received this email in error, please immediately notify the sender via telephone or return mail. _______________________________________________ HCP-Users mailing list [email protected] http://lists.humanconnectome.org/mailman/listinfo/hcp-users