Hi,

Martin Reiß-Zimmermann has kindly alerted me to the fact that "Sense 1
Coil Combine" was NOT checked in the protocols I have uploaded to share.
This was an error, and "Sense 1" SHOULD be better checked under the
"Sequence / Special" tab.
I am actually using "Sense 1" (cf. Q2-Appendix recommendations of the HCP)
- the error was due to exporting from a backup folder where I had not
checked this.

I also want to mention that the protocol >should< be compatible to be
processed by eddy, including a possible lsr restoration of future releases
Jesper Andersson is working on.

Cheers,
Andreas

Am 14.07.13 17:50 schrieb "Andreas Bartsch" unter <[email protected]>:

>Hi,
>
>I have put my 1.8mm isotropic Trio diffusion protocol for download here:
>Archiv.zip <http://www.logicalenemy.net/data/wue/hcp/Archiv.zip>
>
>
>This is without GRAPPA but using the latest MB-GRAPPA reco refinements
>(cf. Steen Moeller's ISMRM poster this year) you may be able to use
>GRAPPA 
>at an acceleration of 2, preferably with 72 reference lines, and cut down
>the TE to something like 73ms.
>I had observed a predominance of V1 RL/LR orientations (a "reddish
>bleeding" on the ColFA or dtifit V1/FA overlays) when using GRAPPA on our
>system. That does not seem to be completely resolved even though it has
>improved. It does not seem to affect probabilistic tractography with Xing
>fibers but it is the reason why I am still staying away from GRAPPA.
>
>Feedback and improvements are welcome, as is a notice and acknowledgement
>if you should use this for your studies.
>Cheers,
>Andreas
>
>
>Am 10.07.13 19:00 schrieb "[email protected]" unter
><[email protected]>:
>
>>Send HCP-Users mailing list submissions to
>>      [email protected]
>>
>>To subscribe or unsubscribe via the World Wide Web, visit
>>      http://lists.humanconnectome.org/mailman/listinfo/hcp-users
>>or, via email, send a message with subject or body 'help' to
>>      [email protected]
>>
>>You can reach the person managing the list at
>>      [email protected]
>>
>>When replying, please edit your Subject line so it is more specific
>>than "Re: Contents of HCP-Users digest..."
>>
>>
>>Today's Topics:
>>
>>   1. Re: HCP-Users Digest, Vol 8, Issue 6 (Rei?-Zimmermann)
>>
>>
>>----------------------------------------------------------------------
>>
>>Message: 1
>>Date: Tue, 9 Jul 2013 17:25:33 +0000
>>From: Rei?-Zimmermann, Martin
>>      <[email protected]>
>>Subject: Re: [HCP-Users] HCP-Users Digest, Vol 8, Issue 6
>>To: Andreas Bartsch <[email protected]>,
>>      "[email protected]" <[email protected]>
>>Message-ID:
>>      <41e41d2635df1d4fb3430e087ba6f49328738...@s050003233.medizin.uni-leipzig
>>.
>>de>
>>      
>>Content-Type: text/plain; charset="iso-8859-1"
>>
>>Great news! Thank you very much Andreas.
>>
>>Best,
>>Martin
>>________________________________________
>>Von: Andreas Bartsch [[email protected]]
>>Gesendet: Montag, 8. Juli 2013 22:16
>>An: [email protected]
>>Betreff: Re: [HCP-Users] HCP-Users Digest, Vol 8, Issue 6
>>
>>Hi,
>>
>>I can share a 1.8 mm isotropic mb-diffusion EPI from a 'standard' Trio.
>>Cheers,
>>Andreas Bartsch
>>
>>Am 08.07.13 19:00 schrieb "[email protected]" unter
>><[email protected]>:
>>
>>>Send HCP-Users mailing list submissions to
>>>       [email protected]
>>>
>>>To subscribe or unsubscribe via the World Wide Web, visit
>>>       http://lists.humanconnectome.org/mailman/listinfo/hcp-users
>>>or, via email, send a message with subject or body 'help' to
>>>       [email protected]
>>>
>>>You can reach the person managing the list at
>>>       [email protected]
>>>
>>>When replying, please edit your Subject line so it is more specific
>>>than "Re: Contents of HCP-Users digest..."
>>>
>>>
>>>Today's Topics:
>>>
>>>   1. Re: HCP-dMRI gradient table on a TRIO - any experience?
>>>      (Harms, Michael)
>>>
>>>
>>>----------------------------------------------------------------------
>>>
>>>Message: 1
>>>Date: Mon, 8 Jul 2013 14:44:26 +0000
>>>From: "Harms, Michael" <[email protected]>
>>>Subject: Re: [HCP-Users] HCP-dMRI gradient table on a TRIO - any
>>>       experience?
>>>To: Rei?-Zimmermann, Martin
>>>       <[email protected]>, "Xu, Junqian"
>>>       <[email protected]>
>>>Cc: "[email protected]" <[email protected]>
>>>Message-ID: <ce003573.1b7fe%[email protected]>
>>>Content-Type: text/plain; charset="iso-8859-1"
>>>
>>>
>>>Hi Martin and Satra,
>>>The HCP itself has not extensively evaluated a 1.5 mm multiband DWI
>>>acquisition on the Trio platform, so we don't have anything concrete to
>>>share at this time.  Perhaps we should rephrase that sentence in the
>>>Appendix I release guide, but it was included primarily to encourage
>>>individuals on other platforms to pilot  beyond the "standard" 2 mm
>>>resolution.
>>>
>>>I'd encourage you to start by piloting ~1.8 mm isotropic on Trio/Verio
>>>scanners with a multiband factor of 3-4.
>>>
>>>cheers,
>>>-MH
>>>
>>>--
>>>Michael Harms, Ph.D.
>>>
>>>-----------------------------------------------------------
>>>Conte Center for the Neuroscience of Mental Disorders
>>>Washington University School of Medicine
>>>Department of Psychiatry, Box 8134
>>>660 South Euclid Ave.           Tel: 314-747-6173
>>>St. Louis, MO  63110                    Email: [email protected]
>>>
>>>
>>>
>>>
>>>On 7/7/13 10:46 AM, "Rei?-Zimmermann, Martin"
>>><[email protected]> wrote:
>>>
>>>>Hello Gordon,
>>>>
>>>>thank you very much for your answers. Due to our resolution limited to
>>>>2
>>>>mm isotropic we have a substantial lower TR (3270 vs. HCP 5520 ms) and
>>>>a
>>>>TE of 110.4 ms (HCP 89.50 ms). I wasn't sure if it is the one gradient
>>>>direction that needs to be changed (since it always stopped at the same
>>>>point) or something else (in this case using a longer TR/TE).
>>>>Thanks to your answer, I will now gradually increase TR in a
>>>>trial-and-error-fashion.
>>>>
>>>>Any chance, the mentioned 1.5mm-DTI-TRIO-sequence can be posted here?
>>>>(I
>>>>guess, there are a couple more TRIO users in the mailing list)
>>>>
>>>>Best regards,
>>>>Martin
>>>>
>>>>
>>>>________________________________________
>>>>Von: Xu, Junqian [[email protected]]
>>>>Gesendet: Samstag, 6. Juli 2013 16:57
>>>>An: Rei?-Zimmermann, Martin
>>>>Cc: [email protected]
>>>>Betreff: Re: [HCP-Users] HCP-dMRI gradient table on a TRIO - any
>>>>experience?
>>>>
>>>>Martin,
>>>>
>>>>Sorry that I misunderstood your problem. However It is still gradient
>>>>duty cycle related. The Siemens VB software version doesn't estimate
>>>>gradient heating as accurately/conservatively as the Siemens VD
>>>>software
>>>>version. Hence the minimum TE/TR allowed by the sequence can still trip
>>>>the gradient duty cycle limit in real time, causing scan crash.
>>>>
>>>>The solution is to relax the TE/TR little by little with try-and-error
>>>>until you find a stable combination that passes your test.
>>>>
>>>>Gordon
>>>>
>>>>On Jul 5, 2013, at 9:12 PM, "Xu, Junqian" <[email protected]> wrote:
>>>>
>>>>> Martin,
>>>>>
>>>>> The culprit is the gradient power amplifier (GPA) duty cycle too high
>>>>>on the Y axis (i.e. the consecutive high value of 0.992 and -0.975).
>>>>>We
>>>>>have encountered similar problem in HCP and have implemented a
>>>>>solution
>>>>>to solve the gradient duty cycle issue.
>>>>>
>>>>> You can find the solution to this problem in Sotiropoulos et al.
>>>>>NeuroImage special issue
>>>>>
>>>>> http://www.ncbi.nlm.nih.gov/pubmed/23702418
>>>>>
>>>>> Section: Q-space sampling
>>>>>
>>>>> "However, even when realistic sequences of diffusion gradient
>>>>>orientations were used (even with interleaved shells), the gradient
>>>>>duty
>>>>>cycle, due to accumulated heating, could still limit the achievable
>>>>>minimal TR. The order of the diffusion gradient orientations in the
>>>>>final HCP dMRI protocol was therefore optimised by swapping adjacent
>>>>>orientations to reduce the auto-correlation (lag = 1) of absolute
>>>>>gradient amplitude (~ < 0.5) for each physical gradient axis."
>>>>>
>>>>> Please let me know if this is clear enough for you to solve your
>>>>>specific case.
>>>>>
>>>>> Regards
>>>>> Gordon
>>>>>
>>>>> On Jul 5, 2013, at 11:06 AM, "Rei?-Zimmermann, Martin"
>>>>><[email protected]> wrote:
>>>>>
>>>>>> Dear all,
>>>>>>
>>>>>> I tried implementing and converting the HCP dMRI-sequences for our
>>>>>>TRIO. I followed the suggestions in the Appendix I and with minor
>>>>>>modifications (2 mm isotropic) it works well with MB3 and the
>>>>>>MDDW-gradient table (60 directions).
>>>>>>
>>>>>> However, when I use the HCP gradient table the measurement stops at
>>>>>>reproducable stops within the sequence (see following) and always
>>>>>>with
>>>>>>the same error (hope I can translate that somehow from German into
>>>>>>English):
>>>>>> Error in gradient amplifier, GPA not ready. duty cycle is to high.
>>>>>>Please check sequence parameters.
>>>>>>
>>>>>> The sequences always stop at these points:
>>>>>>
>>>>>> 95dir-dMRI: direction number 47 (-1.000 , -0.001 , -0.011)
>>>>>> 96dir-dMRI: direction number 75 (-0.094 , 0.992 , -0.082)
>>>>>> 97dir-dMRI: direction number 55 (-0.134 , -0.975 , 0.180)
>>>>>>
>>>>>> The images before look great!
>>>>>>
>>>>>> I took a look at the page of Emmanuel Caruyer - but if I create a
>>>>>>gradient table of 90 dir (the HCP 96dir-table seems to have 6 b0 and
>>>>>>the rest b1000-3000, the E.C.-table seems to encode b>0 values,
>>>>>>right?)
>>>>>>it is totally different from the HCP one.
>>>>>>
>>>>>> I also read the note in the HCP gradient table that these gradients
>>>>>>are customized for the Connectome-Scanner.
>>>>>>
>>>>>> So my question is: Shall I just use one created on Emmanuel's page
>>>>>>or
>>>>>>can I (or better: someone else) change the one from HCP?
>>>>>>
>>>>>> The Appendix I also mentions that the dMRI-sequence has been
>>>>>>successfully tested on a TRIO and Verio with 1.5 mm isotropic
>>>>>>resolution (and b2000). Is it possible that you could post the
>>>>>>protocol
>>>>>>(including the gradient table).
>>>>>>
>>>>>> Thank you very much. Best regards,
>>>>>> Martin
>>>>>> ____________________________________________________
>>>>>> Dr. med. Martin Reiss-Zimmermann
>>>>>> Abteilung f?r Neuroradiologie
>>>>>> Universit?tsklinikum Leipzig A?R
>>>>>> Liebigstr. 20
>>>>>> 04103 Leipzig
>>>>>>
>>>>>> Mail: [email protected]
>>>>>> Fon:  +49 - 341 - 97 17 585
>>>>>> Fax:   +49 - 341 - 97 17 490
>>>>>> Net:   http://www.neuroradiologie-leipzig.de
>>>>>>
>>>>>>.....................................................................
>>>>>>.
>>>>>>.
>>>>>>.
>>>>>>.....................................................................
>>>>>>.
>>>>>>.
>>>>>>.
>>>>>>........................................................
>>>>>> Universit?tsklinikum Leipzig - Anstalt ?ffentlichen Rechts |
>>>>>>Aufsichtsratsvorsitzender: Prof. Dr. Wolfram Knapp
>>>>>> Medizinischer Vorstand und Sprecher: Prof. Dr. med. Wolfgang E.
>>>>>>Fleig
>>>>>>| Kaufm?nnischer Vorstand: Dipl.-Kfm. Ekkehard Zimmer
>>>>>> Sitz: Liebigstra?e 18, 04103 Leipzig | Telefon +49-341-971709 |
>>>>>>St.Nr.: 23114904074 | IK: 261401052
>>>>>> Bankverbindung: Deutsche Bank | BLZ: 86070000 | Kto.: 1247998 |
>>>>>>Swift-Code: DEUTDE8L | IBAN: DE27860700000124799800
>>>>>> Internet: www.uniklinikum-leipzig.de
>>>>>>
>>>>>> Universit?t Leipzig | Ritterstr. 26, 04109 Leipzig | K?rperschaft
>>>>>>des
>>>>>>?ffentlichen Rechts | Vertreten durch die Rektorin Prof. Dr. Beate
>>>>>>Sch?cking
>>>>>> Zust?ndige Aufsichtsbeh?rde: SMWK, Wigardstr. 17, 01097 Dresden,
>>>>>>www.smwk.de
>>>>>> Umsatzsteuer-Identifikationsnummer gem?? ?27a Umsatzsteuergesetz DE
>>>>>>141510383
>>>>>>
>>>>>> Diese E-Mail enth?lt vertrauliche und/oder rechtlich gesch?tzte
>>>>>>Informationen. Sollten Sie nicht der richtige Empf?nger sein oder Sie
>>>>>>haben diese E-Mail irrt?mlich erhalten, informieren Sie uns bitte
>>>>>>sofort und vernichten diese E-Mail. Das unerlaubte Kopieren sowie die
>>>>>>unbefugte Weitergabe dieser E-Mail und der darin enthaltenen
>>>>>>Informationen ist nicht gestattet.
>>>>>>
>>>>>> _______________________________________________
>>>>>> HCP-Users mailing list
>>>>>> [email protected]
>>>>>> http://lists.humanconnectome.org/mailman/listinfo/hcp-users
>>>>>
>>>>>
>>>>> _______________________________________________
>>>>> HCP-Users mailing list
>>>>> [email protected]
>>>>> http://lists.humanconnectome.org/mailman/listinfo/hcp-users
>>>>
>>>>
>>>>_______________________________________________
>>>>HCP-Users mailing list
>>>>[email protected]
>>>>http://lists.humanconnectome.org/mailman/listinfo/hcp-users
>>>
>>>
>>>________________________________
>>>The materials in this message are private and may contain Protected
>>>Healthcare Information or other information of a sensitive nature. If
>>>you
>>>are not the intended recipient, be advised that any unauthorized use,
>>>disclosure, copying or the taking of any action in reliance on the
>>>contents of this information is strictly prohibited. If you have
>>>received
>>>this email in error, please immediately notify the sender via telephone
>>>or return mail.
>>>
>>>
>>>
>>>------------------------------
>>>
>>>_______________________________________________
>>>HCP-Users mailing list
>>>[email protected]
>>>http://lists.humanconnectome.org/mailman/listinfo/hcp-users
>>>
>>>
>>>End of HCP-Users Digest, Vol 8, Issue 6
>>>***************************************
>>
>>
>>
>>
>>
>>
>>------------------------------
>>
>>_______________________________________________
>>HCP-Users mailing list
>>[email protected]
>>http://lists.humanconnectome.org/mailman/listinfo/hcp-users
>>
>>
>>End of HCP-Users Digest, Vol 8, Issue 8
>>***************************************



_______________________________________________
HCP-Users mailing list
[email protected]
http://lists.humanconnectome.org/mailman/listinfo/hcp-users

Reply via email to