Find your channel in the automation explorer where you set up your
virtual channels. Select properties for the channel configured for
temperature. You may specify the CJC Source select built in by using
the drop down box.

In the devices where your SCXI Chassis is listed, you will need to
make sure you have the correct terminal block attached to the module.
For the ll02 you indicated in your question you have the 1303 or the
1300, check your terminal block and make sure you have the correct one
selected. This is important.

For changing the resistors I recommend you read the documentation for
the device to make sure you change the correct channels. Also you will
find this documentation better written and pictures help illustrate
the instructions.

If your thermocouple is not grounded and you do have the 1303 terminal
block, the factory shipped configuration will work. But it is probably
best for you to set up the block as suggested for thermocouples. The
10 Mohms (10x-1-106) resistor is recommended.

As I mentioned the documentation supplied with these products are a
bettor source of information for some of the questions you are asking.