Hi Sudheer,
As Dr. Irit answered. I would like to add that the Nanodrop actually gives
the absorbance and not the concentration, but shows the absorbance units as
mg/mL concentrations. So if you read 2.7 (absorbance units, AU), the
Nanodrop will show it as 2.7 mg/mL.
If I remember,the pathlength of the Nanodrop for the Beer-Lambert law is 1
mm, but the software shows the calculation converted for 10 mm (1 cm,
therefore  l = 1).
You can submit the protein sequence to this online tool:
http://web.expasy.org/protparam/
It will shows you the extinction coefficient, and the expected abosrbance
for a 1 mg/mL solution of your protein.

Again, for your experiment, if you read on the Nanodrop for example an
absorbance of 5, you need to use A=C.e.l
l is 1, A is 5, and after you get e (ext. coeff.) from the online tool, you
can calculate the actual C for your protein.
Regards
toufic el arnaout


Today's Topics:
>
>    1. protein extinction coefficient (Sudheer Sangeetham)
>    2. Re: protein extinction coefficient (Irit Rappley)
>
>
> ----------------------------------------------------------------------
>
> Message: 1
> Date: Wed, 30 Jan 2013 18:33:17 +0100
> From: Sudheer Sangeetham <[email protected]>
> Subject: protein extinction coefficient
> To: [email protected]
> Message-ID:
>         <CAAdHTim=
> [email protected]>
> Content-Type: text/plain; charset=ISO-8859-1
>
> Hello everyone
>
> I wanted to check my protein concentration based on its extinction
> coefficient by using nanodrop rather than doing Bradford or Lowry methods.
> I was checking the article in one article, The protein concentrations were
> estimated using the extinction coefficient of 2.70 at 280 nm for a 1mg/ml
> solution, but to the same protein in another article they determined the
> concentration by giving 66350 /m/cm. If i want to check my protein
> concentration by using nanodrop which value I should concern and how did
> they get value of 2.7 ???
>
> if anyone has idea please reply me
>
> Thanking you in advance
>
> Cheers
>
> --
> Sudheer Babu.S
> Research Fellow
> Institute of Biochemistry
> Biological Research Center
> Szeged,Hungary.
>
>
> ------------------------------
>
> Message: 2
> Date: Wed, 30 Jan 2013 11:36:13 -0800
> From: Irit Rappley <[email protected]>
> Subject: Re: protein extinction coefficient
> To: Sudheer Sangeetham <[email protected]>
> Cc: "[email protected]" <[email protected]>
> Message-ID: <[email protected]>
> Content-Type: text/plain; charset="iso-8859-1"
>
> Hi Sudheer,
>
> The units for the extinction coefficient are 1/(M*cm), which match the
> second value you mention. I would use that number. The calculation is (OD
> 280nm) / (extinction coefficient). This will give you a concentration in
> moles. You can multiply by 10^6 to get the concentration in micromoles,
> which is more likely to be in the right range for your protein.
>
> Cheers,
> Irit
>
>
>
> On Jan 30, 2013, at 9:33 AM, Sudheer Sangeetham wrote:
>
> > Hello everyone
> >
> > I wanted to check my protein concentration based on its extinction
> > coefficient by using nanodrop rather than doing Bradford or Lowry
> methods.
> > I was checking the article in one article, The protein concentrations
> were
> > estimated using the extinction coefficient of 2.70 at 280 nm for a 1mg/ml
> > solution, but to the same protein in another article they determined the
> > concentration by giving 66350 /m/cm. If i want to check my protein
> > concentration by using nanodrop which value I should concern and how did
> > they get value of 2.7 ???
> >
> > if anyone has idea please reply me
> >
> > Thanking you in advance
> >
> > Cheers
> >
> > --
> > Sudheer Babu.S
> > Research Fellow
> > Institute of Biochemistry
> > Biological Research Center
> > Szeged,Hungary.
> > _______________________________________________
> > Methods mailing list
> > [email protected]
> > http://www.bio.net/biomail/listinfo/methods
>
>
>
>
> ------------------------------
>
> _______________________________________________
> Methods mailing list
> [email protected]
> http://www.bio.net/biomail/listinfo/methods
>
> End of Methods Digest, Vol 92, Issue 4
> **************************************
>
>


-- 
******************************************************