You’re exactly right Subutai. 
I’m running a large swarm again with the max properly set. Will share the 
results as they come.

Nick


> On Apr 16, 2015, at 11:28 PM, Subutai Ahmad <[email protected]> wrote:
> 
> 
> Looks like the score dropping to 0 coincided with a sharp rise in the data. I 
> suspect the encoder params (such as min/max) were not set correctly. If the 
> input hits the max value a constant signal will be sent to the HTM and the 
> anomaly score will be constant.
> 
> --Subutai
> 
> On Thu, Apr 16, 2015 at 8:38 AM, Matthew Taylor <[email protected] 
> <mailto:[email protected]>> wrote:
> Right after the apnea, the anomaly score dropped to 0 and flatlined. That 
> doesn't seem right at all. Is your code online anywhere? 
> 
> ---------
> Matt Taylor
> OS Community Flag-Bearer
> Numenta
> 
> On Thu, Apr 16, 2015 at 7:00 AM, Nicholas Mitri <[email protected] 
> <mailto:[email protected]>> wrote:
> Hey all, 
> 
> I’ve procured data from a CPAP machine (waveform plotted below wrapping from 
> subplot to the next). 
> Is this data suitable for HTM anomaly detection and specifically for NAB?
> 
> The peaks in the data are sighs while short flat regions are apneas. The 
> baseline for breathing also changes as seen by the slower trends in the data. 
> I’ve tried using HTM and the anomalies I got weren’t very consistent for 
> sighs (2nd figure). Even worse, apneas were learned immediately, probably due 
> to being quantized to a single value. 
> 
> I’m defending my thesis in 2 weeks and this is the last section to highlight 
> anomaly detection in HTM. I’d really like the thesis to end on a good note 
> with one of the better features of HTM so any advice on how to handle this 
> data is greatly appreciated. 
> 
> Thanks,
> Nick
> <patient1_cpap_25hz.png>
> <Screen Shot 2014-12-17 at 12.54.48 AM.png>
> 
> 

Reply via email to