1. What equipment do you use for Alpha smear and air sample counting?
Canberra / Tennelec 5XLB and Ludlum 2000 with 43/10 detector, we have
iSOLO but not yet in service
2. What equipment do you use for Beta smear and air sample counting?
Canberra / Tennelec 5XLB
3. What statistical tests (LAC) do you perform on your scaler counting
equipment, e.g., Chi-Squared Testing, Confidence Interval Testing,
etc.?
Chi-square test when establishing QC chart data following calibration,
daily checks of source count with +/- 3 sigma
4. Where / when do you perform long term performance checks, e.g., are
after instrument calibration, in the Cal Lab facility or at the use
location? Is this a RP Tech function or is a separate qualification
required?
daily checks when in use, RP tech function
5. For daily performance checks, what source confidence levels do you
use, e.g., within 2 sigma standard deviation?
source counts within +/- 3 sigma, in process of implementing option to
use +/ 3% of mean (from MARLAP manual). background values use fixed
(upper only) levels for pass / fail