Here at Los Alamos National Lab we use the following standard setup of our
AMS-4s and have the latitude to change the settings for specific
radionuclides.

Slow Alarm Interval at 10 minutes
Slow Alarm Setpoint at 5.00E-9 uCi/cc
Fast Alarm Interval at 60 seconds
Fast Alarm Setpoint at 7.00E-8 uCi/cc
Net Alarm Interval at 60 seconds
Net Alarm Setpoint at 6.00E+4 CPM
Stack Alarm Setpoint is not used
DAC-Hour Alarm Setpoint at 8 DAC-Hours
(DAC-Hours are an accumulated value from the Slow Alarm Intervals)

We use the DAC concentration values from 10CFR835 which are similar in
value to the 10CFR20 values.  Our typical DAC factors range from 3E-7 to
7E-9 uCi/cc.

I have a pdf of the Eberline AMS-4 Instrument Manual posted on my website
if anyone needs a copy.  (WWW.VOSS-ASSOCIATES.COM)

Our major and most common concern is for the alpha emitters and we
typically set our Alpha CAMs to the 8 DAC-hour value.

Tom

On Wed, May 3, 2017 6:57 am, Derek Hultquist wrote:
> At Cook we use 3E-9 uCi/cc with the AMS-4
>
> Thanks,
>
> Derek Hultquist
> Health Physicist
> D. C. Cook NPP
> (269) 465-5901 x2088
>
>
>
> From: Lamons, John [mailto:[email protected]]
> Sent: Wednesday, May 03, 2017 8:54 AM
> To: [email protected]
> Subject: [EXTERNAL] [powernet] Setting of Continuous Air Monitors
>
> This is an EXTERNAL email. STOP. THINK before you CLICK links or OPEN
> attachments.
>
> ________________________________
> I would like to benchmark what set point the industry uses for CAM units
> (ie AMS-4).  We recently had a question from the NRC as to why we set our
> units at 10.59 Bq/ft^3  (1.01 e-8 uCi/cc) instead of the posting limit, .3
> DAC (about 3e-9 uCi/cc).  Thanks for the help.
>
> John Lamons
> Health Physicist
> Comanche Peak Nuclear Power Plant
>



Powernet - a service of the Health Physics Society Power Reactor Section

To post message: Send or Reply to: [email protected]
To unsubscribe: send message to [email protected]
archive(Nov2015->): http://www.mail-archive.com/[email protected]/
archive(<Oct2015): http://www.mail-archive.com/[email protected]/
If questions, contact E. Scott Medling, CHP at [email protected]

Reply via email to