Hi Edward,

answers below:

On 07.01.2013, at 11:43, "Edward d'Auvergne" <[email protected]> wrote:
>> One strategy to overcome this was to
>> * record a HSQC/TROSY at different temperatures at the different 
>> spectrometers
>>  (conveniently by using Brukers multi_zgvt),
> 
> Did you use the trosy experiment for relaxation measurements?

Yes, that's why I'm asking. ;) 

I have the impression that the methanol method is not always perfectly 
reflecting the situation in an aqueous buffer with all kinds of salts in it, 
and that the huge methanol signals and the resulting radiation damping can make 
determination of the "true" MeOH peak maxima pretty difficult. They seem pretty 
broad to me. The maximum resolution of peak distance I can get with Topspin is 
0.02 ppm, and if I consider that 0.02 ppm already makes a difference of ~ 1 K I 
guess there is the possibility of significant error. We tried using deuterated 
methanol at atmospheric pressure (much less intense signal, sharper peaks, no 
boiling point artifacts due to odd pressures in a sealed tube) to see if there 
are any differences to our standard procedure with protonated MeOH, but I 
couldn't see any. 

To make matters worse, we have four solution spectrometers with Bruker 
cryoprobes and one usually used with a room temperature probe, and of course 
each and avery probe has a different design in the temperature unit. It's not 
hard to start believing that every machine is behaving completely differently. 

To make my point clear I made two figures. 

The following plot shows a series of spectra, which all have the same reference 
frequency. You can see that most signals are shifting, into different 
"directions". One exception is the (alanine) signal at 134 / 9.2 ppm, which is 
pretty stable over a range of 5 K. The series is color coded red = 305 K to 
purple = 310 K according to the usual methanol calibration.

https://dl.dropbox.com/u/4019316/temp-750-zoom.pdf

Now after adding another spectrum from a different spectrometer (in magenta, 
it's also a different sample in this case) I first have the problem of proper 
referencing, as I don't know if the diverging signal positions are a result of 
slightly inaccurate field calibrations or due to different temperatures. I 
referenced it to the said Ala signal which seems to be unimpressed by 
temperature changes.

The "magenta spectrum" should correspond to one of the spectra "in the middle", 
namely 307 K, but in reality it fits the 310 K spectrum much, much better.

https://dl.dropbox.com/u/4019316/temp-750-reference.pdf

What do you think? Isn't it odd that the temperature-calibrated spectra don't 
fit 100%? They should, and the actual sample seems like a better temperature 
indicator to me than a somewhat artificial MeOH sample. 

Currently I'm measuring a whole set of "TROSY-calibrated" spectra, I'll see if 
they give me different results in the consistency tests.

>> * select the temperature where the NH spectra are nearly 100% identical
> 
> If temperature calibration is important, then you should run the
> experiment on methanol or ethylene glycol to calibrate.  Otherwise if
> the spectra are the same but just shifted because of temperature, then
> it's not so important.

I never saw any differences in MeOH proton peak distance when running the 
different T1/T2/HetNOE experiments against a standard methanol sample, 
regardless if it was HSQC-based or TROSY-based. 

If you take your NH spectra from different, MeOH-calibrated magnets/temperature 
units and superimpose them – are they completely identical? Don't you see any 
differences? I always have to re-adjust my reference peak lists to find the 
peaks in the spectra of the different spectrometers.

> run a quick 1D after running a short version of the experiment.

That's what I did. I pulsed for 15-20 minutes for the system to equilibrate and 
immediatley took a proton 1D to determine the peak distance. 

Maybe I should consider using ethylen glycol since the methanol calibration 
method seems to be less suited for temperatures around 300-310 K where we're 
working with? Are you determining the distance "by hand" or via a peak-picking 
mechanism inside topspin? Do you use deuterated methanol or standard protonated 
one?

Cheers and thanks for the (as usual) quick answer
Martin


_______________________________________________
relax (http://www.nmr-relax.com)

This is the relax-users mailing list
[email protected]

To unsubscribe from this list, get a password
reminder, or change your subscription options,
visit the list information page at
https://mail.gna.org/listinfo/relax-users

Reply via email to