Hi,
Wouldn't a change of Ni filter for a thicker one at least reduce
significantly the problem. It would absorb mostly below the edge but still
will absorb a significant amount of all the radiation that goes through
reducing the tails? Or a monochromator is the only solution? Would a tube
replacement reduce some of the problem?
Thanks,
Leo


2013/9/16 Alan Coelho <[email protected]>

> Matt
>
> >I am not sure I understand how you would account for the additional
> >wavelengths in the beam.
>
> Tube tails is spectrally independent and its aberration shape is 2Th
> independent. Its realized in TOPAS by a convolution called stacked_hats
> which, as its name suggests, convolutes a series of stacked impulse
> functions. An intense narrow hat represents the beam and wider hats the
> penetration of the electrons in the tube.
>
> Cheers
> Alan
>
> -----Original Message-----
> From: Matt Beekman [mailto:[email protected]]
> Sent: Tuesday, 17 September 2013 2:08 AM
> To: Peter Y. Zavalij; Alan Coelho; [email protected];
> [email protected]
> Subject: RE: Instrumental resolution function
>
> Alan, Peter,
>
> Would you mind briefly elaborating on the procedure used to "fit tube
> tails?" I personally have only used GSAS for Rietveld refinement, and other
> than adding another phase(s) with slightly different lattice parameter(s) I
> am not sure I understand how you would account for the additional
> wavelengths in the beam.
>
> Many thanks in advance!
>
> Matt
>
> --
> Matt Beekman, Ph.D.
> Assistant Professor of Physics
> Department of Natural Sciences
> Oregon Institute of Technology
> 3201 Campus Drive
> Klamath Falls, OR 97601
> Tel: 541-885-1940
> Fax: 541-885-1849
> Email: [email protected]
> Web:
> http://www.oit.edu/faculty-staff/home-pages/natural-science/matt-beekman
>
>
>
> -----Original Message-----
> From: [email protected] [mailto:[email protected]] On
> Behalf
> Of Peter Y. Zavalij
> Sent: Monday, September 16, 2013 6:11 AM
> To: Alan Coelho; [email protected]; [email protected]
> Subject: RE: Instrumental resolution function
>
> I do exactly the same - fit tube tails using LaB6 standard and use the
> parameters (usually w/o fitting) in all other refinements. Although tube
> tails fitting helps a lot it's not perfect.
> Peter Zavalij
>
> -----Original Message-----
> From: Alan Coelho [mailto:[email protected]]
> Sent: Sunday, September 15, 2013 10:05 PM
> To: Peter Y. Zavalij; [email protected]; [email protected]
> Subject: RE: Instrumental resolution function
>
> Hi Peter, Mario
>
> I happen to have a Ni filter, Cu, LaB6 pattern  that has a very similar
> look
> to the one sent by Mario. Probably best to look at the 21.35 degrees 2Th
> peak. The Ni filter cut-off occurs at 20.6 degrees 2Th, the satellite group
> is hard to see due to the low angle and hence compression of the emission
> profile. What made sense in the pattern was to fit Tube tails. Across the
> whole patter the Rwp was reduced by 2.1% where Tube tails was considered
> and
> the fit at the shoulders was good.
>
> Cheers
> Alan
>
>
> -----Original Message-----
> From: [email protected] [mailto:[email protected]] On
> Behalf
> Of Peter Y. Zavalij
> Sent: Monday, 16 September 2013 10:55 AM
> To: [email protected]; [email protected]
> Subject: RE: Instrumental resolution function
>
> Hi Mario,
> The shoulder you observe is what's left from "white" after it is cut off by
> beta-filer. You could check the absorption edge of Ni and it is right at
> the
> shoulder you observe.
> Peter
>
> __________________________________
> Peter Zavalij
> X-ray Crystallographic Center
> University of Maryland
> College Park, MD 20742
> Phone/Fax: (301)405-1861
> http:/www.chem.umd.edu/crystallography
>
>
>
> -----Original Message-----
> From: [email protected] [mailto:[email protected]] On
> Behalf
> Of [email protected]
> Sent: Sunday, September 15, 2013 6:08 PM
> To: [email protected]
> Subject: Instrumental resolution function
>
> Dear All,
>
> I am calculating the corresponding instrumental resolution function of our
> Bruker diffractometer which is operating in Bragg-Brentano geometry with Cu
> radiation, using a nickel filter and LynxEye detector.
>
> The calculation is carried out by the Rietveld analysis of the LaB6 (660b)
> NIST-standard. We observed the Kbeta peaks, but also some shoulders in the
> left part of the peaks. We have some ideas about this but taking into
> account that I'm not an actual expert, your comments will be very important
> to me.
>
> In the attached file, I send an image to make a better illustration of my
> doubts.
>
> Thanks in advance for your help.
> Best Regards,
>
> Mario A. Macías
> Universidad Industrial de Santander
> Bucaramanga-Colombia.
>
>
>
> ++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++
> Please do NOT attach files to the whole list <[email protected]
> >
> Send commands to <[email protected]> eg: HELP as the subject with no body
> text
> The Rietveld_L list archive is on
> http://www.mail-archive.com/[email protected]/
> ++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++
>
>
>


-- 
Dr. Leopoldo Suescun
Prof. Agr (Assoc. Prof.) de Física       Tel: (+598) 29290648/29249859
Cryssmat-Lab./DETEMA                             Fax: (+598) 29241906