2019-08-30 Thread James Holton

If it is any of these things, then we want to know about it!

Miss-alignment I doubt because the cameras on 8.2.2 are all co-axial 
(looking down the beam).  The beam size is about 1/5 the field of view, 
and the sample stays visible during the exposure.  It would be hard not 
to notice the crystal moving out of the beam.
I also doubt fluorescence scans damaging the crystal because the general 
ALS fluorescence scan protocol attenuates the incident beam by a factor 
of 1000 or more.  Few fluorescence scans give the crystal more dose than 
a single screening shot.
The cryo can be a problem, but I also doubt it in this case.  Over 20 
years, I have seen maybe 3 instances of a cryo system "burping". These 
problems are extremely difficult to detect and even more difficult to 
solve.  Hence the concern here.  Usually this happens if the neck of the 
exhaust vent gets full of ice.  Checking that is routine PM. I just 
double-checked myself and its clean.


 All of the above things are very unlikely to happen twice in a row. 
Hopefully Kimberly can grow more crystals and we can try again (with 
more than a few eyes watching everything very carefully). But right now 
I'd bet even money on equipment vs the sample.


-James Holton
MAD Scientist

On 8/30/2019 11:48 AM, Green, Todd Jason wrote:
Certainly not trying to be insulting by the suggestions but could it 
be as simple an alignment issue with the crystal, issues with the beam 
alignment or crystal damage following a  fluorescence scan? The cryo 
setup?


Like James, I’ve seen crystals decay quickly in high salts at longer 
wavelengths.


Good luck!

- Todd

Sent from my iPhone

On Aug 30, 2019, at 1:41 PM, James Holton <mailto:jmhol...@lbl.gov>> wrote:



That is absolutely mind-blowing.

ALS 8.2.2 has a nominal dose rate of 95 kGy/s, and you are seeing 
global damage go to completion in ~0.4s, or 40 kGy. This is 750 times 
faster than expected.  I have never heard of a cryo-cooled crystal 
decaying that fast.  This may be a new world record!  I know that is 
probably not what you wanted, but it is exciting to people like me.


Being at 4.5A is actually a good thing for radiation damage because 
features at poor resolution tend to evolve more slowly with dose than 
fine details (high-angle spots).  That is probably also not what you 
wanted, but it makes such a fast decay rate even more unusual.


I have encountered reports of extra-sensitive proteins before, but 
they always turn out to be due to something predictable, like having 
a super-high concentration of heavy atom.  For example, having 200 mM 
sodium iodide in your solvent channels will cut a typical crystal 
lifetime in half.  4 M NaI will cut it by a factor of 20, but to get 
1000x the usual decay rate you would need something like 10 M 
Ta6Br12.  And Ta6Br12 is only soluble to 2 mM.


Yes, you have Se atoms in there, but even if you have 1 in 10 
residues containing an Se atom at 0% solvent that is still only 1.0 M 
Se in the crystal.  Only enough for about a factor of 10.


I have also seen errors in beam size and flux lead to apparently 
abnormal radiation damage rates.  These can easily be as large as a 
factor of two, especially with very small beams, but it is hard to 
imagine how one could get a factor of 1000.  ALS 8.2.2 is only a few 
meters from where I'm sitting, and I'm pretty sure the flux density 
is accurate to within 20%, probably better.  I also just checked out 
the cryo stream, and it has none of the usual warning signs.


So, that leaves either a genuinely ultra-sensitive protein, or some 
kind of undetected equipment malfunction.  Either way, we at the ALS 
are very interested in this result. Would you mind trying again?  And 
let me know when you are coming?


-James Holton
MAD Scientist


On 8/29/2019 11:07 AM, Kimberly Stanek wrote:
ALS 822. I tried as low as 0.2 sec exposure but it didn't seem to 
help much.



Kimberly Stanek
Postdoctoral Researcher, Goulding Lab
Co-chair, UCI Postdoctoral Association
University of California, Irvine
Natural Science I, Room 2302
(949) 824-0014

*From:* James Holton 
*Sent:* Thursday, August 29, 2019 10:50 AM
*To:* Kimberly Stanek ; CCP4BB@JISCMAIL.AC.UK 

*Subject:* Re: [ccp4bb] Semet derivative dying almost immediately in 
beam

What exposure time are you using?  And which beamline?

-James Holton
MAD Scientist

On 8/29/2019 10:26 AM, Kimberly Stanek wrote:

Hi folks,

We have a protein that we have been trying to solve the structure 
of for a while now but so far haven't been able to get any 
diffraction better than ~4.5A. I was able to collect a full 360 
degrees of data and index, but MR is failing so we have turned to 
de novo phasing.


Recently we prepared crystals of the Semet derivative under the 
same condition. While these crystals diffracted to about the same 
resolution, I found they were dying after just

Re: [ccp4bb] Semet derivative dying almost immediately in beam

2019-08-30 Thread Green, Todd Jason
Certainly not trying to be insulting by the suggestions but could it be as 
simple an alignment issue with the crystal, issues with the beam alignment or 
crystal damage following a  fluorescence scan? The cryo setup?

Like James, I’ve seen crystals decay quickly in high salts at longer 
wavelengths.

Good luck!

- Todd

Sent from my iPhone

On Aug 30, 2019, at 1:41 PM, James Holton 
mailto:jmhol...@lbl.gov>> wrote:

That is absolutely mind-blowing.

ALS 8.2.2 has a nominal dose rate of 95 kGy/s, and you are seeing global damage 
go to completion in ~0.4s, or 40 kGy.  This is 750 times faster than expected.  
I have never heard of a cryo-cooled crystal decaying that fast.  This may be a 
new world record!  I know that is probably not what you wanted, but it is 
exciting to people like me.

Being at 4.5A is actually a good thing for radiation damage because features at 
poor resolution tend to evolve more slowly with dose than fine details 
(high-angle spots).  That is probably also not what you wanted, but it makes 
such a fast decay rate even more unusual.

I have encountered reports of extra-sensitive proteins before, but they always 
turn out to be due to something predictable, like having a super-high 
concentration of heavy atom.  For example, having 200 mM sodium iodide in your 
solvent channels will cut a typical crystal lifetime in half.  4 M NaI will cut 
it by a factor of 20, but to get 1000x the usual decay rate you would need 
something like 10 M Ta6Br12.  And Ta6Br12 is only soluble to 2 mM.

Yes, you have Se atoms in there, but even if you have 1 in 10 residues 
containing an Se atom at 0% solvent that is still only 1.0 M Se in the crystal. 
 Only enough for about a factor of 10.

I have also seen errors in beam size and flux lead to apparently abnormal 
radiation damage rates.  These can easily be as large as a factor of two, 
especially with very small beams, but it is hard to imagine how one could get a 
factor of 1000.  ALS 8.2.2 is only a few meters from where I'm sitting, and I'm 
pretty sure the flux density is accurate to within 20%, probably better.  I 
also just checked out the cryo stream, and it has none of the usual warning 
signs.

So, that leaves either a genuinely ultra-sensitive protein, or some kind of 
undetected equipment malfunction.  Either way, we at the ALS are very 
interested in this result. Would you mind trying again?  And let me know when 
you are coming?

-James Holton
MAD Scientist


On 8/29/2019 11:07 AM, Kimberly Stanek wrote:
ALS 822. I tried as low as 0.2 sec exposure but it didn't seem to help much.


Kimberly Stanek
Postdoctoral Researcher, Goulding Lab
Co-chair, UCI Postdoctoral Association
University of California, Irvine
Natural Science I, Room 2302
(949) 824-0014

From: James Holton <mailto:jmhol...@lbl.gov>
Sent: Thursday, August 29, 2019 10:50 AM
To: Kimberly Stanek <mailto:ksta...@uci.edu>; 
CCP4BB@JISCMAIL.AC.UK<mailto:CCP4BB@JISCMAIL.AC.UK> 
<mailto:CCP4BB@JISCMAIL.AC.UK>
Subject: Re: [ccp4bb] Semet derivative dying almost immediately in beam

What exposure time are you using?  And which beamline?

-James Holton
MAD Scientist

On 8/29/2019 10:26 AM, Kimberly Stanek wrote:
Hi folks,

We have a protein that we have been trying to solve the structure of for a 
while now but so far haven't been able to get any diffraction better than 
~4.5A. I was able to collect a full 360 degrees of data and index, but MR is 
failing so we have turned to de novo phasing.

Recently we prepared crystals of the Semet derivative under the same condition. 
While these crystals diffracted to about the same resolution, I found they were 
dying after just one or two snaps, even with increased beam attenuation and 
decreased exposure time. I am wondering if anyone has experienced anything like 
this before and had any suggestions on what to do about it.

Thank you,



Kimberly Stanek
Postdoctoral Researcher, Goulding Lab
Co-chair, UCI Postdoctoral Association
University of California, Irvine
Natural Science I, Room 2302
(949) 824-0014



To unsubscribe from the CCP4BB list, click the following link:
https://www.jiscmail.ac.uk/cgi-bin/webadmin?SUBED1=CCP4BB&A=1




To unsubscribe from the CCP4BB list, click the following link:
https://www.jiscmail.ac.uk/cgi-bin/webadmin?SUBED1=CCP4BB&A=1




To unsubscribe from the CCP4BB list, click the following link:
https://www.jiscmail.ac.uk/cgi-bin/webadmin?SUBED1=CCP4BB&A=1



To unsubscribe from the CCP4BB list, click the following link:
https://www.jiscmail.ac.uk/cgi-bin/webadmin?SUBED1=CCP4BB&A=1


Re: [ccp4bb] Semet derivative dying almost immediately in beam

2019-08-30 Thread James Holton

That is absolutely mind-blowing.

ALS 8.2.2 has a nominal dose rate of 95 kGy/s, and you are seeing global 
damage go to completion in ~0.4s, or 40 kGy.  This is 750 times faster 
than expected.  I have never heard of a cryo-cooled crystal decaying 
that fast.  This may be a new world record!  I know that is probably not 
what you wanted, but it is exciting to people like me.


Being at 4.5A is actually a good thing for radiation damage because 
features at poor resolution tend to evolve more slowly with dose than 
fine details (high-angle spots).  That is probably also not what you 
wanted, but it makes such a fast decay rate even more unusual.


I have encountered reports of extra-sensitive proteins before, but they 
always turn out to be due to something predictable, like having a 
super-high concentration of heavy atom.  For example, having 200 mM 
sodium iodide in your solvent channels will cut a typical crystal 
lifetime in half.  4 M NaI will cut it by a factor of 20, but to get 
1000x the usual decay rate you would need something like 10 M Ta6Br12.  
And Ta6Br12 is only soluble to 2 mM.


Yes, you have Se atoms in there, but even if you have 1 in 10 residues 
containing an Se atom at 0% solvent that is still only 1.0 M Se in the 
crystal.  Only enough for about a factor of 10.


I have also seen errors in beam size and flux lead to apparently 
abnormal radiation damage rates.  These can easily be as large as a 
factor of two, especially with very small beams, but it is hard to 
imagine how one could get a factor of 1000.  ALS 8.2.2 is only a few 
meters from where I'm sitting, and I'm pretty sure the flux density is 
accurate to within 20%, probably better.  I also just checked out the 
cryo stream, and it has none of the usual warning signs.


So, that leaves either a genuinely ultra-sensitive protein, or some kind 
of undetected equipment malfunction.  Either way, we at the ALS are very 
interested in this result. Would you mind trying again? And let me know 
when you are coming?


-James Holton
MAD Scientist


On 8/29/2019 11:07 AM, Kimberly Stanek wrote:
ALS 822. I tried as low as 0.2 sec exposure but it didn't seem to help 
much.



Kimberly Stanek
Postdoctoral Researcher, Goulding Lab
Co-chair, UCI Postdoctoral Association
University of California, Irvine
Natural Science I, Room 2302
(949) 824-0014

*From:* James Holton 
*Sent:* Thursday, August 29, 2019 10:50 AM
*To:* Kimberly Stanek ; CCP4BB@JISCMAIL.AC.UK 


*Subject:* Re: [ccp4bb] Semet derivative dying almost immediately in beam
What exposure time are you using?  And which beamline?

-James Holton
MAD Scientist

On 8/29/2019 10:26 AM, Kimberly Stanek wrote:

Hi folks,

We have a protein that we have been trying to solve the structure of 
for a while now but so far haven't been able to get any diffraction 
better than ~4.5A. I was able to collect a full 360 degrees of data 
and index, but MR is failing so we have turned to de novo phasing.


Recently we prepared crystals of the Semet derivative under the same 
condition. While these crystals diffracted to about the same 
resolution, I found they were dying after just one or two snaps, even 
with increased beam attenuation and decreased exposure time. I am 
wondering if anyone has experienced anything like this before and had 
any suggestions on what to do about it.


Thank you,



Kimberly Stanek
Postdoctoral Researcher, Goulding Lab
Co-chair, UCI Postdoctoral Association
University of California, Irvine
Natural Science I, Room 2302
(949) 824-0014



To unsubscribe from the CCP4BB list, click the following link:
https://www.jiscmail.ac.uk/cgi-bin/webadmin?SUBED1=CCP4BB&A=1






To unsubscribe from the CCP4BB list, click the following link:
https://www.jiscmail.ac.uk/cgi-bin/webadmin?SUBED1=CCP4BB&A=1






To unsubscribe from the CCP4BB list, click the following link:
https://www.jiscmail.ac.uk/cgi-bin/webadmin?SUBED1=CCP4BB&A=1


Re: [ccp4bb] Semet derivative dying almost immediately in beam

2019-08-29 Thread Francis Reyes
Ah, low resolution experimental phasing. Fun stuff, seems like a lost art these 
days.

Are the crystals large ? 
Is it possible to do a (continuous helical data) collection?  (Manually or if 
your beamline permits you to). 

Granted this is anomalous, is it possible to then merge enough partial datasets 
together? (Very carefully of course). If not there’s always SIRAS. 

With this rapid decay, are you using a strategy? (Another lost art these days)

Given the difficulty of this project, have you pursued other derivatives (via 
soaking)? 


> On Aug 29, 2019, at 11:26 AM, Kimberly Stanek  wrote:
> 
> Hi folks,
> 
> We have a protein that we have been trying to solve the structure of for a 
> while now but so far haven't been able to get any diffraction better than 
> ~4.5A. I was able to collect a full 360 degrees of data and index, but MR is 
> failing so we have turned to de novo phasing.
> 
> Recently we prepared crystals of the Semet derivative under the same 
> condition. While these crystals diffracted to about the same resolution, I 
> found they were dying after just one or two snaps, even with increased beam 
> attenuation and decreased exposure time. I am wondering if anyone has 
> experienced anything like this before and had any suggestions on what to do 
> about it.
> 
> Thank you,
> 
> 
> 
> Kimberly Stanek
> Postdoctoral Researcher, Goulding Lab
> Co-chair, UCI Postdoctoral Association
> University of California, Irvine
> Natural Science I, Room 2302
> (949) 824-0014
> 
> To unsubscribe from the CCP4BB list, click the following link:
> https://www.jiscmail.ac.uk/cgi-bin/webadmin?SUBED1=CCP4BB&A=1



To unsubscribe from the CCP4BB list, click the following link:
https://www.jiscmail.ac.uk/cgi-bin/webadmin?SUBED1=CCP4BB&A=1


Re: [ccp4bb] Semet derivative dying almost immediately in beam

2019-08-29 Thread graeme.win...@diamond.ac.uk
Their web page indicates they’re running a 315, so there are limits on useful 
low dose data collection. If you can get to a beamline with a pixel array 
detector and mount your samples carefully you’ll probably get more from low 
dose methods.

The key thing with this is minimising background from loop, drop etc so the 
photons you measure are as far as possible from the sample under study. You’ll 
probably benefit from combining data from multiple samples if they are 
isomorphous.

Obviously your ultimate resolution will be limited by the number of photons 
your sample can scatter before suffering measurable damage... but combining 
samples can help you here

Best wishes Graeme

On 29 Aug 2019, at 19:07, Kimberly Stanek 
mailto:ksta...@uci.edu>> wrote:

ALS 822. I tried as low as 0.2 sec exposure but it didn't seem to help much.


Kimberly Stanek
Postdoctoral Researcher, Goulding Lab
Co-chair, UCI Postdoctoral Association
University of California, Irvine
Natural Science I, Room 2302
(949) 824-0014

From: James Holton mailto:jmhol...@lbl.gov>>
Sent: Thursday, August 29, 2019 10:50 AM
To: Kimberly Stanek mailto:ksta...@uci.edu>>; 
CCP4BB@JISCMAIL.AC.UK<mailto:CCP4BB@JISCMAIL.AC.UK> 
mailto:CCP4BB@JISCMAIL.AC.UK>>
Subject: Re: [ccp4bb] Semet derivative dying almost immediately in beam

What exposure time are you using?  And which beamline?

-James Holton
MAD Scientist

On 8/29/2019 10:26 AM, Kimberly Stanek wrote:
Hi folks,

We have a protein that we have been trying to solve the structure of for a 
while now but so far haven't been able to get any diffraction better than 
~4.5A. I was able to collect a full 360 degrees of data and index, but MR is 
failing so we have turned to de novo phasing.

Recently we prepared crystals of the Semet derivative under the same condition. 
While these crystals diffracted to about the same resolution, I found they were 
dying after just one or two snaps, even with increased beam attenuation and 
decreased exposure time. I am wondering if anyone has experienced anything like 
this before and had any suggestions on what to do about it.

Thank you,



Kimberly Stanek
Postdoctoral Researcher, Goulding Lab
Co-chair, UCI Postdoctoral Association
University of California, Irvine
Natural Science I, Room 2302
(949) 824-0014



To unsubscribe from the CCP4BB list, click the following link:
https://www.jiscmail.ac.uk/cgi-bin/webadmin?SUBED1=CCP4BB&A=1




To unsubscribe from the CCP4BB list, click the following link:
https://www.jiscmail.ac.uk/cgi-bin/webadmin?SUBED1=CCP4BB&A=1

-- 
This e-mail and any attachments may contain confidential, copyright and or 
privileged material, and are for the use of the intended addressee only. If you 
are not the intended addressee or an authorised recipient of the addressee 
please notify us of receipt by returning the e-mail and do not use, copy, 
retain, distribute or disclose the information in or attached to the e-mail.
Any opinions expressed within this e-mail are those of the individual and not 
necessarily of Diamond Light Source Ltd. 
Diamond Light Source Ltd. cannot guarantee that this e-mail or any attachments 
are free from viruses and we cannot accept liability for any damage which you 
may sustain as a result of software viruses which may be transmitted in or with 
the message.
Diamond Light Source Limited (company no. 4375679). Registered in England and 
Wales with its registered office at Diamond House, Harwell Science and 
Innovation Campus, Didcot, Oxfordshire, OX11 0DE, United Kingdom




To unsubscribe from the CCP4BB list, click the following link:
https://www.jiscmail.ac.uk/cgi-bin/webadmin?SUBED1=CCP4BB&A=1


Re: [ccp4bb] Semet derivative dying almost immediately in beam

2019-08-29 Thread Kimberly Stanek
ALS 822. I tried as low as 0.2 sec exposure but it didn't seem to help much.


Kimberly Stanek
Postdoctoral Researcher, Goulding Lab
Co-chair, UCI Postdoctoral Association
University of California, Irvine
Natural Science I, Room 2302
(949) 824-0014

From: James Holton 
Sent: Thursday, August 29, 2019 10:50 AM
To: Kimberly Stanek ; CCP4BB@JISCMAIL.AC.UK 

Subject: Re: [ccp4bb] Semet derivative dying almost immediately in beam

What exposure time are you using?  And which beamline?

-James Holton
MAD Scientist

On 8/29/2019 10:26 AM, Kimberly Stanek wrote:
Hi folks,

We have a protein that we have been trying to solve the structure of for a 
while now but so far haven't been able to get any diffraction better than 
~4.5A. I was able to collect a full 360 degrees of data and index, but MR is 
failing so we have turned to de novo phasing.

Recently we prepared crystals of the Semet derivative under the same condition. 
While these crystals diffracted to about the same resolution, I found they were 
dying after just one or two snaps, even with increased beam attenuation and 
decreased exposure time. I am wondering if anyone has experienced anything like 
this before and had any suggestions on what to do about it.

Thank you,



Kimberly Stanek
Postdoctoral Researcher, Goulding Lab
Co-chair, UCI Postdoctoral Association
University of California, Irvine
Natural Science I, Room 2302
(949) 824-0014



To unsubscribe from the CCP4BB list, click the following link:
https://www.jiscmail.ac.uk/cgi-bin/webadmin?SUBED1=CCP4BB&A=1




To unsubscribe from the CCP4BB list, click the following link:
https://www.jiscmail.ac.uk/cgi-bin/webadmin?SUBED1=CCP4BB&A=1


Re: [ccp4bb] Semet derivative dying almost immediately in beam

2019-08-29 Thread James Holton

What exposure time are you using?  And which beamline?

-James Holton
MAD Scientist

On 8/29/2019 10:26 AM, Kimberly Stanek wrote:

Hi folks,

We have a protein that we have been trying to solve the structure of 
for a while now but so far haven't been able to get any diffraction 
better than ~4.5A. I was able to collect a full 360 degrees of data 
and index, but MR is failing so we have turned to de novo phasing.


Recently we prepared crystals of the Semet derivative under the same 
condition. While these crystals diffracted to about the same 
resolution, I found they were dying after just one or two snaps, even 
with increased beam attenuation and decreased exposure time. I am 
wondering if anyone has experienced anything like this before and had 
any suggestions on what to do about it.


Thank you,



Kimberly Stanek
Postdoctoral Researcher, Goulding Lab
Co-chair, UCI Postdoctoral Association
University of California, Irvine
Natural Science I, Room 2302
(949) 824-0014



To unsubscribe from the CCP4BB list, click the following link:
https://www.jiscmail.ac.uk/cgi-bin/webadmin?SUBED1=CCP4BB&A=1






To unsubscribe from the CCP4BB list, click the following link:
https://www.jiscmail.ac.uk/cgi-bin/webadmin?SUBED1=CCP4BB&A=1


[ccp4bb] Semet derivative dying almost immediately in beam

2019-08-29 Thread Kimberly Stanek
Hi folks,

We have a protein that we have been trying to solve the structure of for a 
while now but so far haven't been able to get any diffraction better than 
~4.5A. I was able to collect a full 360 degrees of data and index, but MR is 
failing so we have turned to de novo phasing.

Recently we prepared crystals of the Semet derivative under the same condition. 
While these crystals diffracted to about the same resolution, I found they were 
dying after just one or two snaps, even with increased beam attenuation and 
decreased exposure time. I am wondering if anyone has experienced anything like 
this before and had any suggestions on what to do about it.

Thank you,



Kimberly Stanek
Postdoctoral Researcher, Goulding Lab
Co-chair, UCI Postdoctoral Association
University of California, Irvine
Natural Science I, Room 2302
(949) 824-0014



To unsubscribe from the CCP4BB list, click the following link:
https://www.jiscmail.ac.uk/cgi-bin/webadmin?SUBED1=CCP4BB&A=1