[ccp4bb] X-Stream 2000 problem - ICING

2009-01-12 Thread Mark Agacan
Apologies for this slightly off topic question:  

I am having a great deal of trouble with my X-Stream 2000 cryostream system and 
I wondered if other users have similar problems.  

I've replaced almost all components (new GAST compressors, helium recharges, 
filters, etc., etc.) in the last couple of months but there is almost always 
icing of any cryo within 10 - 20 minutes of mounting a loop, and it is 
adversely affecting data collections.  

It appears like there is too much moisture in the cold or wam streams but the 
tubes have been fully dried out as per Rigaku advice.  

This X-Stream is attached to a generator with inverted phi axis and and i'm 
wondering if this could be the source of the problem, as the X-Stream for 
another generator in the same laboratory with normal phi axis does not ice up.  
Can some sort of turbulence around the loop caused by backdraft from the cryo 
hitting the inverted phi axis / camera mount cause excess humidity and lead to 
icing on the pin, loop and crystal?  

Has anyone else got this problem?  Any suggestions would be very gratefully 
appreciated.

Best Wishes,

Mark


_
Dr Mark Agacan
Scientific Officer,
Division of Biological Chemistry 
and Drug Discovery,
Wellcome Trust Biocentre,
College of Life Sciences,
Dow St., 
University of Dundee,
Dundee, DD1 5EH
Tel: +44 1382 388751
Fax: +44 1382 345764
_
The University of Dundee is a registered Scottish charity, No: SC015096


Re: [ccp4bb] X-Stream 2000 problem - ICING

2009-01-12 Thread Frank von Delft

Hi Mark

We had a LOT of pain with icing, and it really comes down to one thing:  
water in the gN2.  And don't expect to measure some other way whether 
you have it, because your X-stream (or Cobra) is the most sensitive 
water meter there is. 

In our case, the symptom was the X-stream (and later Cobra) blocking up 
after between several days and several hours.  And we solved it by 
ditching the gN2 generators we were using, and organising boil-off gN2 
(much purer).


The secondary effect of sample icing:  we'd see this as well if we had 
something (e.g. collimator) poking into the cold stream*:  that causes 
turbulence which draws in moisture.  Worst case you see ice flakes flick 
onto the crystal in real time;  best case you get an ice ball after a 
few minutes to hours. 


(* technically, the stationary phase between the cold and warm streams.)

Hope that helps.
phx


Mark Agacan wrote:
Apologies for this slightly off topic question:  

I am having a great deal of trouble with my X-Stream 2000 cryostream system and I wondered if other users have similar problems.  

I've replaced almost all components (new GAST compressors, helium recharges, filters, etc., etc.) in the last couple of months but there is almost always icing of any cryo within 10 - 20 minutes of mounting a loop, and it is adversely affecting data collections.  

It appears like there is too much moisture in the cold or wam streams but the tubes have been fully dried out as per Rigaku advice.  

This X-Stream is attached to a generator with inverted phi axis and and i'm wondering if this could be the source of the problem, as the X-Stream for another generator in the same laboratory with normal phi axis does not ice up.  Can some sort of turbulence around the loop caused by backdraft from the cryo hitting the inverted phi axis / camera mount cause excess humidity and lead to icing on the pin, loop and crystal?  


Has anyone else got this problem?  Any suggestions would be very gratefully 
appreciated.

Best Wishes,

Mark


_
Dr Mark Agacan
Scientific Officer,
Division of Biological Chemistry 
and Drug Discovery,

Wellcome Trust Biocentre,
College of Life Sciences,
Dow St., 
University of Dundee,

Dundee, DD1 5EH
Tel: +44 1382 388751
Fax: +44 1382 345764
_
The University of Dundee is a registered Scottish charity, No: SC015096
  


Re: [ccp4bb] X-Stream 2000 problem - ICING

2009-01-12 Thread Johan P. Turkenburg

Hi,

If you have ice on the crystal (loop) but no blockages of the cryostream 
itself (and this seems to be what you're saying), it is unlikely to be a 
problem with the LN2 (although I only have experience with Oxford 
Crystreams, but I imagine this also applies to X-streams).


Our main problems are turbulence (collimator poking into the gas 
stream(s) or gas 'bouncing' back from something just beyond the crystal 
- moving things even a little may help here) and drafts due to 
'powerful' air-conditioning.


As you state that one of your setups is affected and not the other, it 
may well be the case that an air conditioning vent is blowing onto your 
setup, and disturbing the gas streams. We resorted to hanging plastic 
sheeting around two sides of our set-up and this made a vast difference.


HTH,

Johan

Mark Agacan wrote:
Apologies for this slightly off topic question:  

I am having a great deal of trouble with my X-Stream 2000 cryostream system and I wondered if other users have similar problems.  

I've replaced almost all components (new GAST compressors, helium recharges, filters, etc., etc.) in the last couple of months but there is almost always icing of any cryo within 10 - 20 minutes of mounting a loop, and it is adversely affecting data collections.  

It appears like there is too much moisture in the cold or wam streams but the tubes have been fully dried out as per Rigaku advice.  

This X-Stream is attached to a generator with inverted phi axis and and i'm wondering if this could be the source of the problem, as the X-Stream for another generator in the same laboratory with normal phi axis does not ice up.  Can some sort of turbulence around the loop caused by backdraft from the cryo hitting the inverted phi axis / camera mount cause excess humidity and lead to icing on the pin, loop and crystal?  


Has anyone else got this problem?  Any suggestions would be very gratefully 
appreciated.

Best Wishes,

Mark


_
Dr Mark Agacan
Scientific Officer,
Division of Biological Chemistry 
and Drug Discovery,

Wellcome Trust Biocentre,
College of Life Sciences,
Dow St., 
University of Dundee,

Dundee, DD1 5EH
Tel: +44 1382 388751
Fax: +44 1382 345764
_
The University of Dundee is a registered Scottish charity, No: SC015096





--
+
Dr. Johan P. Turkenburg X-ray facilities manager
York Structural Biology Laboratory  
University of York  Phone (+) 44 1904 328251
York YO10 5DD   UK  Fax   (+) 44 1904 328266
+


Re: [ccp4bb] X-Stream 2000 problem - ICING

2009-01-12 Thread Santarsiero, Bernard D.
We've been able to run months with an old Xstream 2000 system, so that
shouldn't be the problem. Unlike Frank, we haven't had problems with water
in the nitrogen from a nitrogen generator.

If Frank is correct, that it's water, then either the molecular sieves
need to be replaced, or there is ice buildup and blockage in the coldhead.
Usually blockage in the coldhead means that you can't get down in temp, or
can't maintain it within 1 degree or less. We do bring the temp up, run it
at RT over the weekend to dry things out, and then bring it down for
routine cooling.

If it's turbulence, then you don't have the warm and cold flow rates set
correctly. One thing you can see immediately, upon getting down to 100K,
is how far is the frosting plume from the tip of the nozzle. If you adjust
the flow rates, you can move that plume, caused by turbulence, closer or
further from the nozzle. It's also possible that there is something in the
nozzle that is causing turbulence no matter what the flow rates are, and
your vendor needs to look at that.

Bernie Santarsiero


On Mon, January 12, 2009 6:11 am, Frank von Delft wrote:
 Hi Mark

 We had a LOT of pain with icing, and it really comes down to one thing:
 water in the gN2.  And don't expect to measure some other way whether
 you have it, because your X-stream (or Cobra) is the most sensitive
 water meter there is.

 In our case, the symptom was the X-stream (and later Cobra) blocking up
 after between several days and several hours.  And we solved it by
 ditching the gN2 generators we were using, and organising boil-off gN2
 (much purer).

 The secondary effect of sample icing:  we'd see this as well if we had
 something (e.g. collimator) poking into the cold stream*:  that causes
 turbulence which draws in moisture.  Worst case you see ice flakes flick
 onto the crystal in real time;  best case you get an ice ball after a
 few minutes to hours.

 (* technically, the stationary phase between the cold and warm streams.)

 Hope that helps.
 phx


 Mark Agacan wrote:
 Apologies for this slightly off topic question:

 I am having a great deal of trouble with my X-Stream 2000 cryostream
 system and I wondered if other users have similar problems.

 I've replaced almost all components (new GAST compressors, helium
 recharges, filters, etc., etc.) in the last couple of months but there
 is almost always icing of any cryo within 10 - 20 minutes of mounting a
 loop, and it is adversely affecting data collections.

 It appears like there is too much moisture in the cold or wam streams
 but the tubes have been fully dried out as per Rigaku advice.

 This X-Stream is attached to a generator with inverted phi axis and and
 i'm wondering if this could be the source of the problem, as the
 X-Stream for another generator in the same laboratory with normal phi
 axis does not ice up.  Can some sort of turbulence around the loop
 caused by backdraft from the cryo hitting the inverted phi axis / camera
 mount cause excess humidity and lead to icing on the pin, loop and
 crystal?

 Has anyone else got this problem?  Any suggestions would be very
 gratefully appreciated.

 Best Wishes,

 Mark


 _
 Dr Mark Agacan
 Scientific Officer,
 Division of Biological Chemistry
 and Drug Discovery,
 Wellcome Trust Biocentre,
 College of Life Sciences,
 Dow St.,
 University of Dundee,
 Dundee, DD1 5EH
 Tel: +44 1382 388751
 Fax: +44 1382 345764
 _
 The University of Dundee is a registered Scottish charity, No: SC015096




Re: [ccp4bb] X-Stream 2000 problem - ICING

2009-01-12 Thread mjvdwoerd

 Mark,

What bothers me about your message is that you already have talked to Rigaku. 
Until now we have never been able to create a problem that they could not 
diagnose and help me solve from remote. In danger of offending ccp4 readers: 
specialized Rigaku experts are a remarkable source for information and 
solutions, probably better than we are.

Your most likely problem is that your nitrogen is not dry? Specifically, check 
your air dryer (sorry, nitrogen dryer) that it works appropriately. Very 
specifically, there reside two compressors inside the air dryer and if one no 
longer works, the quality of your nitrogen stream degrades. It may not be 
apparent if both compressors work, one can supply all the pressure and volume 
you need and is sufficiently noisy that you would not notice the second being 
silent. Of course this problem becomes obvious when you open up the cabinet. 
(Yes, of course this happened to us once before and in our case the compressor 
wiring was fickle, as in, working when the cabinet was open and not (always) 
working when the cabinet was closed; took FOREVER to find the problem.)

Your second most likely reason is that the warm stream (outer stream) is not 
sufficiently protecting your cold stream from humidity, but this is not 
affected by your phi-axis position. We have two inverted phi-axes and we do not 
see icing, so there is no fundamental reason why the phi-axis should not be 
inverted.

Mark
? 


 


 

-Original Message-
From: Mark Agacan m.aga...@dundee.ac.uk
To: CCP4BB@JISCMAIL.AC.UK
Sent: Mon, 12 Jan 2009 3:42 am
Subject: [ccp4bb] X-Stream 2000 problem - ICING










Apologies for this slightly off topic question:  

I am having a great deal of trouble with my X-Stream 2000 cryostream system and 
I wondered if other users have similar problems.  

I've replaced almost all components (new GAST compressors, helium recharges, 
filters, etc., etc.) in the last couple of months but there is almost always 
icing of any cryo within 10 - 20 minutes of mounting a loop, and it is 
adversely 
affecting data collections.  

It appears like there is too much moisture in the cold or wam streams but the 
tubes have been fully dried out as per Rigaku advice.  

This X-Stream is attached to a generator with inverted phi axis and and i'm 
wondering if this could be the source of the problem, as the X-Stream for 
another generator in the same laboratory with normal phi axis does not ice up.  
Can some sort of turbulence around the loop caused by backdraft from the cryo 
hitting the inverted phi axis / camera mount cause excess humidity and lead to 
icing on the pin, loop and crystal?  

Has anyone else got this problem?  Any suggestions would be very gratefully 
appreciated.

Best Wishes,

Mark


_
Dr Mark Agacan
Scientific Officer,
Division of Biological Chemistry 
and Drug Discovery,
Wellcome Trust Biocentre,
College of Life Sciences,
Dow St., 
University of Dundee,
Dundee, DD1 5EH
Tel: +44 1382 388751
Fax: +44 1382 345764
_
The University of Dundee is a registered Scottish charity, No: SC015096