In general, most protein quantification methods
  have significant problems with idiosyncracy, because proteins are
  quite variable in composition and structure. The best method is to
  use is the absorption at 280 nm, but this is quantitatively useful
  only if the molar absorptivity is known (and the protein is pure,
  or at least purified away from UV-absorbing materials). We work
  extensively with metalloenzymes, and quantifying protein:metal
  ratios accurately is highly problematic. Here is my take:
  


  The Bradford (dye-binding) assay is quite dependent on protein
composition (hydrophobic content and perhaps other amino acid
composition). We see variances of apparent to actual protein
concentration of 50% or more
  The Lowry-type assays based on molybdenum blue formation (this
includes the BCA assay which is a variant of the classic Lowry
assay) are also highly composition dependent. We see variances
of apparent to acutal protein concentration of up to 100%,
depending on the protein involved.
  The Biuret assay is dependent primarily on the peptide
backbone content, and has small variation of response to
composition. It is not very sensitive, however. The microbiuret
assay, which is similar to biuret but with measurements made at
330 nm is tolerable. Protein must be homogeneous, of course. We
see relatively little variation (20% or so) between apparent and
true concentrations of protein using this method, but it is
subject to a variety of interferences.
  
  Far UV absorption of proteins is nearly independent of
composition, as absorption in this region is due almost entirely
to peptide bonds. This requires purified protein, and the
absence of other UV-absorbing materials, but is otherwise quite
tolerant. Maximal consistency can be obtained by carrying out
measurements in denaturing detergents. We have used 0.01-0.1%
Triton X-100 with success. The asborptivity of 1mg/mL solutions
is approximately 20.5 at 210 nm and around 34 at 205 nm. various
corrections can be applied to correct for absorption of
aromatics at 205 nm. See Scopes (1974) Anal Biochem 59, 277-282.
The systematic variance of this method is generally small, maybe
20% or less.
  

I have found that methods 3-4 frequently agree, and give good
agreement with well-characterized metalloproteins where the metal
can be very accurately measured using ICP-OES.

Cheers.

-- 
  

Roger S. Rowlett
Professor
Department of Chemistry
Colgate University
13 Oak Drive
Hamilton, NY 13346

tel: (315)-228-7245
ofc: (315)-228-7395
fax: (315)-228-7935
email: rrowl...@colgate.edu
  

  



Re: [ccp4bb] how to quantitate protein which dont have ne aromatic residue

2011-04-10 Thread E rajakumar
Hi Arpita

You can try QUANTI-iT
Protein assay kit from Invitrogen.

But still there is
nearly 20-50% discrepancy between this method and a Abosorbance at 280.

I also faced same
problem with a protein, then re-cloned by adding a Trp at the C-terminus.

Raj

 
E. Rajakumara
Postdoctoral Fellow  Strcutural Biology Program  Memorial Sloan-Kettering 
Cancer Center  New York-10021  NY  001 212 639 7986 (Lab)  001 917 674 6266 
(Mobile)

--- On Sat, 9/4/11, Arpit Mishra ar...@igib.in wrote:

From: Arpit Mishra ar...@igib.in
Subject: [ccp4bb] how to quantitate protein which dont have ne aromatic residue
To: CCP4BB@JISCMAIL.AC.UK
Date: Saturday, 9 April, 2011, 3:22 PM

hello everybody
i am working on the protien which dont have any aromatic residue  i do fplc 
other purification using 220 absorption, but i want to quantitate protein 
precisely i have tried using BCA nd bradford but both methods quantification is 
not matching,,so any one is having sum idea how to quantitate it precisely 

thanks in advance for your valuable suggestion..

Arpit Mishra


Re: [ccp4bb] how to quantitate protein which dont have ne aromatic residue

2011-04-10 Thread gauri misra
Hi,
You can read the spectrophotometric absorption at 280 nm and 200nm in UV
range.
It should serve your purpose and provide a decent idea for the amount of
protein in the sample.
Provided that absorbance at 280nm is given by aromatic rings but at the same
time absorbance at 200nm is contributed by peptide bonds. A simultaneous
reading at 260 nm which can be deducted later will substract the effect of
nucleic acids.
The relation between the readings at these values and protein concentration
you can easily find anywhere on the web.

Best wishes
Gauri


[ccp4bb] how to quantitate protein which dont have ne aromatic residue

2011-04-09 Thread Arpit Mishra
hello everybody

i am working on the protien which dont have any aromatic residue  i do fplc
other purification using 220 absorption, but i want to quantitate protein
precisely i have tried using BCA nd bradford but both methods quantification
is not matching,,so any one is having sum idea how to quantitate it
precisely

thanks in advance for your valuable suggestion..


Arpit Mishra


Re: [ccp4bb] how to quantitate protein which dont have ne aromatic residue

2011-04-09 Thread VAN RAAIJ , MARK JOHAN
you can do amino acid analysis on your pure protein, using a commercial or 
academic service - I hope these are still around. You should only need to do 
this once, then relate the result to your A220, BCA and Bradford assays. 
Mark 

Quoting Arpit Mishra:

 hello everybody

 i am working on the protien which dont have any aromatic residue  i do fplc
 other purification using 220 absorption, but i want to quantitate protein
 precisely i have tried using BCA nd bradford but both methods quantification
 is not matching,,so any one is having sum idea how to quantitate it
 precisely

 thanks in advance for your valuable suggestion..


 Arpit Mishra


Mark J van Raaij
Laboratorio M-4
Dpto de Estructura de Macromoléculas
Centro Nacional de Biotecnología - CSIC
c/Darwin 3, Campus Cantoblanco
28049 Madrid
tel. 91 585 4616
email: mjvanra...@cnb.csic.es 

Re: [ccp4bb] how to quantitate protein which dont have ne aromatic residue

2011-04-09 Thread Michael Thompson
It is not surprising that your bradford and BCA assays don't agree if you have 
no aromatic amino acids in your protein. Bradford dye binds to hydrophobic 
residues, mainly aromatics, so I would guess your bradford is consistantly 
giving lower measurements than the BCA assay. I also wouldn't be surprised if 
the results of your Bradford vary significantly between replicates. The BCA 
assay reagent interacts with the backbone amides, not with any sidechains, so I 
would tend to believe that measurement more than anything else you have done. 

I work with a protein that has very few hydrophobics (only one aromatic - a 
Phe) and I have found that Bradfords are unreliable, but the BCA assay tends to 
be consistent.

Mike




- Original Message -
From: Arpit Mishra ar...@igib.in
To: CCP4BB@JISCMAIL.AC.UK
Sent: Saturday, April 9, 2011 2:52:21 AM GMT -08:00 US/Canada Pacific
Subject: [ccp4bb] how to quantitate protein which dont have ne aromatic residue

hello everybody 


i am working on the protien which dont have any aromatic residue i do fplc 
other purification using 220 absorption, but i want to quantitate protein 
precisely i have tried using BCA nd bradford but both methods quantification is 
not matching,,so any one is having sum idea how to quantitate it precisely 


thanks in advance for your valuable suggestion.. 




Arpit Mishra 

-- 
Michael C. Thompson

Graduate Student

Biochemistry  Molecular Biology Division

Department of Chemistry  Biochemistry

University of California, Los Angeles

mi...@chem.ucla.edu


Re: [ccp4bb] how to quantitate protein which dont have ne aromatic residue

2011-04-09 Thread Philippe Dumas

Michael Thompson mi...@chem.ucla.edu a écrit :

There is a very simple and very quick  method that yields an answer  
approx. 15% reliable: measuring the increment of index of refraction  
due to the protein. The measurement of an index of refraction can be  
very accurate.  You only need something like a 5µl drop at 1 mg/ml  
(the order of magnitude should be correct...). Unfortunately, a  
refractometer is not common in biology labs, but this is a very  
valuable method.
The link between the increment of index of refraction and the protein  
conc. can be found easily on the web.

Philippe Dumas

It is not surprising that your bradford and BCA assays don't agree   
if you have no aromatic amino acids in your protein. Bradford dye   
binds to hydrophobic residues, mainly aromatics, so I would guess   
your bradford is consistantly giving lower measurements than the BCA  
 assay. I also wouldn't be surprised if the results of your Bradford  
 vary significantly between replicates. The BCA assay reagent   
interacts with the backbone amides, not with any sidechains, so I   
would tend to believe that measurement more than anything else you   
have done.


I work with a protein that has very few hydrophobics (only one   
aromatic - a Phe) and I have found that Bradfords are unreliable,   
but the BCA assay tends to be consistent.


Mike




- Original Message -
From: Arpit Mishra ar...@igib.in
To: CCP4BB@JISCMAIL.AC.UK
Sent: Saturday, April 9, 2011 2:52:21 AM GMT -08:00 US/Canada Pacific
Subject: [ccp4bb] how to quantitate protein which dont have ne   
aromatic residue


hello everybody


i am working on the protien which dont have any aromatic residue i   
do fplc other purification using 220 absorption, but i want to   
quantitate protein precisely i have tried using BCA nd bradford but   
both methods quantification is not matching,,so any one is having   
sum idea how to quantitate it precisely



thanks in advance for your valuable suggestion..




Arpit Mishra

--
Michael C. Thompson

Graduate Student

Biochemistry  Molecular Biology Division

Department of Chemistry  Biochemistry

University of California, Los Angeles

mi...@chem.ucla.edu



Re: [ccp4bb] how to quantitate protein which dont have ne aromatic residue

2011-04-09 Thread John A. Newitt

At 9:47 AM -0700 4/9/11, Michael Thompson wrote:


Bradford dye binds to hydrophobic residues, mainly aromatics,


The statement above is not accurate.

Compton and Jones. Anal. Biochem. 151(2): 369-374, 1985

- John
--


Re: [ccp4bb] how to quantitate protein which dont have ne aromatic residue

2011-04-09 Thread Dima Klenchin
It is not surprising that your bradford and BCA assays don't agree if you 
have no aromatic amino acids in your protein. Bradford dye binds to 
hydrophobic residues, mainly aromatics, so I would guess your bradford is 
consistantly giving lower measurements than the BCA assay. I also wouldn't 
be surprised if the results of your Bradford vary significantly between 
replicates. The BCA assay reagent interacts with the backbone amides, not 
with any sidechains, so I would tend to believe that measurement more than 
anything else you have done.


I work with a protein that has very few hydrophobics (only one aromatic - 
a Phe) and I have found that Bradfords are unreliable, but the BCA assay 
tends to be consistent.



Bradford reagent is colloidal Coomassie G-250 and it's binding to proteins 
is very complex, depending on local structure, hydrophobic interaction and 
basic charges (mainly Arg residues). So yes, it is quite variable protein 
to protein but it is not a simple function of aromatics.


- Dima


Re: [ccp4bb] how to quantitate protein which dont have ne aromatic residue

2011-04-09 Thread Michael Thompson
John,

I believe my statement is accurate. The Compton  Jones paper you cite states 
in the abstract: Interactions are chiefly with arginine rather than primary 
amino groups; the other basic (His, Lys) and aromatic residues (Try, Tyr, and 
Phe) give slight responses. The binding behavior is attributed to Van der Waals 
forces and hydrophobic interactions. Also, 23 years later, the following paper 
provides a detailed study of the mechanism of Coomassie Brilliant Blue G-250 
binding to proteins and discusses the hydrophobic interactions I mentioned 
between aromatic residues and the dye.

Georgiou, et al. Anal Bioanal Chem. 2008 May;391(1):391-403. 

Cheers,

Mike



- Original Message -
From: John A. Newitt newit...@gmail.com
To: CCP4BB@JISCMAIL.AC.UK
Sent: Saturday, April 9, 2011 12:06:23 PM GMT -08:00 US/Canada Pacific
Subject: Re: [ccp4bb] how to quantitate protein which dont have ne aromatic 
residue

At 9:47 AM -0700 4/9/11, Michael Thompson wrote:

Bradford dye binds to hydrophobic residues, mainly aromatics,

The statement above is not accurate.

Compton and Jones. Anal. Biochem. 151(2): 369-374, 1985

- John
-- 

-- 
Michael C. Thompson

Graduate Student

Biochemistry  Molecular Biology Division

Department of Chemistry  Biochemistry

University of California, Los Angeles