Re: [ccp4bb] Ca or Zn

2012-11-01 Thread Armin Wagner

Dear John,

Thanks for pointing to the long-wavelength MX beamline project here at Diamond.

The core wavelength range will be from 3 to 8 keV (1.5 to 4.1 A) and mainly 
targeting anomalous phasing experiments from native proteins/DNA/RNA exploiting 
anomalous differences from S or P. But of course, the energy range will allow 
to unambiguously determine cations such as K+ or Ca++. 

We have just started X-ray commissioning of the beamline optics and will expect 
first users from autumn 2013.

Best regards,

 Armin


-Original Message-
From: CCP4 bulletin board [mailto:CCP4BB@JISCMAIL.AC.UK] On Behalf Of Jrh
Sent: 31 October 2012 08:20
To: ccp4bb
Subject: Re: [ccp4bb] Ca or Zn

Dear Ethan,
Yes indeed. 
An exciting development underway at Diamond Light Source led by Armin Wagner 
will greatly improve the ease of measurement at eg the calcium edge but also 
extending that wavelength range. The  furin paper I quoted  did nevertheless 
successfully show structural detail from those measurements.

The Ga and Zn edges wavelengths are an easier challenge to access, i agree of 
course, rather the interest I tried to stress was the getting of the sigmas on 
the occupancies as well as the occupancies themselves, and how we did that and 
checked them via more than one software is also hopefully of interest.

Greetings from Taipei,
John 

Prof John R Helliwell DSc 
 
 

On 31 Oct 2012, at 10:37, Ethan Merritt merr...@u.washington.edu wrote:

 On Tuesday, 30 October 2012, Jrh wrote:
 This paper describes use of data either side of the calcium edge:-
 
 http://dx.doi.org/10.1107/S0907444905002556
 
 I think that counts as not amenable (which is not quite the same as 
 impossible.  From the Methods section of that paper:
 
  Measurements in the vicinity of the K absorption edge of  calcium 
 (3.07 Å) are close to or beyond the physical limit  of most beamlines 
 typically used for X-ray crystallography  [...] It was not possible to 
 observe interpretable  diffraction patterns at λ = 3 Å with the 
 weakly diffracting  furin crystals using the MAR CCD detector and 
 exposure  times up to 20 min per degree of rotation.
 
 They did soldier on and managed to collect extremely weak data below 
 the Ca edge and stronger but still very weak data above the edge where 
 the Ca f term was appreciable.  But this is far from a routine 
 experiment.
 
 Another approach dating back to work in 1972 by Peter Coleman and 
 Brian Matthews http://dx.doi.org/10.1016/0006-291X(72)90750-4
 is to replace the Ca with a rare earth having similar chemistry (e.g. 
 La, whose L-1 edge is at 1.98Å).
 
 
 This next paper describes a case of gallium and zinc mix at one site 
 with occupancy AND sigmas estimated with different software.
 This example is however much better diffraction resolution than that 
 you may have. But hopefully will still be of interest:-
 http://dx.doi.org/10.1107/S0108768110011237
 
 Ga and Zn, sure.  That's an easy one. 
 The Ga edge is at 1.196Å and the Zn edge is at 1.284Å, both edges are 
 nicely in range for data collection and they are close enough together 
 that little or no beamline readjustment is needed when jumping from 
 one to the other.
 
Ethan
 
 
 
 
 Prof John R Helliwell DSc
 
 
 
 On 31 Oct 2012, at 04:53, Ethan Merritt merr...@u.washington.edu wrote:
 
 On Tuesday, October 30, 2012 01:44:43 pm Adrian Goldman wrote:
 
 The coordination is indicative but not conclusive but, as I 
 responded to the original poster, I think the best approach is to 
 use anomalous scattering.  You can measure just below and above the 
 Ca edge,
 
 Actually, you can't.  The Ca K-edge is at 3.07Å, which is not a 
 wavelength amenable to macromolecular data collection.
 
   cheers,
 
   Ethan
 
 
 and similarly with the Zn, and those maps will be _highly_ indicative of 
 the relative amounts of metal ion present.  In fact, you can deconvolute 
 so that you know the occupancy of the metals at the various sites.
 
 Adrian
 
 
 On 30 Oct 2012, at 22:37, Chittaranjan Das wrote:
 
 Veerendra,
 
 You can rule out if zinc has replaced calcium ions (although I agree with 
 Nat and others that looking at the coordination sphere should give a big 
 clue) by taking a few crystals, washing them a couple of times and 
 subjecting them to ICP-MS analysis, if you have access to this technique. 
 You can learn how many zinc, if any, have bound per one protein molecule 
 in the dissolved crystal.
 
 Best
 Chitta
 
 
 
 - Original Message -
 From: Veerendra Kumar veerendra.ku...@uconn.edu
 To: CCP4BB@JISCMAIL.AC.UK
 Sent: Tuesday, October 30, 2012 2:55:33 PM
 Subject: [ccp4bb] Ca or Zn
 
 Dear CCP4bb users,
 
 I am working on a Ca2+ binding protein. it has 4-5 ca2+ binding sites.  I 
 purified the protein  in presence of Ca2+ and crystallized the Ca2+ bound 
 protein. I got crystal and solved the structure by SAD phasing at 2.1A 
 resolution. I can see the clear density in the difference map for metals

Re: [ccp4bb] Ca or Zn

2012-10-31 Thread Jrh
Dear Ethan,
Yes indeed. 
An exciting development underway at Diamond Light Source led by Armin Wagner 
will greatly improve the ease of measurement at eg the calcium edge but also 
extending that wavelength range. The  furin paper I quoted  did nevertheless 
successfully show structural detail from those measurements.

The Ga and Zn edges wavelengths are an easier challenge to access, i agree of 
course, rather the interest I tried to stress was the getting of the sigmas on 
the occupancies as well as the occupancies themselves, and how we did that and 
checked them via more than one software is also hopefully of interest.

Greetings from Taipei,
John 

Prof John R Helliwell DSc 
 
 

On 31 Oct 2012, at 10:37, Ethan Merritt merr...@u.washington.edu wrote:

 On Tuesday, 30 October 2012, Jrh wrote:
 This paper describes use of data either side of the calcium edge:-
 
 http://dx.doi.org/10.1107/S0907444905002556
 
 I think that counts as not amenable (which is not quite the same
 as impossible.  From the Methods section of that paper:
 
  Measurements in the vicinity of the K absorption edge of
  calcium (3.07 Å) are close to or beyond the physical limit
  of most beamlines typically used for X-ray crystallography
  [...] It was not possible to observe interpretable
  diffraction patterns at λ = 3 Å with the weakly diffracting
  furin crystals using the MAR CCD detector and exposure
  times up to 20 min per degree of rotation.
 
 They did soldier on and managed to collect extremely weak data
 below the Ca edge and stronger but still very weak data above the
 edge where the Ca f term was appreciable.  But this is far from a
 routine experiment.
 
 Another approach dating back to work in 1972 by Peter Coleman
 and Brian Matthews http://dx.doi.org/10.1016/0006-291X(72)90750-4
 is to replace the Ca with a rare earth having similar chemistry 
 (e.g. La, whose L-1 edge is at 1.98Å).  
 
 
 This next paper describes a case of gallium and zinc mix at 
 one site with occupancy AND sigmas estimated with different software. 
 This example is however much better diffraction resolution than 
 that you may have. But hopefully will still be of interest:-
 http://dx.doi.org/10.1107/S0108768110011237
 
 Ga and Zn, sure.  That's an easy one. 
 The Ga edge is at 1.196Å and the Zn edge is at 1.284Å,
 both edges are nicely in range for data collection and they are
 close enough together that little or no beamline readjustment
 is needed when jumping from one to the other.
 
Ethan
 
 
 
 
 Prof John R Helliwell DSc
 
 
 
 On 31 Oct 2012, at 04:53, Ethan Merritt merr...@u.washington.edu wrote:
 
 On Tuesday, October 30, 2012 01:44:43 pm Adrian Goldman wrote:
 
 The coordination is indicative but not conclusive but, as I responded to 
 the original poster, I think the best approach is to use anomalous 
 scattering.  You can measure just below and above the Ca edge, 
 
 Actually, you can't.  The Ca K-edge is at 3.07Å, which is not a wavelength
 amenable to macromolecular data collection.  
 
   cheers,
 
   Ethan
 
 
 and similarly with the Zn, and those maps will be _highly_ indicative of 
 the relative amounts of metal ion present.  In fact, you can deconvolute 
 so that you know the occupancy of the metals at the various sites.
 
 Adrian
 
 
 On 30 Oct 2012, at 22:37, Chittaranjan Das wrote:
 
 Veerendra,
 
 You can rule out if zinc has replaced calcium ions (although I agree with 
 Nat and others that looking at the coordination sphere should give a big 
 clue) by taking a few crystals, washing them a couple of times and 
 subjecting them to ICP-MS analysis, if you have access to this technique. 
 You can learn how many zinc, if any, have bound per one protein molecule 
 in the dissolved crystal.
 
 Best
 Chitta
 
 
 
 - Original Message -
 From: Veerendra Kumar veerendra.ku...@uconn.edu
 To: CCP4BB@JISCMAIL.AC.UK
 Sent: Tuesday, October 30, 2012 2:55:33 PM
 Subject: [ccp4bb] Ca or Zn
 
 Dear CCP4bb users,
 
 I am working on a Ca2+ binding protein. it has 4-5 ca2+ binding sites.  I 
 purified the protein  in presence of Ca2+ and crystallized the Ca2+ bound 
 protein. I got crystal and solved the structure by SAD phasing at 2.1A 
 resolution. I can see the clear density in the difference map for metals 
 at the expected binding sites. However I had ZnCl2 in the crystallization 
 conditions. Now i am not sure whether the observed density is for Ca or 
 Zn or how many of them are ca or  zn? Since Ca (20 elctron) and Zn (30 
 electron), is this value difference can be used to make a guess about 
 different ions? 
 is there any way we can find the electron density value at different 
 peaks? 
 
 Thank you
 
 Veerendra 
 
 
 
 


Re: [ccp4bb] Ca or Zn

2012-10-30 Thread Bosch, Juergen
calculate an anomalous map, you should see the Zn signal even if you collected 
at the SeMet peak.
Jürgen
..
Jürgen Bosch
Johns Hopkins University
Bloomberg School of Public Health
Department of Biochemistry  Molecular Biology
Johns Hopkins Malaria Research Institute
615 North Wolfe Street, W8708
Baltimore, MD 21205
Office: +1-410-614-4742
Lab:  +1-410-614-4894
Fax:  +1-410-955-2926
http://lupo.jhsph.edu



On Oct 30, 2012, at 2:55 PM, Kumar, Veerendra wrote:

Dear CCP4bb users,

I am working on a Ca2+ binding protein. it has 4-5 ca2+ binding sites.  I 
purified the protein  in presence of Ca2+ and crystallized the Ca2+ bound 
protein. I got crystal and solved the structure by SAD phasing at 2.1A 
resolution. I can see the clear density in the difference map for metals at the 
expected binding sites. However I had ZnCl2 in the crystallization conditions. 
Now i am not sure whether the observed density is for Ca or Zn or how many of 
them are ca or  zn? Since Ca (20 elctron) and Zn (30 electron), is this value 
difference can be used to make a guess about different ions?
is there any way we can find the electron density value at different peaks?

Thank you

Veerendra




Re: [ccp4bb] Ca or Zn

2012-10-30 Thread Jim Pflugrath
How would you distinguish between a mixture of Ca and Zn in the same locations?


From: CCP4 bulletin board [CCP4BB@JISCMAIL.AC.UK] on behalf of Kumar, Veerendra 
[veerendra.ku...@uconn.edu]
Sent: Tuesday, October 30, 2012 1:55 PM
To: CCP4BB@JISCMAIL.AC.UK
Subject: [ccp4bb] Ca or Zn

Dear CCP4bb users,

I am working on a Ca2+ binding protein. it has 4-5 ca2+ binding sites.  I 
purified the protein  in presence of Ca2+ and crystallized the Ca2+ bound 
protein. I got crystal and solved the structure by SAD phasing at 2.1A 
resolution. I can see the clear density in the difference map for metals at the 
expected binding sites. However I had ZnCl2 in the crystallization conditions. 
Now i am not sure whether the observed density is for Ca or Zn or how many of 
them are ca or  zn? Since Ca (20 elctron) and Zn (30 electron), is this value 
difference can be used to make a guess about different ions?
is there any way we can find the electron density value at different peaks?

Thank you

Veerendra


Re: [ccp4bb] Ca or Zn

2012-10-30 Thread Nat Echols
On Tue, Oct 30, 2012 at 12:12 PM, Jim Pflugrath
jim.pflugr...@rigaku.com wrote:
 How would you distinguish between a mixture of Ca and Zn in the same 
 locations?

How often would they be likely to bind in the same place?  Some of the
other transition metals are difficult to tell apart, but Ca and Zn
have very different coordination preferences.

-Nat


Re: [ccp4bb] Ca or Zn

2012-10-30 Thread David Schuller

On 10/30/12 15:02, Bosch, Juergen wrote:
calculate an anomalous map, you should see the Zn signal even if you 
collected at the SeMet peak.

Jürgen
..


Ca can have a noticeable anomalous signal of its own, if your data are good.

If the possibility to collect new data exists, collecting above and 
below the Zn edge should be informative. (Zn K edge 1.28 A = 9.6586 keV)


--
===
All Things Serve the Beam
===
   David J. Schuller
   modern man in a post-modern world
   MacCHESS, Cornell University
   schul...@cornell.edu



Re: [ccp4bb] Ca or Zn

2012-10-30 Thread Bosch, Juergen
If you calculate an edgeplot via Ethan's server:
http://skuld.bmsc.washington.edu/scatter/AS_form.html
you'll see that Zn @1Å has about 3 anomalous electrons whereas Ca less than 1, 
so assuming occupancy of 1 the stronger anomalous signal should give you a 
hint, second looking at the refined B-values if you placed the wrong metal 
should also coincide with your correct choice of metal.

The question with the mixture, well lets wave some hands and bring in 
biological relevance. If the protein is supposed to do something with Ca then I 
would say so, even if Zn is bound. If the anomalous signal is comparable to the 
SeMet sites then I would model Zn there but say it's likely not physiological 
as Ca is required by this enzyme.

Jürgen

On Oct 30, 2012, at 3:12 PM, Jim Pflugrath wrote:

 How would you distinguish between a mixture of Ca and Zn in the same 
 locations?
 
 
 From: CCP4 bulletin board [CCP4BB@JISCMAIL.AC.UK] on behalf of Kumar, 
 Veerendra [veerendra.ku...@uconn.edu]
 Sent: Tuesday, October 30, 2012 1:55 PM
 To: CCP4BB@JISCMAIL.AC.UK
 Subject: [ccp4bb] Ca or Zn
 
 Dear CCP4bb users,
 
 I am working on a Ca2+ binding protein. it has 4-5 ca2+ binding sites.  I 
 purified the protein  in presence of Ca2+ and crystallized the Ca2+ bound 
 protein. I got crystal and solved the structure by SAD phasing at 2.1A 
 resolution. I can see the clear density in the difference map for metals at 
 the expected binding sites. However I had ZnCl2 in the crystallization 
 conditions. Now i am not sure whether the observed density is for Ca or Zn or 
 how many of them are ca or  zn? Since Ca (20 elctron) and Zn (30 electron), 
 is this value difference can be used to make a guess about different ions?
 is there any way we can find the electron density value at different peaks?
 
 Thank you
 
 Veerendra

..
Jürgen Bosch
Johns Hopkins University
Bloomberg School of Public Health
Department of Biochemistry  Molecular Biology
Johns Hopkins Malaria Research Institute
615 North Wolfe Street, W8708
Baltimore, MD 21205
Office: +1-410-614-4742
Lab:  +1-410-614-4894
Fax:  +1-410-955-2926
http://lupo.jhsph.edu


Re: [ccp4bb] Ca or Zn

2012-10-30 Thread James Holton
Do occupancy refinement.  Especially if you use the new anomalous 
refinement option available in refmac, phenix.refine and other packages 
as well, you should be able to get a pretty reliable number for how many 
real electrons as well as anomalous electrons are at each site.  At 
almost any reasonable B-factor a Ca+2 looks a LOT like a Zn+2 at 64% 
occupancy, and the f will also be very different (depends on your 
wavelength).


If the results are not abundantly clear, then you can get a kinda-sorta 
error bar by starting refinement with different occupancies, and 
perhaps kicking the starting model coordinates around (to remove 
bias).  If the occupancy keeps refining back to the same value, then 
you can be reasonably confident that it is well-determined by your data.


If the occupancy doesn't appear to stably return to the same value, or 
at least values different enough to say it is Zn vs Ca, then the 
identity of your metal sites is NOT well-determined by your data.


-James Holton
MAD Scientist

On 10/30/2012 7:55 PM, Kumar, Veerendra wrote:

Dear CCP4bb users,

I am working on a Ca2+ binding protein. it has 4-5 ca2+ binding sites.  I 
purified the protein  in presence of Ca2+ and crystallized the Ca2+ bound 
protein. I got crystal and solved the structure by SAD phasing at 2.1A 
resolution. I can see the clear density in the difference map for metals at the 
expected binding sites. However I had ZnCl2 in the crystallization conditions. 
Now i am not sure whether the observed density is for Ca or Zn or how many of 
them are ca or  zn? Since Ca (20 elctron) and Zn (30 electron), is this value 
difference can be used to make a guess about different ions?
is there any way we can find the electron density value at different peaks?

Thank you

Veerendra


Re: [ccp4bb] Ca or Zn

2012-10-30 Thread Chittaranjan Das
Veerendra,
 
You can rule out if zinc has replaced calcium ions (although I agree with Nat 
and others that looking at the coordination sphere should give a big clue) by 
taking a few crystals, washing them a couple of times and subjecting them to 
ICP-MS analysis, if you have access to this technique. You can learn how many 
zinc, if any, have bound per one protein molecule in the dissolved crystal.

Best
Chitta

 

- Original Message -
From: Veerendra Kumar veerendra.ku...@uconn.edu
To: CCP4BB@JISCMAIL.AC.UK
Sent: Tuesday, October 30, 2012 2:55:33 PM
Subject: [ccp4bb] Ca or Zn

Dear CCP4bb users,

I am working on a Ca2+ binding protein. it has 4-5 ca2+ binding sites.  I 
purified the protein  in presence of Ca2+ and crystallized the Ca2+ bound 
protein. I got crystal and solved the structure by SAD phasing at 2.1A 
resolution. I can see the clear density in the difference map for metals at the 
expected binding sites. However I had ZnCl2 in the crystallization conditions. 
Now i am not sure whether the observed density is for Ca or Zn or how many of 
them are ca or  zn? Since Ca (20 elctron) and Zn (30 electron), is this value 
difference can be used to make a guess about different ions? 
is there any way we can find the electron density value at different peaks? 

Thank you

Veerendra 


Re: [ccp4bb] Ca or Zn

2012-10-30 Thread Adrian Goldman
This doesn't really give a useful answer.  It tells you the overall 
composition, but there is no means of knowing whether the metal ions are 
equally present at all sites: some sites can favour Zn over Ca and vice versa.  
Flame spectroscopy also works but has the same issue - and both have the 
problem that you don't really know whether you have  got rid of all metal ions 
bound weakly to the surface of the protein/crystal and most crystals, being 
small, have a humongous surface-to-volume ratio.

The coordination is indicative but not conclusive but, as I responded to the 
original poster, I think the best approach is to use anomalous scattering.  You 
can measure just below and above the Ca edge, and similarly with the Zn, and 
those maps will be _highly_ indicative of the relative amounts of metal ion 
present.  In fact, you can deconvolute so that you know the occupancy of the 
metals at the various sites.

Adrian


On 30 Oct 2012, at 22:37, Chittaranjan Das wrote:

 Veerendra,
 
 You can rule out if zinc has replaced calcium ions (although I agree with Nat 
 and others that looking at the coordination sphere should give a big clue) by 
 taking a few crystals, washing them a couple of times and subjecting them to 
 ICP-MS analysis, if you have access to this technique. You can learn how many 
 zinc, if any, have bound per one protein molecule in the dissolved crystal.
 
 Best
 Chitta
 
 
 
 - Original Message -
 From: Veerendra Kumar veerendra.ku...@uconn.edu
 To: CCP4BB@JISCMAIL.AC.UK
 Sent: Tuesday, October 30, 2012 2:55:33 PM
 Subject: [ccp4bb] Ca or Zn
 
 Dear CCP4bb users,
 
 I am working on a Ca2+ binding protein. it has 4-5 ca2+ binding sites.  I 
 purified the protein  in presence of Ca2+ and crystallized the Ca2+ bound 
 protein. I got crystal and solved the structure by SAD phasing at 2.1A 
 resolution. I can see the clear density in the difference map for metals at 
 the expected binding sites. However I had ZnCl2 in the crystallization 
 conditions. Now i am not sure whether the observed density is for Ca or Zn or 
 how many of them are ca or  zn? Since Ca (20 elctron) and Zn (30 electron), 
 is this value difference can be used to make a guess about different ions? 
 is there any way we can find the electron density value at different peaks? 
 
 Thank you
 
 Veerendra 


Re: [ccp4bb] Ca or Zn

2012-10-30 Thread Ethan Merritt
On Tuesday, October 30, 2012 01:44:43 pm Adrian Goldman wrote:

 The coordination is indicative but not conclusive but, as I responded to the 
 original poster, I think the best approach is to use anomalous scattering.  
 You can measure just below and above the Ca edge, 

Actually, you can't.  The Ca K-edge is at 3.07Å, which is not a wavelength
amenable to macromolecular data collection.  

cheers,

Ethan


 and similarly with the Zn, and those maps will be _highly_ indicative of the 
 relative amounts of metal ion present.  In fact, you can deconvolute so that 
 you know the occupancy of the metals at the various sites.
 
 Adrian
 
 
 On 30 Oct 2012, at 22:37, Chittaranjan Das wrote:
 
  Veerendra,
  
  You can rule out if zinc has replaced calcium ions (although I agree with 
  Nat and others that looking at the coordination sphere should give a big 
  clue) by taking a few crystals, washing them a couple of times and 
  subjecting them to ICP-MS analysis, if you have access to this technique. 
  You can learn how many zinc, if any, have bound per one protein molecule in 
  the dissolved crystal.
  
  Best
  Chitta
  
  
  
  - Original Message -
  From: Veerendra Kumar veerendra.ku...@uconn.edu
  To: CCP4BB@JISCMAIL.AC.UK
  Sent: Tuesday, October 30, 2012 2:55:33 PM
  Subject: [ccp4bb] Ca or Zn
  
  Dear CCP4bb users,
  
  I am working on a Ca2+ binding protein. it has 4-5 ca2+ binding sites.  I 
  purified the protein  in presence of Ca2+ and crystallized the Ca2+ bound 
  protein. I got crystal and solved the structure by SAD phasing at 2.1A 
  resolution. I can see the clear density in the difference map for metals at 
  the expected binding sites. However I had ZnCl2 in the crystallization 
  conditions. Now i am not sure whether the observed density is for Ca or Zn 
  or how many of them are ca or  zn? Since Ca (20 elctron) and Zn (30 
  electron), is this value difference can be used to make a guess about 
  different ions? 
  is there any way we can find the electron density value at different peaks? 
  
  Thank you
  
  Veerendra 
 

-- 
Ethan A Merritt
Biomolecular Structure Center,  K-428 Health Sciences Bldg
University of Washington, Seattle 98195-7742


Re: [ccp4bb] Ca or Zn

2012-10-30 Thread Ethan Merritt
On Tuesday, 30 October 2012, Jrh wrote:
 This paper describes use of data either side of the calcium edge:-
 
 http://dx.doi.org/10.1107/S0907444905002556

I think that counts as not amenable (which is not quite the same
as impossible.  From the Methods section of that paper:

  Measurements in the vicinity of the K absorption edge of
  calcium (3.07 Å) are close to or beyond the physical limit
  of most beamlines typically used for X-ray crystallography
  [...] It was not possible to observe interpretable
  diffraction patterns at λ = 3 Å with the weakly diffracting
  furin crystals using the MAR CCD detector and exposure
  times up to 20 min per degree of rotation.

They did soldier on and managed to collect extremely weak data
below the Ca edge and stronger but still very weak data above the
edge where the Ca f term was appreciable.  But this is far from a
routine experiment.

Another approach dating back to work in 1972 by Peter Coleman
and Brian Matthews http://dx.doi.org/10.1016/0006-291X(72)90750-4
is to replace the Ca with a rare earth having similar chemistry 
(e.g. La, whose L-1 edge is at 1.98Å).  


 This next paper describes a case of gallium and zinc mix at 
 one site with occupancy AND sigmas estimated with different software. 
 This example is however much better diffraction resolution than 
 that you may have. But hopefully will still be of interest:-
 http://dx.doi.org/10.1107/S0108768110011237

Ga and Zn, sure.  That's an easy one. 
The Ga edge is at 1.196Å and the Zn edge is at 1.284Å,
both edges are nicely in range for data collection and they are
close enough together that little or no beamline readjustment
is needed when jumping from one to the other.

Ethan



 
 Prof John R Helliwell DSc
  
  
 
 On 31 Oct 2012, at 04:53, Ethan Merritt merr...@u.washington.edu wrote:
 
  On Tuesday, October 30, 2012 01:44:43 pm Adrian Goldman wrote:
  
  The coordination is indicative but not conclusive but, as I responded to 
  the original poster, I think the best approach is to use anomalous 
  scattering.  You can measure just below and above the Ca edge, 
  
  Actually, you can't.  The Ca K-edge is at 3.07Å, which is not a wavelength
  amenable to macromolecular data collection.  
  
 cheers,
  
 Ethan
  
  
  and similarly with the Zn, and those maps will be _highly_ indicative of 
  the relative amounts of metal ion present.  In fact, you can deconvolute 
  so that you know the occupancy of the metals at the various sites.
  
  Adrian
  
  
  On 30 Oct 2012, at 22:37, Chittaranjan Das wrote:
  
  Veerendra,
  
  You can rule out if zinc has replaced calcium ions (although I agree with 
  Nat and others that looking at the coordination sphere should give a big 
  clue) by taking a few crystals, washing them a couple of times and 
  subjecting them to ICP-MS analysis, if you have access to this technique. 
  You can learn how many zinc, if any, have bound per one protein molecule 
  in the dissolved crystal.
  
  Best
  Chitta
  
  
  
  - Original Message -
  From: Veerendra Kumar veerendra.ku...@uconn.edu
  To: CCP4BB@JISCMAIL.AC.UK
  Sent: Tuesday, October 30, 2012 2:55:33 PM
  Subject: [ccp4bb] Ca or Zn
  
  Dear CCP4bb users,
  
  I am working on a Ca2+ binding protein. it has 4-5 ca2+ binding sites.  I 
  purified the protein  in presence of Ca2+ and crystallized the Ca2+ bound 
  protein. I got crystal and solved the structure by SAD phasing at 2.1A 
  resolution. I can see the clear density in the difference map for metals 
  at the expected binding sites. However I had ZnCl2 in the crystallization 
  conditions. Now i am not sure whether the observed density is for Ca or 
  Zn or how many of them are ca or  zn? Since Ca (20 elctron) and Zn (30 
  electron), is this value difference can be used to make a guess about 
  different ions? 
  is there any way we can find the electron density value at different 
  peaks? 
  
  Thank you
  
  Veerendra