Re: [ccp4bb] off topic: good peak on gel filtration

2013-07-27 Thread Edward A. Berry

OK, lets see-
25000 plates/m = 7500 plates in a 30 cm column, so variance in elution volume
of the molecules should be elution volume/sqrt(7500) = .0115 Ve.

Full width at half height is 2*sqrt(2ln2), or 2.35, times sigma, so 0.0275 
times elution volume.
For elution volumes 10 - 15 ml, FWHH of peak should be 0.27 to 0.4 ml,
if the sample is applied in an infinitely small volume.
For finite samples I guess width of the peak would be the sample volume
increased by .27 to .4 ml due to peak broadening.
eab

Zhijie Li wrote:

Hi Ed,

I guess by 24mL SD200 Peter meant the Superdex 200 10/300 column, which
most of us should be quite familiar with. According to GE healthcare, a new
Superdex 200 10/300 GL column should have TP 25000/m. For comparison, a new
Superdex200 16/600 PG, which uses bigger beads, has TP 13000/m. The TP
difference of the two should be mainly caused by the different resin sizes.
Of course in reality columns change over time and in cases like Peter's, it
might be a good idea to test the performance of the column before drawing a
conclusion. When we are concerned about resolution of a column, we load a
standard sample and calculate the TP based on the peak shape. As I remember,
GE healthcare's SEC manuals has recommended procedures on TP determination.

Zhijie


-Original Message-
From: Edward A. Berry
Sent: Saturday, June 29, 2013 7:43 PM
To: CCP4BB@JISCMAIL.AC.UK
Subject: Re: [ccp4bb] off topic: good peak on gel filtration

Peter Hsu wrote:

Hi all,

I've generally always thought as long as the peak was symmetrical and not
too broad would suggest a good sample. However, looking at my previous
runs in the past, I've had peaks as narrow as 1.5-2mL on a 24mL SD200, or
slightly broader peaks with about 3mL (all symmetrical peaks, roughly
similar amounts loaded on the columns). I'm curious to see what people's
views are as far as what constitutes a broad peak and how much that can
end up affecting crystallization of the sample.

Thanks for any responses.

Peter


The width itself may not be a good indicator unless its always the same
protein- in general a molecule that elutes later
will have a broader peak.
Supposing that each time a molecule diffuses into the stationary phase it
resides there for a certain time on the
average, then the extra retention time is proportional to that time, times
the number of times it enters stationary
phase (N, theoretical plates). The variance in elution time is
proportional to the square root of N (like standard
error of the mean) and the dwell time. This gives sigma/(retention time) =
1/sqrt(N). If N is the same for all
molecules, the criterion to look at is peak width divided by retention time.
If it varies (the reason some molecules
elute slower is not just that they stay in the stationary phase longer, but
also they enter more often; k-on as well as
k-off) that would still be better than just peak width.  People don't talk
about theoretical plats and HTEP much any
more, perhaps because the driving force in chromatography is HPLC and FPLC,
and fast chromatography is antithetical to
good resolution?

However I'm not familiar with this column and can't advise. You can
calculate N more exactly (see wikipedia van Deemter
equation) as 8*ln(2)*square of (elution volume over width at half height),
divide length of column by that to get HETP,
and compare with values like .7 mm reported for resins like ultragel A at
optimum (very slow) flow rate.

eab




Re: [ccp4bb] off topic: good peak on gel filtration

2013-07-02 Thread Gloria Borgstahl
In our lab we do DLS on the SEC fractions and only set up those that are
monodisperse  :-)


On Mon, Jul 1, 2013 at 7:37 PM, El Arnaout, Toufic 
elarnao...@biochem.wustl.edu wrote:

 Hello Peter,
 In addition to the great comments/details, please check the following
 points I have now in mind.. since you want to relate the size exclusion
 peak/profile to the crystallization:
 - occasionnaly, some perfect peaks that you might think are homogeneous
 actually correspond to a sample of hetergeneous protein (maybe the target
 protein will still crystallize, but problems happen during crystal
 optimization, or/and observing missing electronic density of the N- or
 C-terminus for example). It might also be reflected on the specific
 activity (btw if the protein you have is easy to assay, you could check the
 activity from different fractions if you think broad peak = problem). In
 some occasions, analyzing fractions from a perfect peak shows on SDS-PAGE
 a double band or sometimes far bands (I won't comment on oligomerization in
 this case).
 - not all proteins from good SE chromatograms crystallize...
 - some people only collect fractions from the centre of the peak (or for
 example they measure the A280 max, divide it by two, draw an horizontal
 line at that value, and collect the fractions/projection between where the
 line crosses the peak from both sides.. If you add one fraction before or
 after, the protein might not crystallize anymore.
 - from the literature, I have seen many shapes of chromatograms: perfect,
 bleeding, skewed (tail) to the right or left, little broad, etc.. which
 resulted in diffraction quality crystals.
 - re-running a protein sample (that crystallizes after the first SEC) for
 a second SEC might cause the protein not to crystallize anymore (for
 example membrane proteins might lose lipids etc).
 - for proteins that are subjected to SEC straight after Ni-NTA, and which
 have a perfect peak shape: a 4-16 hours delay before injection might show
 the aggregation effect. A peak shoulder will form, which you might not have
 seen if the sample was directly injected. The point is: maybe what you
 incubate for crystallization or work on after SEC might not be anymore that
 protein with the beautiful peak you had. Using different
 additives/chemicals/mutations may help in making the protein more
 stable/thermostable. You can check previous publications of Dr Tate CG,
 GPCRs for example. This is also a nice paper:
 https://www.ncbi.nlm.nih.gov/pmc/articles/PMC3111809.
 Regards



 toufic el arnaout
 School of Medicine - 660 S Euclid Ave
 Washington University in St. Louis
 St Louis, MO 63110, USA

 
 From: CCP4 bulletin board [CCP4BB@JISCMAIL.AC.UK] on behalf of Edward A.
 Berry [ber...@upstate.edu]
 Sent: Saturday, June 29, 2013 9:34 PM
 To: CCP4BB@JISCMAIL.AC.UK
 Subject: Re: [ccp4bb] off topic: good peak on gel filtration

 Thanks- guess I'm old-fashioned, using low-pressure columns.
 So apparently theoretical plates are still calculated, and have
 improved a lot- 25000/m is HETP .04 mm, way better than the
 figure I mentioned. (TP per dollar not so much.)
 No more sour grapes from me-
 eab

 Zhijie Li wrote:
  Hi Ed,
 
  I guess by 24mL SD200 Peter meant the Superdex 200 10/300 column, which
  most of us should be quite familiar with. According to GE healthcare, a
 new
  Superdex 200 10/300 GL column should have TP 25000/m. For comparison, a
 new
  Superdex200 16/600 PG, which uses bigger beads, has TP 13000/m. The TP
  difference of the two should be mainly caused by the different resin
 sizes.
  Of course in reality columns change over time and in cases like Peter's,
 it
  might be a good idea to test the performance of the column before
 drawing a
  conclusion. When we are concerned about resolution of a column, we load a
  standard sample and calculate the TP based on the peak shape. As I
 remember,
  GE healthcare's SEC manuals has recommended procedures on TP
 determination.
 
  Zhijie
 
 
  -Original Message-
  From: Edward A. Berry
  Sent: Saturday, June 29, 2013 7:43 PM
  To: CCP4BB@JISCMAIL.AC.UK
  Subject: Re: [ccp4bb] off topic: good peak on gel filtration
 
  Peter Hsu wrote:
  Hi all,
 
  I've generally always thought as long as the peak was symmetrical and
 not
  too broad would suggest a good sample. However, looking at my previous
  runs in the past, I've had peaks as narrow as 1.5-2mL on a 24mL SD200,
 or
  slightly broader peaks with about 3mL (all symmetrical peaks, roughly
  similar amounts loaded on the columns). I'm curious to see what people's
  views are as far as what constitutes a broad peak and how much that can
  end up affecting crystallization of the sample.
 
  Thanks for any responses.
 
  Peter
 
  The width itself may not be a good indicator unless its always the same
  protein- in general a molecule that elutes later
  will have a broader peak.
  Supposing that each time a molecule diffuses

Re: [ccp4bb] off topic: good peak on gel filtration

2013-07-01 Thread El Arnaout, Toufic
Hello Peter,
In addition to the great comments/details, please check the following points I 
have now in mind.. since you want to relate the size exclusion peak/profile to 
the crystallization:
- occasionnaly, some perfect peaks that you might think are homogeneous 
actually correspond to a sample of hetergeneous protein (maybe the target 
protein will still crystallize, but problems happen during crystal 
optimization, or/and observing missing electronic density of the N- or 
C-terminus for example). It might also be reflected on the specific activity 
(btw if the protein you have is easy to assay, you could check the activity 
from different fractions if you think broad peak = problem). In some occasions, 
analyzing fractions from a perfect peak shows on SDS-PAGE a double band or 
sometimes far bands (I won't comment on oligomerization in this case).
- not all proteins from good SE chromatograms crystallize...
- some people only collect fractions from the centre of the peak (or for 
example they measure the A280 max, divide it by two, draw an horizontal line at 
that value, and collect the fractions/projection between where the line crosses 
the peak from both sides.. If you add one fraction before or after, the protein 
might not crystallize anymore.
- from the literature, I have seen many shapes of chromatograms: perfect, 
bleeding, skewed (tail) to the right or left, little broad, etc.. which 
resulted in diffraction quality crystals.
- re-running a protein sample (that crystallizes after the first SEC) for a 
second SEC might cause the protein not to crystallize anymore (for example 
membrane proteins might lose lipids etc).
- for proteins that are subjected to SEC straight after Ni-NTA, and which have 
a perfect peak shape: a 4-16 hours delay before injection might show the 
aggregation effect. A peak shoulder will form, which you might not have seen if 
the sample was directly injected. The point is: maybe what you incubate for 
crystallization or work on after SEC might not be anymore that protein with the 
beautiful peak you had. Using different additives/chemicals/mutations may help 
in making the protein more stable/thermostable. You can check previous 
publications of Dr Tate CG, GPCRs for example. This is also a nice paper: 
https://www.ncbi.nlm.nih.gov/pmc/articles/PMC3111809.
Regards



toufic el arnaout
School of Medicine - 660 S Euclid Ave
Washington University in St. Louis
St Louis, MO 63110, USA


From: CCP4 bulletin board [CCP4BB@JISCMAIL.AC.UK] on behalf of Edward A. Berry 
[ber...@upstate.edu]
Sent: Saturday, June 29, 2013 9:34 PM
To: CCP4BB@JISCMAIL.AC.UK
Subject: Re: [ccp4bb] off topic: good peak on gel filtration

Thanks- guess I'm old-fashioned, using low-pressure columns.
So apparently theoretical plates are still calculated, and have
improved a lot- 25000/m is HETP .04 mm, way better than the
figure I mentioned. (TP per dollar not so much.)
No more sour grapes from me-
eab

Zhijie Li wrote:
 Hi Ed,

 I guess by 24mL SD200 Peter meant the Superdex 200 10/300 column, which
 most of us should be quite familiar with. According to GE healthcare, a new
 Superdex 200 10/300 GL column should have TP 25000/m. For comparison, a new
 Superdex200 16/600 PG, which uses bigger beads, has TP 13000/m. The TP
 difference of the two should be mainly caused by the different resin sizes.
 Of course in reality columns change over time and in cases like Peter's, it
 might be a good idea to test the performance of the column before drawing a
 conclusion. When we are concerned about resolution of a column, we load a
 standard sample and calculate the TP based on the peak shape. As I remember,
 GE healthcare's SEC manuals has recommended procedures on TP determination.

 Zhijie


 -Original Message-
 From: Edward A. Berry
 Sent: Saturday, June 29, 2013 7:43 PM
 To: CCP4BB@JISCMAIL.AC.UK
 Subject: Re: [ccp4bb] off topic: good peak on gel filtration

 Peter Hsu wrote:
 Hi all,

 I've generally always thought as long as the peak was symmetrical and not
 too broad would suggest a good sample. However, looking at my previous
 runs in the past, I've had peaks as narrow as 1.5-2mL on a 24mL SD200, or
 slightly broader peaks with about 3mL (all symmetrical peaks, roughly
 similar amounts loaded on the columns). I'm curious to see what people's
 views are as far as what constitutes a broad peak and how much that can
 end up affecting crystallization of the sample.

 Thanks for any responses.

 Peter

 The width itself may not be a good indicator unless its always the same
 protein- in general a molecule that elutes later
 will have a broader peak.
 Supposing that each time a molecule diffuses into the stationary phase it
 resides there for a certain time on the
 average, then the extra retention time is proportional to that time, times
 the number of times it enters stationary
 phase (N, theoretical plates). The variance in elution time is
 proportional

Re: [ccp4bb] off topic: good peak on gel filtration

2013-06-29 Thread Edward A. Berry

Peter Hsu wrote:

Hi all,

I've generally always thought as long as the peak was symmetrical and not too 
broad would suggest a good sample. However, looking at my previous runs in the 
past, I've had peaks as narrow as 1.5-2mL on a 24mL SD200, or slightly broader 
peaks with about 3mL (all symmetrical peaks, roughly similar amounts loaded on 
the columns). I'm curious to see what people's views are as far as what 
constitutes a broad peak and how much that can end up affecting crystallization 
of the sample.

Thanks for any responses.

Peter

The width itself may not be a good indicator unless its always the same protein- in general a molecule that elutes later 
will have a broader peak.
Supposing that each time a molecule diffuses into the stationary phase it resides there for a certain time on the 
average, then the extra retention time is proportional to that time, times the number of times it enters stationary 
phase (N, theoretical plates). The variance in elution time is proportional to the square root of N (like standard 
error of the mean) and the dwell time. This gives sigma/(retention time) = 1/sqrt(N). If N is the same for all 
molecules, the criterion to look at is peak width divided by retention time. If it varies (the reason some molecules 
elute slower is not just that they stay in the stationary phase longer, but also they enter more often; k-on as well as 
k-off) that would still be better than just peak width.  People don't talk about theoretical plats and HTEP much any 
more, perhaps because the driving force in chromatography is HPLC and FPLC, and fast chromatography is antithetical to 
good resolution?


However I'm not familiar with this column and can't advise. You can calculate N more exactly (see wikipedia van Deemter 
equation) as 8*ln(2)*square of (elution volume over width at half height), divide length of column by that to get HETP, 
and compare with values like .7 mm reported for resins like ultragel A at optimum (very slow) flow rate.


eab


Re: [ccp4bb] off topic: good peak on gel filtration

2013-06-29 Thread Zhijie Li

Hi Ed,

I guess by 24mL SD200 Peter meant the Superdex 200 10/300 column, which 
most of us should be quite familiar with. According to GE healthcare, a new 
Superdex 200 10/300 GL column should have TP 25000/m. For comparison, a new 
Superdex200 16/600 PG, which uses bigger beads, has TP 13000/m. The TP 
difference of the two should be mainly caused by the different resin sizes.
Of course in reality columns change over time and in cases like Peter's, it 
might be a good idea to test the performance of the column before drawing a 
conclusion. When we are concerned about resolution of a column, we load a 
standard sample and calculate the TP based on the peak shape. As I remember, 
GE healthcare's SEC manuals has recommended procedures on TP determination.


Zhijie


-Original Message- 
From: Edward A. Berry

Sent: Saturday, June 29, 2013 7:43 PM
To: CCP4BB@JISCMAIL.AC.UK
Subject: Re: [ccp4bb] off topic: good peak on gel filtration

Peter Hsu wrote:

Hi all,

I've generally always thought as long as the peak was symmetrical and not 
too broad would suggest a good sample. However, looking at my previous 
runs in the past, I've had peaks as narrow as 1.5-2mL on a 24mL SD200, or 
slightly broader peaks with about 3mL (all symmetrical peaks, roughly 
similar amounts loaded on the columns). I'm curious to see what people's 
views are as far as what constitutes a broad peak and how much that can 
end up affecting crystallization of the sample.


Thanks for any responses.

Peter

The width itself may not be a good indicator unless its always the same 
protein- in general a molecule that elutes later

will have a broader peak.
Supposing that each time a molecule diffuses into the stationary phase it 
resides there for a certain time on the
average, then the extra retention time is proportional to that time, times 
the number of times it enters stationary
phase (N, theoretical plates). The variance in elution time is 
proportional to the square root of N (like standard
error of the mean) and the dwell time. This gives sigma/(retention time) = 
1/sqrt(N). If N is the same for all
molecules, the criterion to look at is peak width divided by retention time. 
If it varies (the reason some molecules
elute slower is not just that they stay in the stationary phase longer, but 
also they enter more often; k-on as well as
k-off) that would still be better than just peak width.  People don't talk 
about theoretical plats and HTEP much any
more, perhaps because the driving force in chromatography is HPLC and FPLC, 
and fast chromatography is antithetical to

good resolution?

However I'm not familiar with this column and can't advise. You can 
calculate N more exactly (see wikipedia van Deemter
equation) as 8*ln(2)*square of (elution volume over width at half height), 
divide length of column by that to get HETP,
and compare with values like .7 mm reported for resins like ultragel A at 
optimum (very slow) flow rate.


eab 


Re: [ccp4bb] off topic: good peak on gel filtration

2013-06-29 Thread Michael Thompson
Hi Peter,

I won't claim to be an expert, however, there is a small textbook by Robert K. 
Scopes called Protein Purification: Principles and Practice, that has very 
good practical and theoretical information about column chromatography (among 
other methods). I have found this book useful a number of times when I have had 
similar questions.

Good luck,

Mike







- Original Message -
From: Peter Hsu hsuu...@u.washington.edu
To: CCP4BB@JISCMAIL.AC.UK
Sent: Saturday, June 29, 2013 2:01:15 PM GMT -08:00 US/Canada Pacific
Subject: [ccp4bb] off topic: good peak on gel filtration

Hi all,

I've generally always thought as long as the peak was symmetrical and not too 
broad would suggest a good sample. However, looking at my previous runs in the 
past, I've had peaks as narrow as 1.5-2mL on a 24mL SD200, or slightly broader 
peaks with about 3mL (all symmetrical peaks, roughly similar amounts loaded on 
the columns). I'm curious to see what people's views are as far as what 
constitutes a broad peak and how much that can end up affecting crystallization 
of the sample. 

Thanks for any responses.

Peter

-- 
Michael C. Thompson

Graduate Student

Biochemistry  Molecular Biology Division

Department of Chemistry  Biochemistry

University of California, Los Angeles

mi...@chem.ucla.edu


Re: [ccp4bb] off topic: good peak on gel filtration

2013-06-29 Thread Edward A. Berry

Thanks- guess I'm old-fashioned, using low-pressure columns.
So apparently theoretical plates are still calculated, and have
improved a lot- 25000/m is HETP .04 mm, way better than the
figure I mentioned. (TP per dollar not so much.)
No more sour grapes from me-
eab

Zhijie Li wrote:

Hi Ed,

I guess by 24mL SD200 Peter meant the Superdex 200 10/300 column, which
most of us should be quite familiar with. According to GE healthcare, a new
Superdex 200 10/300 GL column should have TP 25000/m. For comparison, a new
Superdex200 16/600 PG, which uses bigger beads, has TP 13000/m. The TP
difference of the two should be mainly caused by the different resin sizes.
Of course in reality columns change over time and in cases like Peter's, it
might be a good idea to test the performance of the column before drawing a
conclusion. When we are concerned about resolution of a column, we load a
standard sample and calculate the TP based on the peak shape. As I remember,
GE healthcare's SEC manuals has recommended procedures on TP determination.

Zhijie


-Original Message-
From: Edward A. Berry
Sent: Saturday, June 29, 2013 7:43 PM
To: CCP4BB@JISCMAIL.AC.UK
Subject: Re: [ccp4bb] off topic: good peak on gel filtration

Peter Hsu wrote:

Hi all,

I've generally always thought as long as the peak was symmetrical and not
too broad would suggest a good sample. However, looking at my previous
runs in the past, I've had peaks as narrow as 1.5-2mL on a 24mL SD200, or
slightly broader peaks with about 3mL (all symmetrical peaks, roughly
similar amounts loaded on the columns). I'm curious to see what people's
views are as far as what constitutes a broad peak and how much that can
end up affecting crystallization of the sample.

Thanks for any responses.

Peter


The width itself may not be a good indicator unless its always the same
protein- in general a molecule that elutes later
will have a broader peak.
Supposing that each time a molecule diffuses into the stationary phase it
resides there for a certain time on the
average, then the extra retention time is proportional to that time, times
the number of times it enters stationary
phase (N, theoretical plates). The variance in elution time is
proportional to the square root of N (like standard
error of the mean) and the dwell time. This gives sigma/(retention time) =
1/sqrt(N). If N is the same for all
molecules, the criterion to look at is peak width divided by retention time.
If it varies (the reason some molecules
elute slower is not just that they stay in the stationary phase longer, but
also they enter more often; k-on as well as
k-off) that would still be better than just peak width.  People don't talk
about theoretical plats and HTEP much any
more, perhaps because the driving force in chromatography is HPLC and FPLC,
and fast chromatography is antithetical to
good resolution?

However I'm not familiar with this column and can't advise. You can
calculate N more exactly (see wikipedia van Deemter
equation) as 8*ln(2)*square of (elution volume over width at half height),
divide length of column by that to get HETP,
and compare with values like .7 mm reported for resins like ultragel A at
optimum (very slow) flow rate.

eab