Re: UL544 Leakage Limits Patient Equipment*

2000-03-29 Thread Robert Johnson

It may not be important whether you call it an applied part or not. You are not
making a direct electrical connection to the patient. In any case you should
meet the leakage requirement. This should be easy with properly insulated
earphones.
IEC 60950 also addresses this issue and throws in more considerations. For the
condition of telephone type headsets they are worried about transients from
telecom or power lines, and therefore have additional isolation requirements.
Clause 6.2.2 requires an impulse test of 2.5 KV or hipot of 1.5 KV (in Australia
this is increased to both 7KV impulse and 3KV hipot) applied to foil wrapped
around the headset.
Bob Johnson

Kenneth McCormick wrote:

 I also agree that the headphones are a patient connection.

 Hi,

 Its been a while since I looked at UL 544, but a few comments.

 1.  Headphones are an applied part (UL2601-1) or patient connection
 (UL544).  They come into direct contact with the patient.  You can argue
 that they are an ordinary patient connection (Clause 2.18 of UL544).  But,
 it is still a patient connection.  The limit is 50 uA for ordinary patient
 connection.

 2.  How long is this product going to be in the market?  UL 544 goes
 away on 1/1/2003 for new products and 1/1/2005 for all products.  You might
 be better off going to UL2601-1 now.  In UL2601-1 this type of patient
 connect is Type BF.  The patient leakage current limit is 100 uA in normal
 condition and 500 uA in single fault condition.

 3.  Call UL.  Maybe I missed some out in UL544.

 Ned Devine
 Entela, Inc.
 Program Manager III
 Phone 616 248 9671
 Fax  616 574 9752
 e-mail  ndev...@entela.com

 -Original Message-
 From: me...@aol.com [mailto:me...@aol.com]
 Sent: Friday, March 24, 2000 3:36 PM
 To: jjuh...@fiberoptions.com; m.r...@ieee.org;
 emc-p...@majordomo.ieee.org
 Subject: UL544 Leakage Limits Patient Equipment*

 Most esteemed colleagues, we are looking for your professional opinion on a
 UL 544 leakage limit (I think you will find this interesting):

 1. This particular product uses a UL 544 evaluated direct plug in power
 supply with outputs to the patient care equipment. In this case it is a
 diagnostic unit that sends an audible tone to headphones (audiometer). The
 plug in power supply Conditions of Acceptability indicate the outputs are
 not evaluated for patient leads (i.e. applied parts).

 2. Table 42.1 of UL 544 specifies leakage limits. patient connection
 footnote a references testing of patient leads (applied parts)
 connections. There is no written definition for patient leads or applied
 parts in UL544.

 As such NFPA 99 supplements UL 544 as it draws from the NEC and NFPA 99
 (referenced in UL 544):

 NFPA 99 defines the US definition of  Patient Lead = A deliberate
 electrical connection that can carry current between an appliance and
 patient. It is not intended to include adventitious or casual contacts such
 as a push button, bed surface, lamp, hand held appliance, etc.

 3. As the headphones of this audiometer are clearly not deliberate
 electrical connections we conclude these are not patient leads (applied
 parts) which would not fall under the limits for patient connection
 limits
 per 544. The applicable limits would be as defined under enclosure or
 chassis grounded  or double insulated Now be careful not to jump to

 a conclusion yet. You might say enclosure or chassis?, but if you
 examine this, you will find the footnotes reference UL 544's Enclosure
 definition:

 Enclosure =  That external portion of an appliance that serves to house or
 support component parts, or both. Enclosure of patient care equipment likely

 to be contacted by a patient include, for example, bedside monitors, bed
 frames, dental chairs, and examination stands.

 Our conclusion: Due to the US definition of patient leads (applied parts),

 the earphones of an audiometer (patient care equipment) are subjected to the

 leakage current limits for enclosure or chassis, and not the limits of
 patient connection. For this particular application, we conclude that
 based
 on the C of As, the output of the power supply has already been evaluated
 for
 enclosure or chassis leakage limits.

 Your Thoughts???

 Drew

 PS: If you care to look, CSA supports this position in that 50uA is related
 to cardiac tissue limits only. See Appendix A of CSA 22.2 125 (500uA).

 ---
 This message is from the IEEE EMC Society Product Safety
 Technical Committee emc-pstc discussion list.

 To cancel your subscription, send mail to:
  majord...@ieee.org
 with the single line:
  unsubscribe emc-pstc

 For help, send mail to the list administrators:
  Jim Bacher:  jim_bac...@mail.monarch.com
  Michael Garretson:pstc_ad...@garretson.org

 For policy questions, send mail to:
  Richard Nute:   ri...@ieee.org

 ---
 This message is from the IEEE EMC 

RE: EMC Directive Revisions

2000-03-29 Thread WOODS

Ed, do you know what group produced this document?
Richard Woods

--
From:  Price, Ed [SMTP:ed.pr...@cubic.com]
Sent:  Wednesday, March 29, 2000 10:17 AM
To:  'emc-p...@ieee.org'
Subject:  EMC Directive Revisions


I found an interesting document a few days ago. This is a draft of
late 1999
revisions, actually amendments, to the EMC Directive. It looks like
this
will supplement, not necessarily replace, the existing Directive.
Since
nothing looked newer, and this is several months old, maybe this is
nearly a
final version. FWIW, no guarantee about the veracity of this
document!

I haven't read it in detail yet, but notice:

1. failure modes of which the user would not be aware

2. testing all manners of use

3.  the need for an analysis

The document is in the form of a 140K MS Word rtf. As I did last
year with
the 461E document, I'll return a copy of this EMC Directive
Amendments draft
by attachment to your email. (Email your request directly to me, not
the
list.)

Ed



:-):-):-):-):-):-):-):-):-):-):-):-):-):-):-):-):-):-):-):-):-):-):-):-):-)
Ed Price
ed.pr...@cubic.com
Electromagnetic Compatibility Lab
Cubic Defense Systems
San Diego, CA.  USA
858-505-2780 (Voice)
858-505-1583 (Fax)
Military  Avionics EMC Services Is Our Specialty
Shake-Bake-Shock - Metrology - Reliability Analysis

:-):-):-):-):-):-):-):-):-):-):-):-):-):-):-):-):-):-):-):-):-):-):-):-):-)


---
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

To cancel your subscription, send mail to:
 majord...@ieee.org
with the single line:
 unsubscribe emc-pstc

For help, send mail to the list administrators:
 Jim Bacher:  jim_bac...@mail.monarch.com
 Michael Garretson:pstc_ad...@garretson.org

For policy questions, send mail to:
 Richard Nute:   ri...@ieee.org


---
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

To cancel your subscription, send mail to:
 majord...@ieee.org
with the single line:
 unsubscribe emc-pstc

For help, send mail to the list administrators:
 Jim Bacher:  jim_bac...@mail.monarch.com
 Michael Garretson:pstc_ad...@garretson.org

For policy questions, send mail to:
 Richard Nute:   ri...@ieee.org



RE: Ambient Cancellation Device for OATS

2000-03-29 Thread Grasso, Charles (Chaz)

Thanks Mike for an excellent post.


-Original Message-
From: Mike Murphy [mailto:mmur...@alesis.com]
Sent: Wednesday, March 29, 2000 10:56 AM
To: emc-p...@majordomo.ieee.org
Subject: RE: Ambient Cancellation Device for OATS



Brent and group,

I can verify that the CASPPER system does perform this feat incredibly well
in most cases IF you have the right conditions. I've had two demos of their
product, one at our company in Santa Monica and another at an OATS. We
devised quite a few scenarios to test its capabilities. For instance, at the
OATS, we found a strong ambient. Then we took a signal generator (acting as
EUT) and moved it slightly off of the center frequency of the ambient in
order to measure its magnitude. We moved the EUT signal back to the exact
center frequency of the ambient and the CASPPER system was able to remove
the ambient and take a reading of the EUT signal to within 0.5dB of our
reading. We then proceeded to AM and FM modulate the EUT signal and still
the CASPPER system did a great job of reading it.

However, if the system does not get the differential it needs between the
ambient sensing antenna and the EUT antenna, the results are underwhelming.
At the OATS (nestled in a valley) the system had problems with multipaths.
At our company in Santa Monica, the system had problems with multipaths from
high rise buildings and the overwhelming magnitude of some ambients.

I packed these two demonstrations with as many engineers and technicians as
I could in order to get a broad range of skepticism and opinion. Our
engineers who have DSP and communications systems experience have no qualms
about the theory and spoke directly with the CASPPER engineer to quiz him on
what they were doing with regard to coherence. We all were satisfied that
the system can do what it theoretically claims. However, it's a bit finicky
and the ambient antenna placement (or multiple antenna placements) seems to
be critical. At this point, I disagree with the marketing of the product
that it is a virtual chamber in that it in no way removes the headaches of
dealing with the multitude of ambients the way that a physical chamber
would. It may even be that the users of the system become hyper-aware of
ambients as they use the system.

I've asked CASPPER to give me the names and phone numbers of customers who
own the system to see how it's performing for them in the real world and
gauge their customer satisfaction, but have not gotten this information yet.

I'm still a little disappointed that I haven't heard any responses from this
list server to my specific questions on this topic (see my 3/21/00 post).

For those of you who have witnessed their demos:
1. Did the system perform to your satisfaction?
2. Were you convinced enough to purchase it?
3. Was it worth the asking price?

Mike Murphy
Compliance Engineer
Alesis Studio Electronics

---
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

To cancel your subscription, send mail to:
 majord...@ieee.org
with the single line:
 unsubscribe emc-pstc

For help, send mail to the list administrators:
 Jim Bacher:  jim_bac...@mail.monarch.com
 Michael Garretson:pstc_ad...@garretson.org

For policy questions, send mail to:
 Richard Nute:   ri...@ieee.org


---
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

To cancel your subscription, send mail to:
 majord...@ieee.org
with the single line:
 unsubscribe emc-pstc

For help, send mail to the list administrators:
 Jim Bacher:  jim_bac...@mail.monarch.com
 Michael Garretson:pstc_ad...@garretson.org

For policy questions, send mail to:
 Richard Nute:   ri...@ieee.org



What's your experience with GTEM?

2000-03-29 Thread Allen Tudor

Greetings,

I would appreciate any input with regard to GTEM results for pre-compliance 
emissions and immunity measurements.  Specifically, is there a limit to the 
size of a product that can be tested?  I have heard that small products such as 
pagers, etc. perform well in a GTEM.  Is there a limit to product volume vs. 
volume of the GTEM?  Also, how about cable radiation.  Will the GTEM 
sufficiently detect cable radiation with comparable results with an OATS?

Thanks in advance for your input.


Allen Tudor, Compliance Engineer
PairGain Technologies  tel:  (919)875-3382
6531 Meridien Drive fax: (919)876-1817
Raleigh, NC  27616   email:  allen_tu...@pairgain.com



---
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

To cancel your subscription, send mail to:
 majord...@ieee.org
with the single line:
 unsubscribe emc-pstc

For help, send mail to the list administrators:
 Jim Bacher:  jim_bac...@mail.monarch.com
 Michael Garretson:pstc_ad...@garretson.org

For policy questions, send mail to:
 Richard Nute:   ri...@ieee.org



Re: Stroboscopic light sources

2000-03-29 Thread Robert Macy

Not sure where to look.

But it is my understanding that the frequency of an impulse stimuli can
lock a person's alpha brain wave inducing trance-like states.  Guess at
around the 5Hz to 10Hz range.

In particularly susceptible people, certain pulsating frequencies can induce
epileptic seizures.

Recommend a search under externally induced epileptic seizures

  - Robert -

   Robert A. Macy, PEm...@california.com
   408 286 3985  fx 408 297 9121
   AJM International Electronics Consultants
   619 North First St,   San Jose, CA  95112


-Original Message-
From: Nick Williams n...@conformance.co.uk
To: emc-p...@majordomo.ieee.org emc-p...@majordomo.ieee.org
List-Post: emc-pstc@listserv.ieee.org
Date: Wednesday, March 29, 2000 9:57 AM
Subject: Stroboscopic light sources



Does anyone know of any good resources or actual requirements for the
safe use of stroboscopic light sources? The application is a public
science display. I recall rumors of stroboscopes operating at certain
frequencies being capable of triggering fits, but whether this is
just an old wive's tale or has some basis in reality I am unclear,
not do I know of any other potential hazards (except the obvious
dangers of syncronisation with moving machinery).

All input gratefully received.

Regards

Nick.



---
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

To cancel your subscription, send mail to:
 majord...@ieee.org
with the single line:
 unsubscribe emc-pstc

For help, send mail to the list administrators:
 Jim Bacher:  jim_bac...@mail.monarch.com
 Michael Garretson:pstc_ad...@garretson.org

For policy questions, send mail to:
 Richard Nute:   ri...@ieee.org



RE: Ambient Cancellation Device for OATS

2000-03-29 Thread Mike Murphy

Brent and group,

I can verify that the CASPPER system does perform this feat incredibly well
in most cases IF you have the right conditions. I've had two demos of their
product, one at our company in Santa Monica and another at an OATS. We
devised quite a few scenarios to test its capabilities. For instance, at the
OATS, we found a strong ambient. Then we took a signal generator (acting as
EUT) and moved it slightly off of the center frequency of the ambient in
order to measure its magnitude. We moved the EUT signal back to the exact
center frequency of the ambient and the CASPPER system was able to remove
the ambient and take a reading of the EUT signal to within 0.5dB of our
reading. We then proceeded to AM and FM modulate the EUT signal and still
the CASPPER system did a great job of reading it.

However, if the system does not get the differential it needs between the
ambient sensing antenna and the EUT antenna, the results are underwhelming.
At the OATS (nestled in a valley) the system had problems with multipaths.
At our company in Santa Monica, the system had problems with multipaths from
high rise buildings and the overwhelming magnitude of some ambients.

I packed these two demonstrations with as many engineers and technicians as
I could in order to get a broad range of skepticism and opinion. Our
engineers who have DSP and communications systems experience have no qualms
about the theory and spoke directly with the CASPPER engineer to quiz him on
what they were doing with regard to coherence. We all were satisfied that
the system can do what it theoretically claims. However, it's a bit finicky
and the ambient antenna placement (or multiple antenna placements) seems to
be critical. At this point, I disagree with the marketing of the product
that it is a virtual chamber in that it in no way removes the headaches of
dealing with the multitude of ambients the way that a physical chamber
would. It may even be that the users of the system become hyper-aware of
ambients as they use the system.

I've asked CASPPER to give me the names and phone numbers of customers who
own the system to see how it's performing for them in the real world and
gauge their customer satisfaction, but have not gotten this information yet.

I'm still a little disappointed that I haven't heard any responses from this
list server to my specific questions on this topic (see my 3/21/00 post).

For those of you who have witnessed their demos:
1. Did the system perform to your satisfaction?
2. Were you convinced enough to purchase it?
3. Was it worth the asking price?

Mike Murphy
Compliance Engineer
Alesis Studio Electronics

---
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

To cancel your subscription, send mail to:
 majord...@ieee.org
with the single line:
 unsubscribe emc-pstc

For help, send mail to the list administrators:
 Jim Bacher:  jim_bac...@mail.monarch.com
 Michael Garretson:pstc_ad...@garretson.org

For policy questions, send mail to:
 Richard Nute:   ri...@ieee.org



RE: EMC - Declaration of Incorporation?

2000-03-29 Thread Gert Gremmen
Nick,

Well said, but:

AS the PLC in question might be a component, it cannot be covered using
a DofC. Period. A DofI might just as well do the job, and the fact that it
has not been
defined in the EMC-directive, does not imply that incorporating one is
illegal.

Furthermore: common sense in fact is the basis of the ce marking system.
Lawyers
have made the system into a system of laws, but at the top EC-level, common
sense is represented by the EC system of due diligence and essential
requirements instead of strict prescripted standards and limits.
Many European courts cannot handle this properly, and your court example is
probably right. You should call for a higher justice (who helps me out with
the language)
to get your case defended right. Until then  you are on your own
;)

Regards,

Gert Gremmen, (Ing)

ce-test, qualified testing

===
Web presence  http://www.cetest.nl
CE-shop http://www.cetest.nl/ce_shop.htm
/-/ Compliance testing is our core business /-/
===


-Original Message-
From: owner-emc-p...@ieee.org [mailto:owner-emc-p...@ieee.org]On Behalf
Of Nick Williams
Sent: Tuesday, March 28, 2000 8:21 AM
To: Dan Kinney (A)
Cc: mertino...@skyskan.com; emc-p...@majordomo.ieee.org
Subject: RE: EMC - Declaration of Incorporation?



As an engineer, I have a lot of sympathy with the position taken by
Mike that the DofI would be a useful document for compliance with the
EMC Directive (and other directives too). However, the DofI is only
an available option for compliance with the Machinery Directive.

It's important to remember that we're talking politics and the law
here, not engineering common sense.  There are three routes to
compliance permitted by the EMC Directive and all require a DofC.
There is no provision for a DofI under the EMC directive so you can't
use it as a means of complying with the directive. Period. That's a
shame, but until there's a change in the law, that's the way it is.

It's all very well applying common sense to areas of the directives
which are not clear (and god knows there are plenty of those!).
However, the requirement for a DofC under the EMC Directive is not
one of them. If you ever found yourself in court trying to defend the
action you've taken to comply with the directive, imagine trying to
persuade the court that your judgement was superior to that of all
the lawyers who wrote the directive, and all those who are now lined
up waiting to demonstrate how much smarter they are than you.

Sorry if this sounds a bit terse: it isn't meant to be, but it's late
and I'm afraid I don't feel able to say it in a nicer way without
undesirable verbosity.

Regards

Nick.




At 10:39 -0500 27/3/2000, Dan Kinney (A) wrote:
Mike,
Wow - that helps a lot.  Thanks for the information.
Dan

Sincerely,
Daniel C. Kinney
Lead Qualification Engineer

Horner APG, LLC
Advanced Products Group
640 N. Sherman Drive
Indianapolis, IN  46201
Phone:   (317) 916-4274 ext. 462
FAX: (317) 916-4287
Email:   dan.kin...@heapg.com
Website:  http://www.heapg.com


  -Original Message-
  From: Michael Mertinooke [SMTP:mertino...@skyskan.com]
  Sent: Monday, March 27, 2000 10:28 AM
  To:   'Dan Kinney (A)'
  Subject:  RE: EMC - Declaration of Incorporation?

  Dan;
  Until 6 minths ago I was working for one of your competitors.
  All PLCs were shipped with Declarations of Conformity. The
  Declaration of Incorporation actually would be more appropriate,
  but we found our customers screaming for a DofC. So fine. We
  hired a Notified Body, set up TCFs, and went with DOCs.

  Also please note that the DOI is only mentioned in the
  Machinery Directive. This has often been interpreted to mean
  that it is only appropriate for mechanical parts. I'm not sure
  I agree with that. I think that the rules are not clear for
  something like a PLC, which has its own enclosure (and therefore
  is a device) but which does not perform a complete standalone
  action (and therefore is a component). In this case none of the
  rules fit exactly - so I am in favor of using whatever existing
  precedents you can find. In this case,  DOI would fit the situation
  perfectly: the device cannot be meaningfully tested all by itself,
  but you need to declare that when properly installed in accordance
  with user instructions, the device will meet all the declared
  requirements.

  One other point is the ongoing debate about random combinations
  of modular products. It is questionable whether the configuration
  you tested actually represents the real world. A DOI would sidestep
  this whole rathole, whereas a DOC is sort of a gamble. If you declare
  absolute conformity with a DOC, how do you know some customer won't
  put together a magic combination of modules that will violate
  emissions or immunity requirements? Personally, I spent a lot of
  test money proving to my satisfaction that I was 

RE: General Product Safety Directive

2000-03-29 Thread Gert Gremmen

Hi John, Richard and group,

I agree with your opinions for most of it but want to add this to it:

The application of the GPSD is meant exactly the way we handle
the generic EMC standards: as a backup to products not covered
by any other more dedicated  (ce) directive.

Of course for a directive to be more specific, it needs a scope,
having loop holes and omissions.

This does not mean that your life as a manufacturer has become
much more difficult, now selecting a directive for products
with multifunction properties may seem to lead to contradictories.

The EC target meant by using ce marking and the GPSD
(which is not ce-mark related - I believe) is to create
a situation in which safe products are created.
Any safety problem with any product will ultimately be judged
by the question: did the manufacturer show due diligence
in creating a safe product, and not by did he select
the right directive .

Directives and standards and essential requirements are a way
to handle this for mature manufacturers. As manufs are familiar with
their product best of all, they can (need to) be trusted to categorize
their product best. Due diligence is part of this idea.

The EC system is not meant to create some framework where every
product is covered by exactly the right directive and standard,
and where a smart manufacturer can escape testing by using aloop hole
in a directive or standards scope.
A manufacturer is not lead by hand, but need to motivate
his choices, and need to make a choice.

In any of the cases below, your reasoning should be:

Is my product safe for all these environments, and yes if
this means that it should be covered by multiple directives
at the same time, then it should be like that.

But then, select one , the one that fits best
to the original description of the product or covers
the most relevant risks
and create your files accordingly,
Fill up the safety loop holes leftover voluntarily.

As far as Richard's remarks concern:

Professional products need to be safe too, if no
other directive fits the product, how to show due diligence ?


Regards,

Gert Gremmen, (Ing)

ce-test, qualified testing

===
Web presence  http://www.cetest.nl
CE-shop http://www.cetest.nl/ce_shop.htm
/-/ Compliance testing is our core business /-/
===


-Original Message-
From: owner-emc-p...@ieee.org [mailto:owner-emc-p...@ieee.org]On Behalf
Of John Allen
Sent: Tuesday, March 28, 2000 4:54 PM
To: emc-p...@majordomo.ieee.org
Subject: RE: General Product Safety Directive



Hi folks

Very interesting and I am glad that Richard has spotted this as I
was under
the same impression (as, I think, were/are many other people).

However, I personally doubt it makes much difference to the TECHNICAL
safety requirements that should be applied to any specific product. These
mainly come from standards which are often sector independent.

Also, it should be noted that the discussion paper also says (in the
middle) that when professional products migrate to the consumer market
then the GPSD DOES now apply - so it might be difficult to
identify exactly
when that occurs.

Personal computers are probably both a good and a bad example:

a) A 230V desktop unit is definitely subject to the LVD and not the GPSD.

b) A notebook PC with an external AC/DC PSU is not subject to the
LVD, and
thus the GPSD probably applies (but see c) below!) - EXCEPT if it
is a very
high-spec high-end unit for professional use when the GPSD would not
appear to apply but it is then covered (according to the discussion
document) by work-place legislation.

c) A telecomms connectable notebook unit is subject to the RTTE
Directive,
not the LVD and that makes no supply-voltage rating distinction, nor any
distinction between consumer and non-consumer products - (and the
high-end
unit will almost certainly have telecomms connectability!).

d) High-spec units of this type often slide down the market to become
consumer products so the GPSD now applies to even the high-spec unit!

e) In all cases EN60950 is the applicable safety standard - so there's no
difference to the technical requirements!

f) Finally, I have seen many definitions of consumer that are not
explicit enough to exclude personnel in their working environment.

However, I can certainly see that where the product has no consumer
applications then the GPSD will not apply - but, again, will the actual
effects on the way we (as equipment suppliers) actually design,
manufacture
and supply products be much different? I rather doubt it.

I think the major effects of the GPSD are related to the needs to inform
various government bodies if you have been a bad boy - or you know of
someone who has - and the on-going implications thereof.

What do others make of this?

John Allen
Racal Defence Electronics Ltd
UK

--
From: wo...@sensormatic.com[SMTP:wo...@sensormatic.com]
Sent: 28 March 2000 13:40
To:  

Duplicate Messages

2000-03-29 Thread Jim Bacher

I have placed a call to the IEEE asking them to investigate the duplicate
messages.  No doubt they will find the problem and fix it.

Jim

---
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

To cancel your subscription, send mail to:
 majord...@ieee.org
with the single line:
 unsubscribe emc-pstc

For help, send mail to the list administrators:
 Jim Bacher:  jim_bac...@mail.monarch.com
 Michael Garretson:pstc_ad...@garretson.org

For policy questions, send mail to:
 Richard Nute:   ri...@ieee.org



EMC Directive Revisions

2000-03-29 Thread Price, Ed

I found an interesting document a few days ago. This is a draft of late 1999
revisions, actually amendments, to the EMC Directive. It looks like this
will supplement, not necessarily replace, the existing Directive. Since
nothing looked newer, and this is several months old, maybe this is nearly a
final version. FWIW, no guarantee about the veracity of this document!

I haven't read it in detail yet, but notice:

1. failure modes of which the user would not be aware

2. testing all manners of use

3.  the need for an analysis

The document is in the form of a 140K MS Word rtf. As I did last year with
the 461E document, I'll return a copy of this EMC Directive Amendments draft
by attachment to your email. (Email your request directly to me, not the
list.)

Ed


:-):-):-):-):-):-):-):-):-):-):-):-):-):-):-):-):-):-):-):-):-):-):-):-):-)
Ed Price
ed.pr...@cubic.com
Electromagnetic Compatibility Lab
Cubic Defense Systems
San Diego, CA.  USA
858-505-2780 (Voice)
858-505-1583 (Fax)
Military  Avionics EMC Services Is Our Specialty
Shake-Bake-Shock - Metrology - Reliability Analysis
:-):-):-):-):-):-):-):-):-):-):-):-):-):-):-):-):-):-):-):-):-):-):-):-):-)


---
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

To cancel your subscription, send mail to:
 majord...@ieee.org
with the single line:
 unsubscribe emc-pstc

For help, send mail to the list administrators:
 Jim Bacher:  jim_bac...@mail.monarch.com
 Michael Garretson:pstc_ad...@garretson.org

For policy questions, send mail to:
 Richard Nute:   ri...@ieee.org



Immediate Job Opening available for Certification Project Manager

2000-03-29 Thread Dieker, Paris

Compaq Computer EMEA BV, the regional headquarter for Europe, Middle East
and Africa of the Compaq Computer Corp. Houston/TX, located in
Munich/Germany is seeking for a Certification Project Manager.

Main responsibilities:
* Manage and coordinate Product Certification for selected countries in
Central and Eastern Europe, Africa and Middle East independently in
accordance with documented processes to obtain required regulatory
approvals.
* Support Compaq's entire product portfolio in the areas of Product Safety,
EMC and Telecommunications.
* Process customer regulatory inquiries in EU countries, Central and Eastern
Europe, Africa and the Middle East.
* Keep up-to-date with regulatory compliance requirements.
* Collect and communicate regional and national regulatory, certification
and standards requirements.
* Represent Compaq EMEA in relevant international standards and regulatory
meetings.
* Interface with regulatory authorities, certification bodies and test labs.
* Support our Product Divisions concerning technical localization of their
products

Education:
* BSEE, BSME or equivalent

Skills  Experience:
* 5+ years experience managing similar functions
* Experience in project and vendor management
* Demonstrated business understanding and development skills
* Strong communication and interpersonal skills. 
* Comfortable working with MS Office applications and project management
software. 
* Technical know-how and good understanding of  international standards
related to electrical safety, electromagnetic compatibility (EMC) and
telecommunication
* Experience in working with Testing Laboratories in the area of Agency
Approval/Product Certification.
* Ability to understand regulatory requirements and relevant technical
standards.
* Excellent oral and written communication skills, English mother tongue
preferred (another European language would be an asset but is not
essential).

Other:
* Teamwork
* Multi cultural work environment
* Relocation Package
* Willing to travel



Contact:
Paris Dieker [paris.die...@compaq.com] or
John W. Smith [johnw.sm...@compaq.com]

Compaq Computer EMEA BV
Einsteinring 30, 85609 Dornach / Munich, Germany
Phone: +49-(0)89 9392-2332 or -2338
Fax: +49-(0)89 9392-2336

---
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

To cancel your subscription, send mail to:
 majord...@ieee.org
with the single line:
 unsubscribe emc-pstc

For help, send mail to the list administrators:
 Jim Bacher:  jim_bac...@mail.monarch.com
 Michael Garretson:pstc_ad...@garretson.org

For policy questions, send mail to:
 Richard Nute:   ri...@ieee.org



Stroboscopic light sources

2000-03-29 Thread Nick Williams


Does anyone know of any good resources or actual requirements for the 
safe use of stroboscopic light sources? The application is a public 
science display. I recall rumors of stroboscopes operating at certain 
frequencies being capable of triggering fits, but whether this is 
just an old wive's tale or has some basis in reality I am unclear, 
not do I know of any other potential hazards (except the obvious 
dangers of syncronisation with moving machinery).


All input gratefully received.

Regards

Nick.

---
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

To cancel your subscription, send mail to:
majord...@ieee.org
with the single line:
unsubscribe emc-pstc

For help, send mail to the list administrators:
Jim Bacher:  jim_bac...@mail.monarch.com
Michael Garretson:pstc_ad...@garretson.org

For policy questions, send mail to:
Richard Nute:   ri...@ieee.org



RE: EN61326-1 Harmonics

2000-03-29 Thread Wagner, John P (John)

At the final stages of development of IEC 61000-3-2, TC77 WG1 responsible
for the document made an editorial change without review by voting bodies.
That change was to add the statement This section is a Product family
standard.  With the exception of a few questionable characters leading WG1,
no one thought of this as a product standard.

We have been at odds over this standard ever since.
John P. Wagner
Lucent Technologies, Bell Labs
11900 N. Pecos St, Room 2F58
Denver CO  80234
email:  johnwag...@lucent.com
phone:  303 538-4241
fax:  303 538-5211

 --
 From: Barry Ma[SMTP:barry...@altavista.com]
 Reply To: Barry Ma
 Sent: Tuesday, March 28, 2000 2:55 PM
 To:   chr...@gnlp.com
 Cc:   bkundew...@qtm.net; nprov...@foxboro.com; emc-p...@ieee.org
 Subject:  RE: EN61326-1 Harmonics
 
 
 Chris,
 
 I am impressed by your gentleman discussion manner. 
 
 We all agree that the committee of EN 61326-1 has very solid reason to
 exclude 3-2  3-3 for Class A equipment. If we had had the vote right we
 might have done the same thing.
 
 Unfortunately 3-2  3-3 became Product Standards with very broad
 definition. The rumor I heard is that these two standards were originally
 drafted as Basic Standards. ... (There must be some esteemed members in
 the EMC-PSTC group able to tell us what really happened.)
 
 If I have a piece of Class A lab equipment (referenced to EN 61326-1) with
 current  16A, I would rather test it for 3-2  3-3. Because I want to be
 prudent and conservative for the best interest of my company, the same
 attitude as you said: 
 
 I don't want cause my company to be seduced by the dark side of
 non-compliance.
 
 Thank you very much.
 Best Regards,
 Barry Ma
 b...@anritsu.com
 
 ---
 On Tue, 28 March 2000, Maxwell, Chris wrote:
  
  Barry,
  
  You have a great point.  It doesn't just apply to Information Technology
  Equipment. I apologize for using the term ITE loosely.  
  
  I feel (don't know) that the lowering of the current limit from 16A per
  phase (one of the main differences between IEC 555-2, 3 and IEC
 1000-3-2,3)
  was aimed at the proliferation of ITE and consumer products.  Someone at
 IEC
  realized that there could be a cumulative effect of harmonic currents.
  However, the scope of the standards is very broad.  It can be
 interpreted to
  include anything that uses an electron :-)   
  
  I felt that the commitee that wrote EN 61326-1 actually looked at the
  difference between Class A and Class B equipment within EN 61326-1 and
  consciously left the harmonics and flicker limits out of the Class A
  requirements.  This was confirmed by Norm Provost's reply to the thread.
 He
  participated in the writing and development of the standard.
  
  I think that you have a valid point in that EN 61326-1 treated EN
 61000-3-2
  and 61000-3-3 as Basic Standards as opposed to Product Standards.
 I
  never considered that point of view before your email.  But I want to
 know
  more.
  
  Now that I have conceded that I used ITE incorrectly, could I get an
  explanation of what makes IEC 1000-3-2 and IEC 1000-3-3 a Product
 Standard
  as opposed to a Basic Standard?
  
  Thanks,
  
  Chris Maxwell, Design Engineer
  GN Nettest Optical Division
  109 N. Genesee St.  
  Utica, NY 13502
  PH:  315-797-4449
  FAX:  315-797-8024
  EMAIL:  chr...@gnlp.com
  
   -Original Message-
   From:Barry Ma [SMTP:barry...@altavista.com]
   Sent:Tuesday, March 28, 2000 12:40 PM
   To:chr...@gnlp.com
   Cc:bkundew...@qtm.net; nprov...@foxboro.com; emc-p...@ieee.org
   Subject:RE: EN61326-1 Harmonics
   
   Hi Chris,
   
   Would you please prove that two product family standards 3-2 and 3-3
 are
   only applicable to ITE?
   
   Thanks.
   Barry Ma
   b...@anritsu.com
 
 
 
 
 For the largest MP3 index on the Web, go to http://mp3.altavista.com
 
 
 
 
 ---
 This message is from the IEEE EMC Society Product Safety
 Technical Committee emc-pstc discussion list.
 
 To cancel your subscription, send mail to:
  majord...@ieee.org
 with the single line:
  unsubscribe emc-pstc
 
 For help, send mail to the list administrators:
  Jim Bacher:  jim_bac...@mail.monarch.com
  Michael Garretson:pstc_ad...@garretson.org
 
 For policy questions, send mail to:
  Richard Nute:   ri...@ieee.org
 
 

---
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

To cancel your subscription, send mail to:
 majord...@ieee.org
with the single line:
 unsubscribe emc-pstc

For help, send mail to the list administrators:
 Jim Bacher:  jim_bac...@mail.monarch.com
 Michael Garretson:pstc_ad...@garretson.org

For policy questions, 

JAPAN

2000-03-29 Thread Rafael González

Dear group,

How can I know the standard code to apply if I want to install a machine
in Japan ?
It is enough to be in compliance with the UL508 ?

Thanks in advance for your help,


Rafael González Licerán
Area Seguridad Eléctrica
rgonzál...@cetecom.es
_
CENTRO DE TECNOLOGIA DE LAS COMUNICACIONES, S.A.
PTA - C/Severo Ochoa 2, 29590 Campanillas (Málaga)
Tel: 952619100 - Fax: 952619113 - Web: http://www.cetecom.es/
_




---
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

To cancel your subscription, send mail to:
 majord...@ieee.org
with the single line:
 unsubscribe emc-pstc

For help, send mail to the list administrators:
 Jim Bacher:  jim_bac...@mail.monarch.com
 Michael Garretson:pstc_ad...@garretson.org

For policy questions, send mail to:
 Richard Nute:   ri...@ieee.org



REQUEST FOR HELP WITH LOOP ANTENNA DESIGN

2000-03-29 Thread Andrew Wood

At the other end of the spectrum from Frank's request- I have misplaced my
design guide for loop antenna. Can anyone help with a source of information?
 You may wish to reply direct to avoid clogging the system.

Thanks in advance.
Andy Wood.

---
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

To cancel your subscription, send mail to:
 majord...@ieee.org
with the single line:
 unsubscribe emc-pstc

For help, send mail to the list administrators:
 Jim Bacher:  jim_bac...@mail.monarch.com
 Michael Garretson:pstc_ad...@garretson.org

For policy questions, send mail to:
 Richard Nute:   ri...@ieee.org



Re: StripLine

2000-03-29 Thread Robert Macy

I've got a copy of DimStrip which calculates their characteristics.  Either
electrical properties from dimensions, or dimensions from required
electrical characteristics for stripline, or microstrip.

Runs under DOS and also has adjacent line capability.

Is there a place I can post it so people can stop by and pick it up and not
have to send the 100K to everyone?

  - Robert -

   Robert A. Macy, PEm...@california.com
   408 286 3985  fx 408 297 9121
   AJM International Electronics Consultants
   619 North First St,   San Jose, CA  95112

-Original Message-
From: Frank Krozel fr...@electronicinstrument.com
To: emc-p...@ieee.org emc-p...@ieee.org
Date: Tuesday, March 28, 2000 12:27 PM
Subject: Re: StripLine


Hi All:
Seem to have misplaced my data on Striplines.
Is there somewhere on the web that I can find the dimensions to build a
stripline?
Frank Krozel


---
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

To cancel your subscription, send mail to:
 majord...@ieee.org
with the single line:
 unsubscribe emc-pstc

For help, send mail to the list administrators:
 Jim Bacher:  jim_bac...@mail.monarch.com
 Michael Garretson:pstc_ad...@garretson.org

For policy questions, send mail to:
 Richard Nute:   ri...@ieee.org



RE: EMC/PSTC/NEBS/TREG

2000-03-29 Thread Lou Gnecco

Im with Dave. it is getting too cumbersome. we are going to have to split it
up eventually.
Lou

At 03:51 PM 3/28/2000 -0500, you wrote:

Average number of messages is now between 30 and 50 a day.  And people want
more!  I wish I had enough free company time to participate in this mountain
of mail.  We need to increase the quality and substance of the messages
rather than to increase the amount.
Dave George
Unisys


-Original Message-
From: pmerguer...@itl.co.il [mailto:pmerguer...@itl.co.il]
Sent: 17 March, 2000 4:04 AM
To: emc-p...@majordomo.ieee.org
Subject: EMC/PSTC/NEBS/TREG



Dear All,

DO NOT SPLIT! I AM ALSO IN FAVOR OF MIGRATING TREG AND NEBS GROUPS INTO TO
THE EMC/PSTC LIST. ARE YOU ALL IN FAVOR?
Peter Merguerian
Managing Director
Product Testing Division
I.T.L. (Product Testing) Ltd.
Hacharoshet 26, POB 211
Or Yehuda 60251, Israel

Tel: 972-3-5339022 Fax: 972-3-5339019
e-mail: pmerguer...@itl.co.il
website: http://www.itl.co.il 






---
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

To cancel your subscription, send mail to:
 majord...@ieee.org
with the single line:
 unsubscribe emc-pstc

For help, send mail to the list administrators:
 Jim Bacher:  jim_bac...@mail.monarch.com
 Michael Garretson:pstc_ad...@garretson.org

For policy questions, send mail to:
 Richard Nute:   ri...@ieee.org


---
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

To cancel your subscription, send mail to:
 majord...@ieee.org
with the single line:
 unsubscribe emc-pstc

For help, send mail to the list administrators:
 Jim Bacher:  jim_bac...@mail.monarch.com
 Michael Garretson:pstc_ad...@garretson.org

For policy questions, send mail to:
 Richard Nute:   ri...@ieee.org




---
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

To cancel your subscription, send mail to:
 majord...@ieee.org
with the single line:
 unsubscribe emc-pstc

For help, send mail to the list administrators:
 Jim Bacher:  jim_bac...@mail.monarch.com
 Michael Garretson:pstc_ad...@garretson.org

For policy questions, send mail to:
 Richard Nute:   ri...@ieee.org



Fwd: Re: modest proposal

2000-03-29 Thread Barry Ma

Sorry for bothering those who are not interested in.
Rene, Please forgive me Fwding your email to the group.
Barry Ma
--- Start of forwarded message ---
 
Subject: Re: modest proposal - unl at unu.edu
To: Barry Ma barry...@altavista.com
From: r...@twn.tuv.com
List-Post: emc-pstc@listserv.ieee.org
Date: Wed, 29 Mar 2000 08:04:29 +0800

There is already a project running that will address item 1) of the mail
shown below:

http://www.unl.ias.unu.edu/

As the central concept of this project seems to be about grammar, it will
probably not be able to fulfil the requirement of item 2) of the mail
shown below.

Learning vocabulary is learning facts, that might be done during sleeping (with
a tape-machine running under the pillow). Grammar in my opinion is more a 
concept, a thing of logic. For learning grammar you probably have to be awake.

Regards

Rene Charton

--

Barry Ma barry...@altavista.com on 03/28/2000 08:07:25 AM

To:   emc-p...@ieee.org
cc:   Lou Gnecco l...@tempest-inc.com (bcc: Rene Charton/TUV-Twn)
Subject:  Re: modest proposal

Hi Lou,

There must be some day in the future, the artificial intelligence has been so
well developed that
(1) An instant interpreting machine built-in to your PC would automatically
transfer any language you input (either typed or voiced) to any languages the
other party would like to have.
(2) Learning foreign language is a very pleasant process and can be completed in
very short period of time even when you are in sleep. ...  :-)

Best Regards,
Barry Ma
b...@anritsu.com
 
--- End of forwarded message ---




For the largest MP3 index on the Web, go to http://mp3.altavista.com




---
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

To cancel your subscription, send mail to:
 majord...@ieee.org
with the single line:
 unsubscribe emc-pstc

For help, send mail to the list administrators:
 Jim Bacher:  jim_bac...@mail.monarch.com
 Michael Garretson:pstc_ad...@garretson.org

For policy questions, send mail to:
 Richard Nute:   ri...@ieee.org



RE: modest proposal

2000-03-29 Thread Joe Finlayson


Am I the only one receiving posts in duplicate?  It seems that this
has happened at least a dozen times.


Joe Finlayson
Manager, Compliance Engineering
Telica, Inc.
734 Foster Street, Bldg. G, Suite 100
Marlboro, MA 01752
Tel:(508) 480-0909 x212
Fax:(508) 480-0922
Email:  jfinlay...@telica.com



-Original Message-
From: Gert Gremmen [mailto:cet...@cetest.nl]
Sent: Tuesday, March 28, 2000 1:28 PM
To: geor...@lexmark.com
Cc: emc-p...@ieee.org
Subject: RE: modest proposal


Thank you, Thank you, and applause

Regards,
 
Gert Gremmen, (Ing)

ce-test, qualified testing
 
===
Web presence  http://www.cetest.nl
CE-shop http://www.cetest.nl/ce_shop.htm
/-/ Compliance testing is our core business /-/
===


-Original Message-
From: owner-emc-p...@ieee.org [mailto:owner-emc-p...@ieee.org]On Behalf
Of geor...@lexmark.com
Sent: Tuesday, March 28, 2000 6:20 PM
To: m.r...@ieee.org
Cc: emc-p...@ieee.org
Subject: Re: modest proposal



Martin,

Your post included the following:

* We, the ugly Americans, want the world to conform
  to our native language.  We're too lazy, stubborn,
  and arrogant to learn another language.

You may be partially right, but I believe there is a
much simpler explanation.  It is human nature to do
only that which we are motivated to do.  The English
speaking world has been fortunate in not having to
learn another predominate language to conduct global
business.  This is probably due to the fact that most
non-English speaking countries do not agree that French,
German, Spanish, etc. are an acceptable alternate
global langauge.  Therefore, English may have won
only by default.

Here is what I remember of the U.S. interest in other
languages.  Prior to WWII U.S. schools taught Latin as
a way to learn the root of words.  There was a little
French, German, and Spanish taught.  After WWII, it was
thought that we should be learning Russian, as the other
major technical country.  Then, in the '70's or so, it
was thought that Japanese may be the main other language
to learn.

In summary, Americans have never had any reason to pick
one particular other language to learn.  Many have studied
other languages, but more for personal than business
reasons. Personally, I studied Spanish in high school, and
German a few years ago, but am not fluent in either,as
there are few opportunities to practive what little I
learned.

There are people in every country that are too lazy,
stubborn, and arrogant to learn another language.  But
I find that educated professionals will learn what they
need to learn to conduct business in their chosen career.

Finally, I greatly respect and am thankful for the many
non-English speaking peoples who have learned this very
difficult langauge for global business purposes.  For
this reason, I am never critical of their English spelling
or errors in grammer.  I can only imagine the result of
my trying to use German, French, etc.!

Regards,

George Alspaugh



---
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

To cancel your subscription, send mail to:
 majord...@ieee.org
with the single line:
 unsubscribe emc-pstc

For help, send mail to the list administrators:
 Jim Bacher:  jim_bac...@mail.monarch.com
 Michael Garretson:pstc_ad...@garretson.org

For policy questions, send mail to:
 Richard Nute:   ri...@ieee.org



---
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

To cancel your subscription, send mail to:
 majord...@ieee.org
with the single line:
 unsubscribe emc-pstc

For help, send mail to the list administrators:
 Jim Bacher:  jim_bac...@mail.monarch.com
 Michael Garretson:pstc_ad...@garretson.org

For policy questions, send mail to:
 Richard Nute:   ri...@ieee.org



Re: modest proposal - unl at unu.edu

2000-03-29 Thread rc

There is already a project running that will address   item 1)of the mail
shown below:

http://www.unl.ias.unu.edu/

As the central concept of this project seems to be about grammar, it will
probably not be able to fulfil the requirement of item 2)of the mail
shown below.

Learning vocabulary is learning facts, that might be done during sleeping (with
a tape-machine running under the pillow).
Grammar in my opinion is more a concept, a thing of logic. For learning grammar
you probably have to be awake.

Regards

Rene Charton






Barry Ma barry...@altavista.com on 03/28/2000 08:07:25 AM

Please respond to Barry Ma barry...@altavista.com

To:   emc-p...@ieee.org
cc:   Lou Gnecco l...@tempest-inc.com (bcc: Rene Charton/TUV-Twn)
Subject:  Re: modest proposal




Hi Lou,

There must be some day in the future, the artificial intelligence has been so
well developed that
(1) An instant interpreting machine built-in to your PC would automatically
transfer any language you input (either typed or voiced) to any languages the
other party would like to have.
(2) Learning foreign language is a very pleasant process and can be completed in
very short period of time even when you are in sleep. ...  :-)

Best Regards,
Barry Ma
b...@anritsu.com











---
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

To cancel your subscription, send mail to:
 majord...@ieee.org
with the single line:
 unsubscribe emc-pstc

For help, send mail to the list administrators:
 Jim Bacher:  jim_bac...@mail.monarch.com
 Michael Garretson:pstc_ad...@garretson.org

For policy questions, send mail to:
 Richard Nute:   ri...@ieee.org



RE: Ambient Cancellation Device for OATS

2000-03-29 Thread pahlbf




-- Forwarded by Brent F Pahl/FOS/PEC on 03/28/2000 03:59 PM
---


Brent F Pahl
03/23/2000 01:43 PM

To:   Lacey,Scott sla...@foxboro.com
cc:

Subject:  RE: Ambient Cancellation Device for OATS  (Document link: Brent F
  Pahl)

I recently attended one of the CASSPER demonstrations, and when asked the
question of how their instrument deals with EUT noise buried under ambient
noise, they gave this response:  The CASSPER system not only compares the
frequency of the incoming ambient (ambients being brought in through the
distant antenna), but it also compares the phase and waveform of the signal.
If the phase and waveform between the two antennae signals does not match, then
the signal is not canceled out.  They specifically claim that they can find EUT
noise buried under ambient noise.  Honestly, I'm still a little skeptical.
Hopefully I'll be able to test this and see how true it is some time soon.

Cheers,
Brent





Lacey,Scott sla...@foxboro.com on 03/23/2000 09:34:35 AM

Please respond to Lacey,Scott sla...@foxboro.com

To:   'Brumbaugh, David' david.brumba...@pss.boeing.com
cc:   'emc-p...@ieee.org' emc-p...@ieee.org (bcc: Brent F Pahl/FOS/PEC)

Subject:  RE: Ambient Cancellation Device for OATS




Exactly.
Scott Lacey

 -Original Message-
 From: Brumbaugh, David [SMTP:david.brumba...@pss.boeing.com]
 Sent: Thursday, March 23, 2000 11:19 AM
 To:  emc-p...@majordomo.ieee.org
 Subject:  RE: Ambient Cancellation Device for OATS


 I would like to add that one must be careful when it comes to using
such a system. I apologize if this has been pointed out already, but it
can't be emphasized enough. When using a system like this, you take the risk
that the ambient environment will mask significant emissions from the EUT.
You may never see these emissions under such conditions, and it will bite
you later.

 I am particularly concerned when I hear this being touted as a
panacea for testing in traditionally noisy environments, such as development
labs. Be careful folks.

 My 2 cents.

 DB
  --
  From:   Price, Ed[SMTP:ed.pr...@cubic.com]
  Reply To:Price, Ed
  Sent:   Thursday, March 23, 2000 9:01 AM
  To: 'Doug'; emc-p...@majordomo.ieee.org
  Subject: RE: Ambient Cancellation Device for OATS
 
 
  Doug:
 
  As I understand the process, there isn't actually cancellation AT
the OATS.
  They have a monitor antenna nearby, but relatively far from the
EUT on the
  OATS. A real EUT signal should fall off considerably by the time
it reaches
  the monitor antenna, but an ambient emission will be relatively
constant at
  the monitor and measurement antennas. All you have to do is sample
the
  ambient, invert it, and sum it with the measurement antenna
voltage. This
  should delete the ambient from the measurement data. (Easy to say!
All you
  need is enough computing power and sufficiently capable processing
  algorithms.)
 
  The cancellation takes place as a data operation within the
Cassper
  hardware (or it's controlling computer), and not as an energy
cancellation
  of the propagating fields actually on the OATS site.
 
  And yes, I would also like to know more about how Cassper manages
to do
  this. But, I also realize they can't give away their trade secrets
either.
 
  Ed
 
 
 
:-):-):-):-):-):-):-):-):-):-):-):-):-):-):-):-):-):-):-):-):-):-):-):-):-)
  Ed Price
  ed.pr...@cubic.com
  Electromagnetic Compatibility Lab
  Cubic Defense Systems
  San Diego, CA.  USA
  858-505-2780 (Voice)
  858-505-1583 (Fax)
  Military  Avionics EMC Services Is Our Specialty
  Shake-Bake-Shock - Metrology - Reliability Analysis
 
:-):-):-):-):-):-):-):-):-):-):-):-):-):-):-):-):-):-):-):-):-):-):-):-):-)
 
   -Original Message-
   From: Doug [SMTP:dmck...@gte.net]
   Sent: Wednesday, March 22, 2000 11:38 PM
   To:   emc-p...@majordomo.ieee.org
   Subject:   Re: Ambient Cancellation Device for OATS
  
  
   Perhaps it's because I've never understood ...
  
   How exactly does one do ambient cancellation at an OATS?
  
 
 
 

 ---
 This message is from the IEEE EMC Society Product Safety
 Technical Committee emc-pstc discussion list.

 To cancel your subscription, send mail to:
  majord...@ieee.org
 with the single line:
  unsubscribe emc-pstc

 For help, send mail to the list administrators:
  Jim Bacher:  jim_bac...@mail.monarch.com
  Michael Garretson:pstc_ad...@garretson.org

 For policy questions, send mail to:
  Richard Nute:   ri...@ieee.org


---
This message is from the IEEE EMC Society Product