2 questions. 1) HP software for 7400A analyzers, 2) FDA letters of Accession

2002-05-14 Thread Gary McInturff

1) I believe someone out there commented on using the HP 74XX series 
analyzer for pre-compliance measurements. Do you also use the additional 
software that can be purchased? 
The unit I played with had no additional software, but I could make 
measurements against limits lines etc, and it factored in the transceiver 
gains/losses and then allowed me to export a summary sheet for any reports I 
wanted to generate. What functions and value does the additional software bring 
to the table.?

2) Does anybody out there get any traction from a "Letters of 
accession" that the FDA sends to a optics vendor after receiving a request for 
a model addition?
This letter says nothing useful for NRTL's and always includes "This 
acknowledgement does not constitute approval or the document".
The FEDS are disavowing any level of conformity assessment, and the 
NRTL's I use tell me they can't use it, even for an unrecognized componet, yet 
the vendors are insistent that I am the only unaccepting curmudgeon in the 
entire universe. 

Sorry if you've heard this before but I just can't believe it keeps 
happening, and that tells me that I should double check my facts.

Thanks
Gary

---
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/

To cancel your subscription, send mail to:
 majord...@ieee.org
with the single line:
 unsubscribe emc-pstc

For help, send mail to the list administrators:
 Ron Pickard:  emc-p...@hypercom.com
 Dave Heald:   davehe...@attbi.com

For policy questions, send mail to:
 Richard Nute:   ri...@ieee.org
 Jim Bacher: j.bac...@ieee.org

All emc-pstc postings are archived and searchable on the web at:
http://ieeepstc.mindcruiser.com/
Click on "browse" and then "emc-pstc mailing list"


RE: Constant for Change of Resistance formula.

2002-05-14 Thread Robert Wilson

Stating the proven principles of accurate thermal analysis is hardly
"muddling the knowledge pool". 

No one has suggested that increased resistance leads to increasing heat
generation in an infinite spiral as you mention. This would obviously be
nonsense, and is not predicted by theory or observed in practice. 

As I mentioned in an earlier post, increased resistance DOES lead to
increased heat generation, which DOES in turn cause further resistance
increase ...but this does not spiral on to infinity for the very reason
that the system is nonlinear. Thermal issues ARE highly nonlinear. What
DOES happen is that a new higher temperature plateau is reached where
(as I wrote in my original post on this subject):

"The final temperature that the "system" stabilizes at, is reached when
the logarithmically increasing (i.e. also very non-linear) heat transfer
to the environment caused by increasing temperature, balances increased
heat being generated."

The bottom line is that heat transfer issues are unavoidably complex and
require iterative solutions for accurate answers. That is why thermal
analysis software is so hideously expensive, and requires such long
times and high computer "horsepower" to converge on an accurate
solution. While it may be possible to arrive at an approximate solution,
for a limited set of parameters under a narrow subset of conditions by
using rule-of-thumb simplification, it doesn't change the fact that an
accurate solution is unavoidably far more complicated than you are
presenting, and such a simplification must be understood to be just
that: a simplification.

Bob Wilson
TIR Systems Ltd.
Vancouver.

-Original Message-
From: Sam Davis [mailto:sda...@ptitest.com] 
Sent: May 14, 2002 8:57 AM
To: emc-p...@majordomo.ieee.org
Subject: RE: Constant for Change of Resistance formula.


OK, I've avoided jumping in on this exhausted thread, but here's my take
on
the situation.

It's obvious that more than a few of the posters have not performed this
test.  When you cannot answer a question with more than a theory, don't
throw it out there as fact.  You muddle the knowledge pool.

First off, why go to the trouble of performing the CoR measurements on a
connector, when you can fairly accurately (and much more simply) use a
thermocouple.  If you decide you need the accuracy of the Change of
Resistance measurement, the equipment used for this test is specialized
and
highly precise, (it's a milliohm meter, not your standard DMM).  The
meters
I've used have been capable of measuring fractions of milliohms, down to
microohms.  And you do end up measuring the resistance of the leads.
That's
why you measure them separately, in the ambient, and subtract them from
the
system resistance in the formula to get the resistance of the EUT only.
Also, there is specially designed equipment to perform this test while
the
EUT is energized, but the normal method is to run the EUT until thermal
stabilization, disconnect power, and measure the resistance as it drops
over
time, and extrapolate back to time 0.  The smaller the EUT, the faster
you
need to get the first measurement, and subsequent measurements because
within seconds, the EUT could drop significantly, making your
extrapolation
inaccurate.

And about the resistance to temp rise to resistance rise to temp rise -
if
it went on infinitum, all conductors (not just those under test) would
eventually ignite.  This only happens when you allow too much current.

Sam
Disclaimer - Sorry if I stepped on any toes, but I've got big feet.

---
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/

To cancel your subscription, send mail to:
 majord...@ieee.org
with the single line:
 unsubscribe emc-pstc

For help, send mail to the list administrators:
 Ron Pickard:  emc-p...@hypercom.com
 Dave Heald:   davehe...@attbi.com

For policy questions, send mail to:
 Richard Nute:   ri...@ieee.org
 Jim Bacher: j.bac...@ieee.org

All emc-pstc postings are archived and searchable on the web at:
http://ieeepstc.mindcruiser.com/
Click on "browse" and then "emc-pstc mailing list"


---
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/

To cancel your subscription, send mail to:
 majord...@ieee.org
with the single line:
 unsubscribe emc-pstc

For help, send mail to the list administrators:
 Ron Pickard:  emc-p...@hypercom.com
 Dave Heald:   davehe...@attbi.com

For policy questions, send mail to:
 Richard Nute:   ri...@ieee.org
 Jim Bacher: j.bac...@ieee.org

All emc-pstc postings are archived and searchable on the web at:
htt

Re: Lightning and power cross in Japan

2002-05-14 Thread Scott Roleson

Hello Joe,

In reply to your TREG message of May 14 regarding lightning incidence in
Japan and any protection requirements on telecom equipment, here are 
several resources that you might find helpful.

The Colorado Lightning Research Center focuses mostly on lightning in 
the U.S.A., but their Web site shows a map of lightning incidence
worldwide.  This map shows Japan as having lightning incidence of 
around 8-20 flashes/sq km/year (if I am interpreting their map correctly), 
which is similar to the midwestern U.S.A. or southern Canada.  
See:  http://www.crh.noaa.gov/pub/ltg.html

The National Lightning Safety Institute also has a map showing 
lightning strike regions in Japan, with data on the average number of
thunderstorms per year at:
http://www.lightningsafety.com/nlsi_info/lightningmaps/JPNghtning.html

You might also check out the Web site of the Society of Atmospheric 
Electricity of Japan at:  
http://lightning.pwr.eng.osaka-u.ac.jp/saej/e_index.html

NTT (Japan) standardization activities are discussed in their "NTT 
Review" periodical.  It seems to me that this might be a place where
research on lightning effects on telecom equipment in Japan might be 
reported.  The NTT Review is available on the web at:  
http://www.ntt.co.jp/RD/Enindex.html 

Hope you find these helpful.

Best regards,

  -- Scott Roleson

-
 Scott Roleson, PE   (MS-61U74) |  E-mail: srole...@ieee.org
 Telecom & EMC Engineer |  http://www.roleson.com
 Corporate External Standards   |  Telephone: +1-858-655-4809
 Hewlett-Packard Company|  FAX:   +1-858-655-5931
 16399 W. Bernardo Drive|  Amateur Radio:  KC7CJ
 San Diego, CA  92127-1899  USA |  Any opinions are my own, not HP's.
-




Lightning and power cross in Japan

2002-05-14 Thread JPR3

Hello All:

I am trying to determine whether there are any published requirements for 
lightning immunity or power cross for wireline PSTN terminal equipment in 
Japan.  The JATE requirements for PSTN terminal equipment do not contain any 
such tests, and I am not aware of any similar requirements imposed by various 
industry specifications.

Do any of you know of any specific requirements for Japan?  

On a related note, how severe is the actual lightning environment in Japan, 
compared to say, North America or Europe?


Joe Randolph
Telecom Design Consultant
Randolph Telecom, Inc.
781-721-2848
http://www.randolph-telecom.com

---
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/

To cancel your subscription, send mail to:
 majord...@ieee.org
with the single line:
 unsubscribe emc-pstc

For help, send mail to the list administrators:
 Ron Pickard:  emc-p...@hypercom.com
 Dave Heald:   davehe...@attbi.com

For policy questions, send mail to:
 Richard Nute:   ri...@ieee.org
 Jim Bacher: j.bac...@ieee.org

All emc-pstc postings are archived and searchable on the web at:
http://ieeepstc.mindcruiser.com/
Click on "browse" and then "emc-pstc mailing list"


RE: Constant for Change of Resistance formula.

2002-05-14 Thread Sam Davis

OK, I've avoided jumping in on this exhausted thread, but here's my take on
the situation.

It's obvious that more than a few of the posters have not performed this
test.  When you cannot answer a question with more than a theory, don't
throw it out there as fact.  You muddle the knowledge pool.

First off, why go to the trouble of performing the CoR measurements on a
connector, when you can fairly accurately (and much more simply) use a
thermocouple.  If you decide you need the accuracy of the Change of
Resistance measurement, the equipment used for this test is specialized and
highly precise, (it's a milliohm meter, not your standard DMM).  The meters
I've used have been capable of measuring fractions of milliohms, down to
microohms.  And you do end up measuring the resistance of the leads.  That's
why you measure them separately, in the ambient, and subtract them from the
system resistance in the formula to get the resistance of the EUT only.
Also, there is specially designed equipment to perform this test while the
EUT is energized, but the normal method is to run the EUT until thermal
stabilization, disconnect power, and measure the resistance as it drops over
time, and extrapolate back to time 0.  The smaller the EUT, the faster you
need to get the first measurement, and subsequent measurements because
within seconds, the EUT could drop significantly, making your extrapolation
inaccurate.

And about the resistance to temp rise to resistance rise to temp rise - if
it went on infinitum, all conductors (not just those under test) would
eventually ignite.  This only happens when you allow too much current.

Sam
Disclaimer - Sorry if I stepped on any toes, but I've got big feet.

---
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/

To cancel your subscription, send mail to:
 majord...@ieee.org
with the single line:
 unsubscribe emc-pstc

For help, send mail to the list administrators:
 Ron Pickard:  emc-p...@hypercom.com
 Dave Heald:   davehe...@attbi.com

For policy questions, send mail to:
 Richard Nute:   ri...@ieee.org
 Jim Bacher: j.bac...@ieee.org

All emc-pstc postings are archived and searchable on the web at:
http://ieeepstc.mindcruiser.com/
Click on "browse" and then "emc-pstc mailing list"


---
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/

To cancel your subscription, send mail to:
 majord...@ieee.org
with the single line:
 unsubscribe emc-pstc

For help, send mail to the list administrators:
 Ron Pickard:  emc-p...@hypercom.com
 Dave Heald:   davehe...@attbi.com

For policy questions, send mail to:
 Richard Nute:   ri...@ieee.org
 Jim Bacher: j.bac...@ieee.org

All emc-pstc postings are archived and searchable on the web at:
http://ieeepstc.mindcruiser.com/
Click on "browse" and then "emc-pstc mailing list"


KOREAN S MARK REQUIREMENTS

2002-05-14 Thread Gordon,Ian

Does anybody know how to achieve compliance with the Korean S mark
legislation?
Ian Gordon



_
This message has been checked for all known viruses by UUNET delivered 
through the MessageLabs Virus Control Centre. For further information visit
http://www.uk.uu.net/products/security/virus/

---
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/

To cancel your subscription, send mail to:
 majord...@ieee.org
with the single line:
 unsubscribe emc-pstc

For help, send mail to the list administrators:
 Ron Pickard:  emc-p...@hypercom.com
 Dave Heald:   davehe...@attbi.com

For policy questions, send mail to:
 Richard Nute:   ri...@ieee.org
 Jim Bacher: j.bac...@ieee.org

All emc-pstc postings are archived and searchable on the web at:
http://ieeepstc.mindcruiser.com/
Click on "browse" and then "emc-pstc mailing list"


RE: Constant for Change of Resistance formula.

2002-05-14 Thread Chris Maxwell

Robert,

Perhaps your take on the situation is indeed on target.  I'm not sure myself if 
I would use the formula in the situation that you describe.   

>The original poster stated that he was trying to determine the change 
of
>temperature in a connector, caused by increased resistance of its
>conductors. This clearly implies that the increased temperature must be
>due to additional resistive heating of these contacts, which in turn
>means that there MUST be sufficient current flowing in the connection 
to
>cause non-negligible heating. After all, if the additional heating due
>to current flow through a more resistive contact material was, in fact,
>negligible, then the connector would not get hotter, and the OP would
>not be worrying about "change of temperature based on change of
>resistance" as he stated. 

Too many variables...current is still flowing...causing heating and dynamic 
resistance change...connectors have mating surfaces...which dominate the 
resistance.  I didn't see terms in the formula that took these factors into 
account.

Another problem that I can see with this is measuring the resistance in the 
first place.  Most connectors have a resistance which is a fraction of an 
Ohm...You would almost need a bridge or other high precision device to measure 
in the first place; and even then you would mostly be measuring the resistance 
of the interface, not the contacts themselves.

I never let pride stand in the way of the facts...I'm starting to lean toward 
your way of thinking on this one.

Chris



---
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/

To cancel your subscription, send mail to:
 majord...@ieee.org
with the single line:
 unsubscribe emc-pstc

For help, send mail to the list administrators:
 Ron Pickard:  emc-p...@hypercom.com
 Dave Heald:   davehe...@attbi.com

For policy questions, send mail to:
 Richard Nute:   ri...@ieee.org
 Jim Bacher: j.bac...@ieee.org

All emc-pstc postings are archived and searchable on the web at:
http://ieeepstc.mindcruiser.com/
Click on "browse" and then "emc-pstc mailing list"


RE: Constant for Change of Resistance formula.

2002-05-14 Thread Bill Ellingford

Hi Robert, Group
Yes, this is a known factor for mis-measurement.  You must always isolate
the component part / winding etc and measure each resistive element to find
the individual temperature rises.  This is common practice when measuring
motor windings, you would take a sample of windings measuring at the brush
caps (hot brushes quickly removed) at pre-marked segments on the commutator.
This allows you to get a picture of the heat characteristics of a number of
windings and hence heat spread throughout the motor windings.  Yes, accuracy
depends on the resistance being measured but where the formula is applied
(motors, relay coils, transformers, solenoids and other mechelec devices),
the resistance is high enough to get a good result as long as you follow the
guidelines (like those above).

Thanks for raising the measurement factors.

Bill Ellingford

-Original Message-
From: Robert Macy [mailto:m...@california.com]
Sent: 14 May 2002 02:00
To: Bill Ellingford; emc-p...@majordomo.ieee.org
Subject: Re: Constant for Change of Resistance formula.


Bill,

Big oops.

Measuring the resistance to determine the temperature is not productive
*unless* the resistance dominates the resistance measurement.

Picture three equal valued resistances in a row.  The middle one gets very
hot (more than 100C rise) and increases over 40%, the two on the edges are
heat sinked and barely increase in temperature.  The resulting change in
resistance is 13% which implies the temperature in there has only gone up
around 33C.

Measuring the resistance doesn't tell you much.  At least with transformers
the dominant resistance is pretty much the bulk resistance.

- Robert -


-Original Message-
From: Bill Ellingford 
To: 'Robert Macy' ; emc-p...@majordomo.ieee.org

List-Post: emc-pstc@listserv.ieee.org
Date: Monday, May 13, 2002 1:56 PM
Subject: RE: Constant for Change of Resistance formula.


>
>Hi Robert / group
>OK, Not the best choice of website to demo the answer.  The differing
>figures are because the formula has been transposed to give Temp from
change
>of R from the original formula which gives R from change of T.  To do this,
>another constant (The 234.5 constant) is required.  This is the implied
>point of zero resistance for copper on the Celsius scale.  The formula we
>use is:
>
>  Rfinal - Rorig
>  x (234.5 + Tamb start) -(Tamb finsh - Tamb start)
>  Rorig
>
>The Tamb start and finish are the changes (if any) in Room ambient.  If the
>room remains at 20c then 234.5 + 20 is the multiplier.
>
>-Original Message-
>From: Robert Macy [mailto:m...@california.com]
>Sent: 13 May 2002 14:54
>To: Bill Ellingford; emc-p...@majordomo.ieee.org
>Subject: Re: Constant for Change of Resistance formula.
>
>
>Bill,
>
>Thanks for the site.
>
>Went there and found the same formula and constant I use.
>
>For copper, Temp Coeff = 3.9 x 10-3
>
>Then I clicked on table of coeff and there was a very long list of
>materials, but the temp coeff of copper there was 6.8 x 10-3  ???!!!
>
>Any ideas for this disparity?
>
>- Robert -
>
>   Robert A. Macy, PEm...@california.com
>   408 286 3985  fx 408 297 9121
>   AJM International Electronics Consultants
>   619 North First St,   San Jose, CA  95112
>
>
>-Original Message-
>From: Bill Ellingford 
>To: 'Colgan, Chris' ;
>emc-p...@majordomo.ieee.org 
>Date: Monday, May 13, 2002 5:38 AM
>Subject: RE: Constant for Change of Resistance formula.
>
>
>
>Hi Folks
>Further to the answer given, here is a little more data.
>The constant used is for the change of resistance with temperature.  metals
>and alloys (conductors) all exhibit a different constant.  This can be used
>for calculating temperature rise or resistance change.  i.e. find the temp
>rise from a start and finish test measurement on a winding (for example) at
>the begining and end of a on load heat run or, find R for a given temp:
>using a table or the formula, resistance at various temperatures can be
>pre-determined from a measurement made at one particular temperature.
>
>A website with the formulae can be found at
>http://hyperphysics.phy-astr.gsu.edu/hbase/electric/restmp.html
>
>Where you have a transition from one metal type to another, you must
measure
>each metal part individually.  If you have only two metals in contact, you
>may be able to apply a combination of the temp coefficient methods and
>transposition of the measurement of change of junction voltage formulae
i.e.
>Thermocouple laws.
>
>Hope this adds some value:  Bill Ellingford
>
>-Original Message-
>From: Colgan, Chris [mailto:chris.col...@tagmclaren.com]
>Sent: 13 May 2002 10:28
>To: emc-p...@majordomo.ieee.org
>Subject: RE: Constant for Change of Resistance formula.
>
>
>
>Ned is referring to the constant used in the "temperature rise calculated
by
>change in resistance formula" ie
>
><<...>>
>
>Where dt is the temperature r

RE: Constant for Change of Resistance formula.

2002-05-14 Thread Crabb, John

The formula for calculating the temperature rise of a winding
can be found in Annex E of IEC 60950 - and all the experts in 
TC 74 can't be wrong ?? (I also found the same formula in my
1985 issue of IEC 380.

Different temperature coefficents for copper have been quoted,
and as far as I can determine, the reason for this is as follows:

Definition of "temperature coefficent of resistance" is "the ratio
of the increase of resistance per degree C rise of temperature
to the resistance at 0 degrees C", and for a "standard annealed
copper conductor", this is 1/234.5 = .004264.

However, "the ratio of the increase of resistance per degree C 
rise of temperature to the resistance at 20 degrees C" is 
apparently .00393.

Regards,
John Crabb, Development Excellence (Product Safety) , 
NCR  Financial Solutions Group Ltd.,  Discovery Centre, 
3 Fulton Road, Dundee, Scotland, DD2 4SW
E-Mail :john.cr...@scotland.ncr.com
Tel: +44 (0)1382-592289  (direct ). Fax +44 (0)1382-622243.   


-Original Message-
From: Robert Macy [mailto:m...@california.com]
Sent: 13 May 2002 14:54
To: Bill Ellingford; emc-p...@majordomo.ieee.org
Subject: Re: Constant for Change of Resistance formula.



Bill,

Thanks for the site.

Went there and found the same formula and constant I use.

For copper, Temp Coeff = 3.9 x 10-3

Then I clicked on table of coeff and there was a very long list of
materials, but the temp coeff of copper there was 6.8 x 10-3  ???!!!

Any ideas for this disparity?

- Robert -

   Robert A. Macy, PEm...@california.com
   408 286 3985  fx 408 297 9121
   AJM International Electronics Consultants
   619 North First St,   San Jose, CA  95112

---
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/

To cancel your subscription, send mail to:
 majord...@ieee.org
with the single line:
 unsubscribe emc-pstc

For help, send mail to the list administrators:
 Ron Pickard:  emc-p...@hypercom.com
 Dave Heald:   davehe...@attbi.com

For policy questions, send mail to:
 Richard Nute:   ri...@ieee.org
 Jim Bacher: j.bac...@ieee.org

All emc-pstc postings are archived and searchable on the web at:
http://ieeepstc.mindcruiser.com/
Click on "browse" and then "emc-pstc mailing list"


Fw: Sr. EMC/EMI engineer position

2002-05-14 Thread John McBain

Hello All --

Any EMC specialist interested in working in Milpitas, California, may want
to contact the following recruiter (do not contact me!).  Apparently some
knowledge of  product safety and NEBS compliance also would be useful.  Good
luck!

Regards,  John McBain

- Original Message -
From: Isaac Ohana 
Sent: Monday, May 13, 2002 4:03 AM
Subject: Sr. EMC/EMI engineer


Hello John, Thank you for your time today.
I look forward to talk your friends and to work with you in the future.
Best Regards,

Isaac Ohana
Sr. Technical recruiter
Davis Search Group
416-782-7191 ext.211
www.davissearch.com



---
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/

To cancel your subscription, send mail to:
 majord...@ieee.org
with the single line:
 unsubscribe emc-pstc

For help, send mail to the list administrators:
 Ron Pickard:  emc-p...@hypercom.com
 Dave Heald:   davehe...@attbi.com

For policy questions, send mail to:
 Richard Nute:   ri...@ieee.org
 Jim Bacher: j.bac...@ieee.org

All emc-pstc postings are archived and searchable on the web at:
http://ieeepstc.mindcruiser.com/
Click on "browse" and then "emc-pstc mailing list"


Re: Constant for Change of Resistance formula.

2002-05-14 Thread Robert Macy

Bill,

Big oops.

Measuring the resistance to determine the temperature is not productive
*unless* the resistance dominates the resistance measurement.

Picture three equal valued resistances in a row.  The middle one gets very
hot (more than 100C rise) and increases over 40%, the two on the edges are
heat sinked and barely increase in temperature.  The resulting change in
resistance is 13% which implies the temperature in there has only gone up
around 33C.

Measuring the resistance doesn't tell you much.  At least with transformers
the dominant resistance is pretty much the bulk resistance.

- Robert -


-Original Message-
From: Bill Ellingford 
To: 'Robert Macy' ; emc-p...@majordomo.ieee.org

List-Post: emc-pstc@listserv.ieee.org
Date: Monday, May 13, 2002 1:56 PM
Subject: RE: Constant for Change of Resistance formula.


>
>Hi Robert / group
>OK, Not the best choice of website to demo the answer.  The differing
>figures are because the formula has been transposed to give Temp from
change
>of R from the original formula which gives R from change of T.  To do this,
>another constant (The 234.5 constant) is required.  This is the implied
>point of zero resistance for copper on the Celsius scale.  The formula we
>use is:
>
>  Rfinal - Rorig
>  x (234.5 + Tamb start) -(Tamb finsh - Tamb start)
>  Rorig
>
>The Tamb start and finish are the changes (if any) in Room ambient.  If the
>room remains at 20c then 234.5 + 20 is the multiplier.
>
>-Original Message-
>From: Robert Macy [mailto:m...@california.com]
>Sent: 13 May 2002 14:54
>To: Bill Ellingford; emc-p...@majordomo.ieee.org
>Subject: Re: Constant for Change of Resistance formula.
>
>
>Bill,
>
>Thanks for the site.
>
>Went there and found the same formula and constant I use.
>
>For copper, Temp Coeff = 3.9 x 10-3
>
>Then I clicked on table of coeff and there was a very long list of
>materials, but the temp coeff of copper there was 6.8 x 10-3  ???!!!
>
>Any ideas for this disparity?
>
>- Robert -
>
>   Robert A. Macy, PEm...@california.com
>   408 286 3985  fx 408 297 9121
>   AJM International Electronics Consultants
>   619 North First St,   San Jose, CA  95112
>
>
>-Original Message-
>From: Bill Ellingford 
>To: 'Colgan, Chris' ;
>emc-p...@majordomo.ieee.org 
>Date: Monday, May 13, 2002 5:38 AM
>Subject: RE: Constant for Change of Resistance formula.
>
>
>
>Hi Folks
>Further to the answer given, here is a little more data.
>The constant used is for the change of resistance with temperature.  metals
>and alloys (conductors) all exhibit a different constant.  This can be used
>for calculating temperature rise or resistance change.  i.e. find the temp
>rise from a start and finish test measurement on a winding (for example) at
>the begining and end of a on load heat run or, find R for a given temp:
>using a table or the formula, resistance at various temperatures can be
>pre-determined from a measurement made at one particular temperature.
>
>A website with the formulae can be found at
>http://hyperphysics.phy-astr.gsu.edu/hbase/electric/restmp.html
>
>Where you have a transition from one metal type to another, you must
measure
>each metal part individually.  If you have only two metals in contact, you
>may be able to apply a combination of the temp coefficient methods and
>transposition of the measurement of change of junction voltage formulae
i.e.
>Thermocouple laws.
>
>Hope this adds some value:  Bill Ellingford
>
>-Original Message-
>From: Colgan, Chris [mailto:chris.col...@tagmclaren.com]
>Sent: 13 May 2002 10:28
>To: emc-p...@majordomo.ieee.org
>Subject: RE: Constant for Change of Resistance formula.
>
>
>
>Ned is referring to the constant used in the "temperature rise calculated
by
>change in resistance formula" ie
>
><<...>>
>
>Where dt is the temperature rise, R1 is start resistance, R2 is end
>resistance, T1 is start ambient and T2 is end ambient.  234.5 is the
formula
>constant for copper.
>
>This formula is used extensively when heat testing transformers and coils.
>
>I'm afraid I don't know the constant for brass but I believe the figure may
>be related to the "inferred absolute zero" of a material.  Try asking a
>metallurgist?
>
>Regards
>
>Chris Colgan
>Compliance Engineer
>TAG McLaren Audio Ltd
>The Summit, Latham Road
>Huntingdon, Cambs, PE29 6ZU
>*Tel: +44 (0)1480 415 627
>*Fax: +44 (0)1480 52159
>* Mailto:chris.col...@tagmclaren.com
>* http://www.tagmclaren.com
>
>
>
>
>> -Original Message-
>> From: Robert Wilson [SMTP:robert_wil...@tirsys.com]
>> Sent: Friday, May 10, 2002 7:00 PM
>> To: emc-p...@majordomo.ieee.org; Ned Devine
>> Subject: RE: Constant for Change of Resistance formula.
>>
>> What are the units? 234.5 ...what?? Looking at what the units are, will
>> basically tell you exactly what the property is related to.
>>
>>
>>
>> Nonetheless, you cannot possibly directly determine w