Re: [time-nuts] Rpair of an 8662 HP Generator
With the improvement in the top radios today, it is no longer possible to test them with my 8662A / 8663A generators. I have had to move to 8642A / 8642B. However for more general sig gen needs, I prefer the 8662A. Rob NC0B -Original Message- From: time-nuts [mailto:time-nuts-boun...@febo.com] On Behalf Of KA2WEU--- via time-nuts Sent: Friday, April 15, 2016 7:26 PM To: paulsw...@gmail.com; time-nuts@febo.com Subject: Re: [time-nuts] Rpair of an 8662 HP Generator Thanks, I do not have a repair manual and I hate to abandon the signal generator after man years of positive use... Ulrich In a message dated 4/15/2016 9:14:21 P.M. Eastern Daylight Time, paulsw...@gmail.com writes: Ulrich To your first question I am unaware of anyone that repairs them. Though there are many places that might, typical cal and repair shops. I have a sick 8662 also (age) and need to dig in. At least I know the board and the most likely issue. The yahoo user group has a lot of useful details. Regards Paul WB8TSL On Fri, Apr 15, 2016 at 7:08 PM, Dr. Ulrich Rohde via time-nuts <_time-nuts@febo.com_ (mailto:time-nuts@febo.com) > wrote: Thanks Sent from my iPhone On Apr 15, 2016, at 6:38 PM, "John Miles" <_john@miles.io_ (mailto:j...@miles.io) > wrote: >> My signal generator has a poor, intermittent power supply. The RF >> section is ok. Who can please tell me which company can and will fix these >> older >> but excellent generators? >> >> Thanks, Ulrich > > Have a look at the capacitors in the voltage divider that drives the bases of the switching transistors. There's a good chance that's your problem, and they probably need to be replaced even if not. 75-TVA1607 (Mouser p/n for Sprague TVA1607) works well. > > It would be nice to use 105C parts if you can find some, but the 85C TVA1607s have worked well for me in multiple 8662As over several years. > > The other likely suspect is one of the large screw-terminal electrolytics on the motherboard. The power supply can almost-but-not-quite start up when one of those is completely open. High ESR is very likely to cause intermittent operation. > > -- john, KE5FX > Miles Design LLC > > ___ > time-nuts mailing list -- _time-nuts@febo.com_ (mailto:time-nuts@febo.com) > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. ___ time-nuts mailing list -- _time-nuts@febo.com_ (mailto:time-nuts@febo.com) To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there. ___ time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there. -- If this email is spam, report it to https://support.onlymyemail.com/view/report_spam/ODExMjI6MTg3Mjk2ODEzMzpyb2JAbmMwYi5jb206ZGVsaXZlcmVk ___ time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there.
Re: [time-nuts] Rpair of an 8662 HP Generator
If you 8662A has gone "bang" yet, replace the electrolytic on the three boards is a good starting place. Once it goes bang you will have a to start replacing solid-state devices. Rob, NC0B -Original Message- From: time-nuts [mailto:time-nuts-boun...@febo.com] On Behalf Of KA2WEU--- via time-nuts Sent: Friday, April 15, 2016 1:04 PM To: time-nuts@febo.com Subject: [time-nuts] Rpair of an 8662 HP Generator My signal generator has a poor, intermittent power supply. The RF section is ok. Who can please tell me which company can and will fix these older but excellent generators? Thanks, Ulrich ___ time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there. ___ time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there. -- If this email is spam, report it to https://support.onlymyemail.com/view/report_spam/ODExMjI6MTg3Mjk0MDA3NTpyb2JAbmMwYi5jb206ZGVsaXZlcmVk ___ time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there.
Re: [time-nuts] Rpair of an 8662 HP Generator
Thanks In a message dated 4/15/2016 10:02:01 P.M. Eastern Daylight Time, scmcgr...@gmail.com writes: Silicon investigations repairs them. Google the name for website Content by Scott Typos by Siri On Apr 15, 2016, at 6:38 PM, John Mileswrote: >> My signal generator has a poor, intermittent power supply. The RF >> section is ok. Who can please tell me which company can and will fix these >> older >> but excellent generators? >> >> Thanks, Ulrich > > Have a look at the capacitors in the voltage divider that drives the bases of the switching transistors. There's a good chance that's your problem, and they probably need to be replaced even if not. 75-TVA1607 (Mouser p/n for Sprague TVA1607) works well. > > It would be nice to use 105C parts if you can find some, but the 85C TVA1607s have worked well for me in multiple 8662As over several years. > > The other likely suspect is one of the large screw-terminal electrolytics on the motherboard. The power supply can almost-but-not-quite start up when one of those is completely open. High ESR is very likely to cause intermittent operation. > > -- john, KE5FX > Miles Design LLC > > ___ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. ___ time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there. ___ time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there.
Re: [time-nuts] Rpair of an 8662 HP Generator
Silicon investigations repairs them. Google the name for website Content by Scott Typos by Siri On Apr 15, 2016, at 6:38 PM, John Mileswrote: >> My signal generator has a poor, intermittent power supply. The RF >> section is ok. Who can please tell me which company can and will fix these >> older >> but excellent generators? >> >> Thanks, Ulrich > > Have a look at the capacitors in the voltage divider that drives the bases of > the switching transistors. There's a good chance that's your problem, and > they probably need to be replaced even if not. 75-TVA1607 (Mouser p/n for > Sprague TVA1607) works well. > > It would be nice to use 105C parts if you can find some, but the 85C TVA1607s > have worked well for me in multiple 8662As over several years. > > The other likely suspect is one of the large screw-terminal electrolytics on > the motherboard. The power supply can almost-but-not-quite start up when one > of those is completely open. High ESR is very likely to cause intermittent > operation. > > -- john, KE5FX > Miles Design LLC > > ___ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. ___ time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there.
Re: [time-nuts] Rpair of an 8662 HP Generator
Thanks, I do not have a repair manual and I hate to abandon the signal generator after man years of positive use... Ulrich In a message dated 4/15/2016 9:14:21 P.M. Eastern Daylight Time, paulsw...@gmail.com writes: Ulrich To your first question I am unaware of anyone that repairs them. Though there are many places that might, typical cal and repair shops. I have a sick 8662 also (age) and need to dig in. At least I know the board and the most likely issue. The yahoo user group has a lot of useful details. Regards Paul WB8TSL On Fri, Apr 15, 2016 at 7:08 PM, Dr. Ulrich Rohde via time-nuts <_time-nuts@febo.com_ (mailto:time-nuts@febo.com) > wrote: Thanks Sent from my iPhone On Apr 15, 2016, at 6:38 PM, "John Miles" <_john@miles.io_ (mailto:j...@miles.io) > wrote: >> My signal generator has a poor, intermittent power supply. The RF >> section is ok. Who can please tell me which company can and will fix these >> older >> but excellent generators? >> >> Thanks, Ulrich > > Have a look at the capacitors in the voltage divider that drives the bases of the switching transistors. There's a good chance that's your problem, and they probably need to be replaced even if not. 75-TVA1607 (Mouser p/n for Sprague TVA1607) works well. > > It would be nice to use 105C parts if you can find some, but the 85C TVA1607s have worked well for me in multiple 8662As over several years. > > The other likely suspect is one of the large screw-terminal electrolytics on the motherboard. The power supply can almost-but-not-quite start up when one of those is completely open. High ESR is very likely to cause intermittent operation. > > -- john, KE5FX > Miles Design LLC > > ___ > time-nuts mailing list -- _time-nuts@febo.com_ (mailto:time-nuts@febo.com) > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. ___ time-nuts mailing list -- _time-nuts@febo.com_ (mailto:time-nuts@febo.com) To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there. ___ time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there.
Re: [time-nuts] Rpair of an 8662 HP Generator
Ulrich To your first question I am unaware of anyone that repairs them. Though there are many places that might, typical cal and repair shops. I have a sick 8662 also (age) and need to dig in. At least I know the board and the most likely issue. The yahoo user group has a lot of useful details. Regards Paul WB8TSL On Fri, Apr 15, 2016 at 7:08 PM, Dr. Ulrich Rohde via time-nuts < time-nuts@febo.com> wrote: > Thanks > > Sent from my iPhone > > On Apr 15, 2016, at 6:38 PM, "John Miles"wrote: > > >> My signal generator has a poor, intermittent power supply. The RF > >> section is ok. Who can please tell me which company can and will fix > these > >> older > >> but excellent generators? > >> > >> Thanks, Ulrich > > > > Have a look at the capacitors in the voltage divider that drives the > bases of the switching transistors. There's a good chance that's your > problem, and they probably need to be replaced even if not. 75-TVA1607 > (Mouser p/n for Sprague TVA1607) works well. > > > > It would be nice to use 105C parts if you can find some, but the 85C > TVA1607s have worked well for me in multiple 8662As over several years. > > > > The other likely suspect is one of the large screw-terminal > electrolytics on the motherboard. The power supply can > almost-but-not-quite start up when one of those is completely open. High > ESR is very likely to cause intermittent operation. > > > > -- john, KE5FX > > Miles Design LLC > > > > ___ > > time-nuts mailing list -- time-nuts@febo.com > > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > > and follow the instructions there. > ___ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. > ___ time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there.
Re: [time-nuts] Rpair of an 8662 HP Generator
Thanks Sent from my iPhone On Apr 15, 2016, at 6:38 PM, "John Miles"wrote: >> My signal generator has a poor, intermittent power supply. The RF >> section is ok. Who can please tell me which company can and will fix these >> older >> but excellent generators? >> >> Thanks, Ulrich > > Have a look at the capacitors in the voltage divider that drives the bases of > the switching transistors. There's a good chance that's your problem, and > they probably need to be replaced even if not. 75-TVA1607 (Mouser p/n for > Sprague TVA1607) works well. > > It would be nice to use 105C parts if you can find some, but the 85C TVA1607s > have worked well for me in multiple 8662As over several years. > > The other likely suspect is one of the large screw-terminal electrolytics on > the motherboard. The power supply can almost-but-not-quite start up when one > of those is completely open. High ESR is very likely to cause intermittent > operation. > > -- john, KE5FX > Miles Design LLC > > ___ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. ___ time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there.
Re: [time-nuts] Rpair of an 8662 HP Generator
On 4/15/2016 3:38 PM, John Miles wrote: Have a look at the capacitors in the voltage divider that drives the bases of the switching transistors. There's a good chance that's your problem, and they probably need to be replaced even if not. 75-TVA1607 (Mouser p/n for Sprague TVA1607) works well. John for what it's worth both my 8662 and 8663 which are late (1994) revisions have these single 15 µF caps replaced by two parallel 10 µFs among other changes in the power supply area. Maybe worth doing if they can be fitted... Dan ___ time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there.
Re: [time-nuts] Rpair of an 8662 HP Generator
> My signal generator has a poor, intermittent power supply. The RF > section is ok. Who can please tell me which company can and will fix these > older > but excellent generators? > > Thanks, Ulrich Have a look at the capacitors in the voltage divider that drives the bases of the switching transistors. There's a good chance that's your problem, and they probably need to be replaced even if not. 75-TVA1607 (Mouser p/n for Sprague TVA1607) works well. It would be nice to use 105C parts if you can find some, but the 85C TVA1607s have worked well for me in multiple 8662As over several years. The other likely suspect is one of the large screw-terminal electrolytics on the motherboard. The power supply can almost-but-not-quite start up when one of those is completely open. High ESR is very likely to cause intermittent operation. -- john, KE5FX Miles Design LLC ___ time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there.
Re: [time-nuts] HP 5360A History?
Are you referring to the 00105-6013? I have one and its info. Ignacio, EB4APL El 15/04/2016 a las 8:31, cdel...@juno.com escribió: The oscillator used in the 5360A is the same one used in the early 105A, the 5061A and the 5065A. I just call them the 105 style oscillator. They have a 115 volt thermoswitch controlled fast warmup heater that is set to switch out once it gets close to the set point of the proportional controller oven that stays on all the time. 10811A units are better performers. Cheers, Corby ___ time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there.
Re: [time-nuts] Fast risetime pulse generator
Hi all, Brooke Clark wrote: > In the HP 5100 Frequency Synthesizer they use a heater on the SRD in order to > increase it's > lifetime so that the 3 MHz drive will generate the needed harmonics. > http://www.prc68.com/I/HP5100.shtml These > assemblies are no longer available and getting an SRD with a lifetime long > enough to work with a 3 MHz input is a common > as chicken teeth. A solution is to get a large stud mounted Varactor and > these do have long lifetimes. I was so pleased to see this. It mirrored my experience exactly. A few years ago I restored a 5100A and its 5110A exciter. The SRD heater in the 5110A had developed a failed thermostat and was a charred blob. I made up a test jig and tested lots and lots of different diodes and as it turned out, the best candidate was a stud mounted varactor out of a junked transmitter from the old VHF cellphone system. It's worked like a charm ever since. This post (and webpage) has explained it beautifully. Morris ___ time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there.
Re: [time-nuts] Fast risetime pulse generator
Bill Now this discussion gets really interesting. You are doing something and will report results. You have my attention. Regards Paul WB8TSL On Fri, Apr 15, 2016 at 12:27 PM, BIll Ezellwrote: > I'm amazed at how much traffic this has generated. > Thanks Brooke, you're exactly on-topic as usual. > > As I said originally, I wanted a quick-and-dirty-and-cheap, and I'm using > an SMMD-835 SRD. Making up the test layout tonight. > I'm starting with a 100 uA bias on the diode, and a 20 pF diode-to-load > coupling cap, driving the SRD with a 5V 1Mhz square-wave. > > I'm also going to try the 2N3904 Vce-reverse-bias avalanche idea just for > fun. > > I'll post again after I have some results. > > PS - as I've mentioned before, I trim these things. I hate the > endlessly-requoted replies, impossible to parse what's actually been said. > > --- Brooke post, trimmed -- > Hi: > > The idea with a step recovery diode is that during forward bias the diode > stores charge. When the polarity reverses the diode continues to conduct > until the charge has been depleted. At that time the diode stops > conducting and opens. > Suppose this was done using a square wave. The length of the positive > (charging) half of waveform needs to be long enough so that the diode if > fully charged (a function of the diode lifetime and charging current). > Then on the negative half (discharging) when all the charge has been > extracted the diode turns of making a sharp edge. So the diode lifetime is > important in relation to input drive frequency. For a sinewave drive the > lifetime should be about 1/4 the period at the recommended drive level. > > -- > Bill Ezell > -- > The day Microsoft makes something that doesn't suck > will be the day they make vacuum cleaners. > Or maybe Windows 10. > > ___ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. > ___ time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there.
[time-nuts] Rpair of an 8662 HP Generator
My signal generator has a poor, intermittent power supply. The RF section is ok. Who can please tell me which company can and will fix these older but excellent generators? Thanks, Ulrich ___ time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there. ___ time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there.
[time-nuts] Science of Time, early registration deadline
Today is the early registration deadline (saving $75) for The Science of Time symposium, to be held by the Harvard-Smithsonian from 5-9 June 2016: http://sot2016.cfa.harvard.edu Your participation would be welcome, and in good company: http://sot2016.cfa.harvard.edu/Program/registrants.html Carpe Diem! Rob Seaman Catalina Sky Survey Lunar and Planetary Laboratory ___ time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there.
Re: [time-nuts] Fast risetime pulse generator
I'm amazed at how much traffic this has generated. Thanks Brooke, you're exactly on-topic as usual. As I said originally, I wanted a quick-and-dirty-and-cheap, and I'm using an SMMD-835 SRD. Making up the test layout tonight. I'm starting with a 100 uA bias on the diode, and a 20 pF diode-to-load coupling cap, driving the SRD with a 5V 1Mhz square-wave. I'm also going to try the 2N3904 Vce-reverse-bias avalanche idea just for fun. I'll post again after I have some results. PS - as I've mentioned before, I trim these things. I hate the endlessly-requoted replies, impossible to parse what's actually been said. --- Brooke post, trimmed -- Hi: The idea with a step recovery diode is that during forward bias the diode stores charge. When the polarity reverses the diode continues to conduct until the charge has been depleted. At that time the diode stops conducting and opens. Suppose this was done using a square wave. The length of the positive (charging) half of waveform needs to be long enough so that the diode if fully charged (a function of the diode lifetime and charging current). Then on the negative half (discharging) when all the charge has been extracted the diode turns of making a sharp edge. So the diode lifetime is important in relation to input drive frequency. For a sinewave drive the lifetime should be about 1/4 the period at the recommended drive level. -- Bill Ezell -- The day Microsoft makes something that doesn't suck will be the day they make vacuum cleaners. Or maybe Windows 10. ___ time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there.
Re: [time-nuts] Lea-6T Math error?
Did you use the CFG-TMODE svinVarLimit to set 350 and the svinMinDur to set 32400? The CFG-TMODE svinVarLimit is a variance in squared mm, not a standard deviation... On Fri, Apr 15, 2016 at 6:27 AM, Bob Stewartwrote: > I hope I haven't posted about this before, but it's something that's bugged > me for awhile. When I run a survey-in on an LEA-6T, the > CFG-TMODE2->fixedPosAcc value seems to be wrong. For instance, I ran a > survey today and told it to run for 32,400 seconds and reach at least 350mm > accuracy. When it finished, TIM-SVIM reported it had run 38,126 seconds with > a mean 3D std deviation of 350mm. OK, so far. So, I go look in the > CFG-TMODE2 structure and it reports 122mm. That seems a bit odd, but OK. > And then I squared 350, and came up with 122500. It appears that instead of > just moving the 350mm value over to the TMODE structure, they've squared it > and divided it by 1000. I've noticed this happening on at least 2 versions > of the 6T. I suppose I should do the same test on the M8T. > > Bob > ___ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. ___ time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there.
Re: [time-nuts] Lea-6T Math error?
Yes, I see: the fixedPosAcc should be 18mm not 122. On Fri, Apr 15, 2016 at 12:15 PM, Azelio Borianiwrote: > Did you use the CFG-TMODE svinVarLimit to set 350 and the svinMinDur > to set 32400? > The CFG-TMODE svinVarLimit is a variance in squared mm, not a standard > deviation... > > On Fri, Apr 15, 2016 at 6:27 AM, Bob Stewart wrote: >> I hope I haven't posted about this before, but it's something that's bugged >> me for awhile. When I run a survey-in on an LEA-6T, the >> CFG-TMODE2->fixedPosAcc value seems to be wrong. For instance, I ran a >> survey today and told it to run for 32,400 seconds and reach at least 350mm >> accuracy. When it finished, TIM-SVIM reported it had run 38,126 seconds >> with a mean 3D std deviation of 350mm. OK, so far. So, I go look in the >> CFG-TMODE2 structure and it reports 122mm. That seems a bit odd, but OK. >> And then I squared 350, and came up with 122500. It appears that instead of >> just moving the 350mm value over to the TMODE structure, they've squared it >> and divided it by 1000. I've noticed this happening on at least 2 versions >> of the 6T. I suppose I should do the same test on the M8T. >> >> Bob >> ___ >> time-nuts mailing list -- time-nuts@febo.com >> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >> and follow the instructions there. ___ time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there.
[time-nuts] HP 5360A History?
The oscillator used in the 5360A is the same one used in the early 105A, the 5061A and the 5065A. I just call them the 105 style oscillator. They have a 115 volt thermoswitch controlled fast warmup heater that is set to switch out once it gets close to the set point of the proportional controller oven that stays on all the time. 10811A units are better performers. Cheers, Corby ___ time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there.
[time-nuts] Lea-6T Math error?
I hope I haven't posted about this before, but it's something that's bugged me for awhile. When I run a survey-in on an LEA-6T, the CFG-TMODE2->fixedPosAcc value seems to be wrong. For instance, I ran a survey today and told it to run for 32,400 seconds and reach at least 350mm accuracy. When it finished, TIM-SVIM reported it had run 38,126 seconds with a mean 3D std deviation of 350mm. OK, so far. So, I go look in the CFG-TMODE2 structure and it reports 122mm. That seems a bit odd, but OK. And then I squared 350, and came up with 122500. It appears that instead of just moving the 350mm value over to the TMODE structure, they've squared it and divided it by 1000. I've noticed this happening on at least 2 versions of the 6T. I suppose I should do the same test on the M8T. Bob ___ time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there.