Re: [time-nuts] PRS10 rubidium lamp module failure
Follow-up: as several time-nuts noted, there was a lot of corrosion on parts of the lamp module. Oddly, there was no corrosion on the rest of the PRS10, or on the TS2500 which it came out of. The resistor lead was severely oxidized and partly melted. It's possible that the inductance of the wire-wound resistor helped sustain an arc. Alternatively, the grounded lead may have arced to the hot end of the lamp drive coil, although the coil itself showed very little damage. Perhaps ozone from an arc accelerated the corrosion. Dunno, it's a mystery. The resistor is specified at 2W, but it's much smaller than a typical 2W resistor. I removed the damaged lead and replaced it with a bit of 16 AWG magnet wire. This gave mechanical support as well as an electrical connection. Buttoned it all back up and applied power, and the lamp lit almost immediately (it wasn't lit before) and it locked quickly. Tuned it to match a GPSDO, and it seems to be working fine. A note on tuning: the documentation says that the "SF" tuning command has a resolution of 1E-12. That's true, but the actual available tuning resolution seems to be more like 1E-11. The actual tuning is done by adjusting the C-field with a 12-bit DAC. The "MR?" command gives the present value of that DAC. Unfortunately, the full range of the SF command (4000 counts) only moves the DAC about 600 counts. You can increment the SF command several times before seeing a response. Anyone planning to use a PRS10 inside a control loop should be aware of this quirk. Cheers! --Stu ___ time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there.
Re: [time-nuts] High rate, high precision/accuracy time interval counter methods
Am 09.05.2016 um 10:08 schrieb Magnus Danielson: Hi, On 05/08/2016 09:53 PM, Attila Kinali wrote: True and not true. Yes, there are many ADCs that do high conversion rates, but these are optimized for piplined applications where conversion happens at a constant rate. Ie they expect a constant conversion clock with a constant rate. If you want to trigger conversion at an arbitrary time, you either have to build your own sampler or need to use one of the non-pipelined ADCs whic are much slower (IIRC they stop around 5-10Msps aka >100ns conversion time). Flash ADCs with direct access to the sampling circuitry are basically extinct. You can let the ADC convert as a continuous process as long as you filter out the samples you are interested in. And, at least for the LTC2165, you could really use the encode clock to read the result of the capacitor immediately. All one would have to do is to flush the pipeline with 5 or 6 clocks after the encode, with a cycle time not much faster than 10 ns and not slower than 1 usec. That seems not too hard to do but costs a differential mux in the encode lines. We then would not need to switch the current source, it could degenerate to a 20V source with a resistor & safety clamps; that would be as linear as it gets. The reset level could be clamped by a Sky or Avago phemt, they can swallow the current easily, are blinding fast and do not drop a voltage for the base current over Re (from Gummel-Poon model). And an inverse transistor may have less Re, but also less beta and therefore more base current. I have already built a time stretcher and was restricted to spaceworthy components; those Intersil transistor arrays had the only acceptable PNPs. These flatpacks are huge, and the legs are endless; anything more complicated than a simple current mirror would oscillate or would at least lurch to its destination, and you cannot simply damp it to death. Not funny. Minimizing the E-E-inductance: The next best thing to an onchip-connection might be abusing 2 Infineon SiGETs on the 2 sides of a thin board; they are optimized for low emitter inductance and have 2 emitter legs each. And their Early voltage is huge, so they are not impressed by the VCE changes of the current source. (SiGe BFP650, 750 etc, available in a tiny package) regards, Gerhard ___ time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there.
[time-nuts] Microchip PIC CTMU
I ran a search through the time-nuts archives and found only one mention of this a couple years ago asking the same question which is, has anybody explored the capabilities of the Microchip PIC CTMU for use as a time to digital converter or programmable delay? The Microchip datasheets lack detailed performance specifications but my conservative estimate is that resolution down to 500ps over 200ns using the built in 10 bit ADC should be possible without complex calibration. (The PIC series ADCs with more than 10 bits are pretty horrible.) See What You Can Do with the CTMU: http://ww1.microchip.com/downloads/en/AppNotes/CTMU%2001375a.pdf Overview of Charge Time Measurement Unit (CTMU): http://www.microchip.com/stellent/groups/SiteComm_sg/documents/DeviceDoc/en542792.pdf PIC18(L)F2X/4XK22 w/10 Bit ADC: http://ww1.microchip.com/downloads/en/DeviceDoc/39977f.pdf PIC18F66K80 w/12 Bit ADC: http://ww1.microchip.com/downloads/en/DeviceDoc/39977f.pdf I prefer more discrete implementations but this might be useful for its higher integration, lower cost, and simplicity in less demanding applications. ___ time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there.
Re: [time-nuts] HP 105B oscillator
Perry, There are several op/svc manuals for the HP 105B. Original models used the 00105 brick oscillator; newer models the 10811. Check the S/N of your manual against the S/N of your hp 105B. Yes, the 105B makes a good test bed. For years I used one to measure lots of 10811's. /tvb - Original Message - From: "Perry Sandeen via time-nuts"To: Sent: Tuesday, May 10, 2016 12:10 AM Subject: [time-nuts] HP 105B oscillator List, I purchased a HP 105B oscillator on ebay and it arrived today. Previous to getting it I D/L the manual and brief;y perused it. So when I popped open the top cover to see the internal parts, what to my wondering eyes appeared Not the A1 module shown in the manual but a 10811 lookalike. The only information on the oscillator is S/N 3010A59493 and R11215k. So what do I have? The next question that comes to mind is the plug in base the same electrical configuration of the 10811? (Being a TN You can see where this is going.) Leaving out the divide issue for the moment, this would seem to be an ideal test bed for my 10811 oscillators as the 3 in my 5370B's have no EFC adjustment installed and I have another 10811 spare. I have enough 5370B's and 5335's in which I could put all my 10811's in to *incubate* for months before attempting to tune them. Also are there upgrades that are practical to add to the present PS for greater stability. I am going to replace the original screw cap capacitor with a new Nichicon 105C cap. Thoughts? Regards, Perrier ___ time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there.
Re: [time-nuts] PRS10 rubidium lamp module failure
I agree with Arthur. It's not possible that such temperature had happened. My unit had a severe board and components electrochemical corrosion probably caused by a mix of factors as flux residues action and the pcb coating that simple dissapear. I think that all together under hot conditions in a small enclosure is the most probably reason to that failure. Also, in picture it can be seen clearly many spots with electrochemical corrosion. - Original Message - From: "Arthur Dent"To: Sent: Tuesday, May 10, 2016 1:06 AM Subject: [time-nuts] PRS10 rubidium lamp module failure ..."apparently got hot enough to create a localized thermal runaway in its lead wire." I don't think that explanation is correct. There are no signs of discoloration caused by overheating anywhere else in the photo and yet the lead wire has apparently melted as you say. That would probably require at least a 2000 degree temp which would cause lots of damage. Are there any signs of chemical contamination that could have caused this? Not the same but In older equipment that had leaky Hg batteries I've seen similar damage. -Arthur ___ time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there. ___ time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there.
Re: [time-nuts] PRS10 rubidium lamp module failure
On Mon, 9 May 2016 15:47:06 -0700 Stewart Cobbwrote: > A half-melted bit of wire fell > away before I could take the picture. Without that heater working, the > lamp is apparently too cold to start. This looks more like it corroded away than melted. Though, it's a bit hard to judge with that angle and only one picture. Attila Kinali -- It is upon moral qualities that a society is ultimately founded. All the prosperity and technological sophistication in the world is of no use without that foundation. -- Miss Matheson, The Diamond Age, Neil Stephenson ___ time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there.
Re: [time-nuts] High rate, high precision/accuracy time interval counter methods
Hoi Bruce, On Mon, 9 May 2016 23:34:24 + (UTC) Bruce Griffithswrote: > Its probably easier/cheaper to construct a suitable filter for a GSPS ADC > than to construct a TAC that is fast enough to suit an ADC with a GHz clock. Probably. > Minimising the emitter to emitter inductance of a longtailed pair or > equivalent is key to achieving a fast enough switching time for a suitable > TAC. How does one minimize this inductance? The only way I am aware of is to use a package with a pair of transistors with common emiter pin. > The somewhat heroic measures employed in the Wavecrest counters is > perhaps the limit of discrete construction techniques. Are these heroic measures documented somewhere? So far I have not even seen a block diagram of a Wavecrest counter, much less a schematic. Attila kinali -- It is upon moral qualities that a society is ultimately founded. All the prosperity and technological sophistication in the world is of no use without that foundation. -- Miss Matheson, The Diamond Age, Neil Stephenson ___ time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there.
Re: [time-nuts] High rate, high precision/accuracy time interval counter methods
Hoi Attila Yes, the only way to reduce emitter-emitter inductance is indeed to connect them on the die. Its even better if the current source transistor collector is also connected to the common emitter node of the long tailed pair on the die as this minimises the capacitance at this node. Alternatively one can add a small series resistance to reduce the associated emitter circuit time constant as Wavecrest do. The Wavecrest patent is somewhat detailed and complete with reasonably detailed circuit diagrams and explanatory text. US6185509 US6194925 *** US6226231US4908784 Bruce On Tuesday, 10 May 2016 9:17 PM, Attila Kinaliwrote: Hoi Bruce, On Mon, 9 May 2016 23:34:24 + (UTC) Bruce Griffiths wrote: > Its probably easier/cheaper to construct a suitable filter for a GSPS ADC > than to construct a TAC that is fast enough to suit an ADC with a GHz clock. Probably. > Minimising the emitter to emitter inductance of a longtailed pair or > equivalent is key to achieving a fast enough switching time for a suitable > TAC. How does one minimize this inductance? The only way I am aware of is to use a package with a pair of transistors with common emiter pin. > The somewhat heroic measures employed in the Wavecrest counters is > perhaps the limit of discrete construction techniques. Are these heroic measures documented somewhere? So far I have not even seen a block diagram of a Wavecrest counter, much less a schematic. Attila kinali -- It is upon moral qualities that a society is ultimately founded. All the prosperity and technological sophistication in the world is of no use without that foundation. -- Miss Matheson, The Diamond Age, Neil Stephenson ___ time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there.
[time-nuts] HP 105B oscillator
List, I purchased a HP 105B oscillator on ebay and it arrived today. Previous to getting it I D/L the manual and brief;y perused it. So when I popped open the top cover to see the internal parts, what to my wondering eyes appeared Not the A1 module shown in the manual but a 10811 lookalike. The only information on the oscillator is S/N 3010A59493 and R11215k. So what do I have? The next question that comes to mind is the plug in base the same electrical configuration of the 10811? (Being a TN You can see where this is going.) Leaving out the divide issue for the moment, this would seem to be an ideal test bed for my 10811 oscillators as the 3 in my 5370B's have no EFC adjustment installed and I have another 10811 spare. I have enough 5370B's and 5335's in which I could put all my 10811's in to *incubate* for months before attempting to tune them. Also are there upgrades that are practical to add to the present PS for greater stability. I am going to replace the original screw cap capacitor with a new Nichicon 105C cap. Thoughts? Regards, Perrier ___ time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there.